Search found 202 matches
- Wed Jan 04, 2023 12:28 am
- Forum: Microscopes and optics
- Topic: Can I remove an excitation filter to increase auto fluorescence?
- Replies: 5
- Views: 1152
Re: Can I remove an excitation filter to increase auto fluorescence?
You really don't want to remove the excitation filter. Although about inch of glass will prevent the transmission of harmful UV rays, the intensity would overwhelm the retina. In addition, the emission from autofluorence is still wavelength dependent. In otherwords it needs the excitation to generat...
- Mon Dec 26, 2022 10:27 pm
- Forum: Microscopes and optics
- Topic: Two questions regarding water immersion
- Replies: 35
- Views: 5505
Re: Two questions regarding water immersion
Thanks for the comment and good luck with your investigations.
- Mon Dec 26, 2022 3:29 am
- Forum: Microscopes and optics
- Topic: Two questions regarding water immersion
- Replies: 35
- Views: 5505
Re: Two questions regarding water immersion
The Biostar was never offered with Leica branded objectives, although it was infinity corrected so someone may have built a system with objective from an DM IRB. On this point I am positive you are wrong, to the extent that it matters. I never bought an 1820 from the manufacturer, but you frequentl...
- Sun Dec 25, 2022 1:31 am
- Forum: Microscopes and optics
- Topic: Two questions regarding water immersion
- Replies: 35
- Views: 5505
Re: Two questions regarding water immersion
Phil, I agree completely.
- Sat Dec 24, 2022 12:20 am
- Forum: Microscopes and optics
- Topic: Two questions regarding water immersion
- Replies: 35
- Views: 5505
Re: Two questions regarding water immersion
The Biostar was never offered with Leica branded objectives, although it was infinity corrected so someone may have built a system with objective from an DM IRB. However, it would require RMS/M25 adapter rings to fit the objective into the nosepiece. Of course, it would have been possible to mount L...
- Mon Nov 28, 2022 1:49 am
- Forum: Microscopes and optics
- Topic: 20 mm working distance air objective, 10x, 0.50 NA
- Replies: 10
- Views: 2780
Re: 20 mm working distance air objective, 10x, 0.50 NA
The article referenced in association with this objective specifically mentioned multi - photon applications. Terminology associated with Confocal Microscopy. I suspect the low mag is desired because of thick tissue samples and the high NA is beneficial because it will aid in presenting an intensity...
- Tue Nov 15, 2022 8:28 pm
- Forum: Microscopes and optics
- Topic: Reichert 1711
- Replies: 2
- Views: 843
Re: Reichert 1711
I had some old literature that provided some detailed information. After the specs is my commentary of the applications and image information. 1711 was used with Auramine stain. (436/450/GG475) This was common for TB bacillus that had excitation in the 436nm range and emissions in the in the yellow ...
- Tue Nov 15, 2022 4:57 am
- Forum: Microscopes and optics
- Topic: 20 mm working distance air objective, 10x, 0.50 NA
- Replies: 10
- Views: 2780
Re: 20 mm working distance air objective, 10x, 0.50 NA
It seems this objective would lend itself to applications with confocal microscopy. High NA provides better light amplitude combined with higher DOF associated with the lower magnification. Thicker samples could be viewed, while eliminating the noise in the interstial tissue areas.
- Thu Nov 03, 2022 4:00 am
- Forum: Microscopes and optics
- Topic: the scale on the condenser's aperture
- Replies: 27
- Views: 5392
Re: the scale on the condenser's aperture
This is a great discussion when it comes to comparing theoretical vs. practical real world application of the condenser diaphragm adjustment. Most of the major manufacturers have used some form of reference to the objective, in use, on the condenser, giving the user a reference point. It doesn’t mat...
- Fri Aug 19, 2022 1:02 am
- Forum: Microscopes and optics
- Topic: Difference between Olympus UVFL and DPlan UV objectives.
- Replies: 4
- Views: 1270
Re: Difference between Olympus UVFL and DPlan UV objectives.
Pete,
Sorry I don’t know the specifics on those objectives. It might be possible to search for the spectral curve on each objective, ideally using the cat number if you happen to have it
Sorry I don’t know the specifics on those objectives. It might be possible to search for the spectral curve on each objective, ideally using the cat number if you happen to have it
- Fri Aug 12, 2022 3:00 am
- Forum: Microscopes and optics
- Topic: Difference between Olympus UVFL and DPlan UV objectives.
- Replies: 4
- Views: 1270
Re: Difference between Olympus UVFL and DPlan UV objectives.
As Pete M Indicated, the UVFL objectives are for vertical microscopy, which really is reflected light, not transmitted applications. I don't have the spectral curves for these objectives, but typically a UV reference is good to something in the range of 365nm. This is typical for a good biological r...
- Fri Aug 12, 2022 2:39 am
- Forum: Microscopes and optics
- Topic: What is the magnification?
- Replies: 10
- Views: 2278
Re: What is the magnification?
Did any of the proposed calculation provide an answer that was comparable?
- Thu Jul 28, 2022 10:28 pm
- Forum: Microscopes and optics
- Topic: What is the magnification?
- Replies: 10
- Views: 2278
Re: What is the magnification?
If you want to know the final magnification, measure the field of view in mm and decide by the FOV of the eyepiece. So if the FOV through microscope is 18mm the total mag is 1x. If the FOV is 1.8mm your total mag is 10x. Etc as long as you know the FOV of the eyepiece, you can determine the total ma...
- Sat May 14, 2022 2:06 am
- Forum: Microscopes and optics
- Topic: Convert a biological microscope into a wafer microscope?
- Replies: 8
- Views: 1904
Re: Convert a biological microscope into a wafer microscope?
Keep in mind almost all industrial microscopes are augmented biological scopes with accommodation for illumination path and specimen manipulation. Then rebranded to reflect those differences
- Sat May 14, 2022 2:02 am
- Forum: Microscopes and optics
- Topic: Leica R.M.S. objective.
- Replies: 12
- Views: 2015
Re: Leica R.M.S. objective.
I can tell you this is definitely not an adapted Reichert industrial objective from the Vistek Polyvar range. When the Vistek name was implemented as a spin off from Leica, the Vienna factory was already out of the microscope business. The continue to make EM instrumentation, today. The 11767xxx cat...
- Fri Apr 15, 2022 3:23 am
- Forum: Microscopes and optics
- Topic: American Optical reflecting objective model 1200
- Replies: 9
- Views: 2848
Re: American Optical reflecting objective model 1200
A couple of additional points of reference relative to this objective. Not only does this objective need a condenser capable of transmitting UV the entire illumination system must also support UV transmission. If this objective was designed for transmitted light applications such as fluorescence, th...
- Thu Mar 24, 2022 2:07 am
- Forum: Microscopes and optics
- Topic: Aligning LED in Leica DMIL
- Replies: 5
- Views: 1906
Re: Aligning LED in Leica DMIL
Is the question how to determine the centration of the optical path, or what is the best way to me mechanically position the LED element? If it is the previous, recall there are conjugate planes in the microscope, of which one is the illumination plane. The objective back aperture, aperture diaphrag...
- Mon Feb 07, 2022 5:21 am
- Forum: Microscopes and optics
- Topic: Reichert zetopan questions
- Replies: 39
- Views: 10185
Re: Reichert zetopan questions
As an addendum to apochronaut’s thread, parfocality on a Reichert Zetopan and later on the Univar/Polyvar scopes was achieved by hand honing the shoulder of the objective. This was specific to the position in the nosepiece and each scope. It was one of the reasons the objective mag position was mark...
- Tue Feb 01, 2022 12:21 am
- Forum: Microscopes and optics
- Topic: Advice on projection system light house 250
- Replies: 11
- Views: 3374
Re: Advice on projection system light house 250
with that spectral curve, that is not a halogen lamp. Typically halogen lamps use a filament. The Leitz projectiion scope usually had a XBO lamp depending on vintage. 150W most likely but early systems used a 450XBO. You can't miss it. It is very large as is the power supply. Also that combo often m...
- Sun Jan 23, 2022 9:16 pm
- Forum: Microscopes and optics
- Topic: Reichert zetopan questions
- Replies: 39
- Views: 10185
Re: Reichert zetopan questions
Absolutely correct. I was oversimplifying, Thanks for clarifying.
- Sun Jan 23, 2022 8:12 pm
- Forum: Microscopes and optics
- Topic: Reichert zetopan questions
- Replies: 39
- Views: 10185
Re: Reichert zetopan questions
Generally when you encounter a short coming such as the stage x-y control there is often a corresponding plus. In this case, the stage level coaxial stage control offer a bilateral access so you can use either the left or right hand to operate the stage movement. It might not be a plus for you, but ...
- Fri Jan 21, 2022 10:21 pm
- Forum: Microscopes and optics
- Topic: Reichert zetopan questions
- Replies: 39
- Views: 10185
Re: Reichert zetopan questions
I haven’t seen a two diaphragm condenser for the Zetopan since the 70, As Phil pointed out, the bottom iris functions as a field diaphragm, while the upper iris is the aperture diaphragm. The dark field stop would need to be at that conjugate plane. I don’t think there is a practical way to accompli...
- Wed Jan 05, 2022 12:23 am
- Forum: Microscopes and optics
- Topic: What was this weird Leitz system used for?
- Replies: 7
- Views: 2451
Re: What was this weird Leitz system used for?
I just took a look at the bid site and a side picture. This system has a unique to Leitz ( Leica) component installed between the lamp house and the microscope stand. It is a variable color filter system that was used to help balance the color between the left and right scopes. As I mentioned earlie...
- Wed Jan 05, 2022 12:19 am
- Forum: Microscopes and optics
- Topic: What was this weird Leitz system used for?
- Replies: 7
- Views: 2451
Re: What was this weird Leitz system used for?
This particular system, simply known as a comparison microscope, is used in the trace evidence labs for forensic investigations. Typical samples would include hair, fibers, glass and paint chips. To a lesser extent, building materials, minerals and other samples that would be suitable for use on a P...
- Fri Dec 31, 2021 7:17 pm
- Forum: Microscopes and optics
- Topic: Manuals For American Optical Microscopes
- Replies: 9
- Views: 3344
Re: Manuals For American Optical Microscopes
Phil's description is spot on. I have not found any in my stash in any case.
- Tue Dec 28, 2021 8:00 pm
- Forum: Microscopes and optics
- Topic: Manuals For American Optical Microscopes
- Replies: 9
- Views: 3344
Re: Manuals For American Optical Microscopes
I used to have a couple of different types that were essentially homemade. I'll see if I can find one in an old tool kit and take a picture. One was wooden, with a soft metal strap, that other was all metal.
- Tue Dec 28, 2021 7:54 pm
- Forum: Microscopes and optics
- Topic: Stereo microscope photo port projection lens not parfocal with eyepieces
- Replies: 38
- Views: 10377
Re: Stereo microscope photo port projection lens not parfocal with eyepieces
Regardless of the microscope model, one factor that should be considered in low mag photomicrography is the parallax that can occur to the conjugate image plane Does the viewing eyepiece have a reticle that can be focused with to the eye? If not, it can be critical to the focus process of the scope ...
- Tue Dec 28, 2021 7:31 pm
- Forum: Microscopes and optics
- Topic: Manuals For American Optical Microscopes
- Replies: 9
- Views: 3344
Re: Manuals For American Optical Microscopes
You will need an objective wrench in order to get them off. An objective wrench in similar in design to a jar opener, only the scale is brought down to fit into the nosepiece and objective.
- Tue Dec 28, 2021 7:29 pm
- Forum: Microscopes and optics
- Topic: Manuals For American Optical Microscopes
- Replies: 9
- Views: 3344
Re: Manuals For American Optical Microscopes
If you are trying to take the objectives off the Serise 50/150 or Series 60/160, you may encounter some difficulty. The objectives on those models were tightened into the nosepiece with objective wrench. This prevented students from removing an subsequently damaging an objective. It they were instal...
- Wed Dec 15, 2021 12:02 am
- Forum: Microscopes and optics
- Topic: measuring eyepiece
- Replies: 8
- Views: 2240
Re: measuring eyepiece
Which eyepiece do you currently have. The 10x/20 eyepiece for the DMLB will accept a reticle. There are a number of them available depending on what and how you want to measure. The reticle is placed at the conjugate image plane of the eyepiece. It will be in focus with the sample. The only need for...