Power supply woes

#1 Post by 75RR » Thu May 07, 2020 12:00 pm

Don't know why this is suddenly happening ... anyone know? Any solutions?

Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Power supply woes

#2 Post by MichaelG. » Thu May 07, 2020 12:35 pm

Does that PSU have a Variac [or similar] under the knob ?
... If so, the wiper brush may be worn, or the track may dirty.

MichaelG.

https://uk.rs-online.com/web/c/power-su ... s/variacs/
Too many 'projects'

Hobbyst46
Posts: 4287
Joined: Mon Aug 21, 2017 9:02 pm

Re: Power supply woes

#3 Post by Hobbyst46 » Thu May 07, 2020 1:25 pm

Sorry for not being able to provide specific info.

Ancient capacitors often deteriorate with age. Oldish cloth-style wire insulation could fail as well.
My Carl Zeiss power supply (for the 6V 15W tungsten lamp) was repaired for free by an electronics technician, but after a while failed again.

Modern good-quality variable power supplies, although not cheap, are much more reliable.
Example (regretfully beyond my budget) is Delta Elektronika model ES300 (0-30VDC, 0-10A, 300W max) or ES015-10 (0-15VDC, 0-10A, 150W max). Lightweight, quiet, stable and reliable bench instrument (long-term user experience, no personal affiliation of any kind).

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#4 Post by 75RR » Thu May 07, 2020 2:05 pm

Thanks MichaelG. and Hobbyst46

Well, wiped the track and the brush with Isopropyl alcohol (does it need a firmer hand?) - plugged back in into another wall socket - same result.

Could a brown out be causing this?
Attachments
570.jpg
570.jpg (43.11 KiB) Viewed 5964 times
Trafo.jpg
Trafo.jpg (116.41 KiB) Viewed 5964 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

BramHuntingNematodes
Posts: 1546
Joined: Tue Jan 21, 2020 1:29 am
Location: Georgia, USA

Re: Power supply woes

#5 Post by BramHuntingNematodes » Thu May 07, 2020 2:07 pm

try some 320 grit sandpaper then isopropyl

edit on the other hand it looks really clean. What's the brush look like?
Last edited by BramHuntingNematodes on Thu May 07, 2020 2:12 pm, edited 1 time in total.
1942 Bausch and Lomb Series T Dynoptic, Custom Illumination

Charles
Posts: 1424
Joined: Mon Mar 02, 2015 11:55 pm

Re: Power supply woes

#6 Post by Charles » Thu May 07, 2020 2:10 pm

From the video it appears your regulator is loosing contact and then re-establishing. It could be the roller or wiring? You can open it up by first removing the power knob and then the four screws on the side. Unplug it first, of course. Then you will see a copper platform with an arm with roller which rolls across the copper platform as you increase/decrease the power knob. Insure the roller contacts the copper platform properly and you can also clean the path on the copper platform.

Charles
Posts: 1424
Joined: Mon Mar 02, 2015 11:55 pm

Re: Power supply woes

#7 Post by Charles » Thu May 07, 2020 2:13 pm

Okay, I was late posting. I see you got it open and tried cleaning. I use an eraser along the track and then wipe with alcohol. Also make sure all the wiring is secure.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#8 Post by 75RR » Thu May 07, 2020 2:33 pm

Thanks for all the replies

Will recheck all connections are secure.

I imagine ideally mark on track should be removed, will try with an eraser, failing that I dug out some 600 sandpaper (also happen to have 320) and will try with that.

Apply to brush as well?
Attachments
Track and brush.jpg
Track and brush.jpg (179.95 KiB) Viewed 5956 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Charles
Posts: 1424
Joined: Mon Mar 02, 2015 11:55 pm

Re: Power supply woes

#9 Post by Charles » Thu May 07, 2020 2:40 pm

Yes, clean the roller/brush as well.

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Power supply woes

#10 Post by MicroBob » Thu May 07, 2020 3:24 pm

Hi Glen,
do I see a broken solder joint here?

Bob
Attachments
Solder joint.jpg
Solder joint.jpg (7.48 KiB) Viewed 5947 times

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#11 Post by 75RR » Thu May 07, 2020 4:08 pm

MicroBob wrote:
Thu May 07, 2020 3:24 pm
Hi Glen,
do I see a broken solder joint here?

Bob
Well spotted! Yes I did. Turned out it was cracked but still connected. Soldered it anyway. Does not look as neat now :( I am going to have to learn to solder!

Double checked all the connections - all solid.

Tried the eraser trick to clean the track - worked very well. Still no luck.

Then I had a light bulb moment ... literally! Bulb was on its last legs.

Power supply now (again) solid as a rock!!!

Many thanks to all, especially Charles - who was the previous (very careful owner) :)


Feel sad about the demise of the bulb, had had it for several years - being always careful to keep below the rated voltage.

That was until I started trying to obtain COL + Pol with the 63x

Will use a LED for that next time!
.
.
.
.
Attachments
bulb.jpg
bulb.jpg (55.77 KiB) Viewed 5939 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Wes
Posts: 1027
Joined: Sat Mar 09, 2019 12:58 pm

Re: Power supply woes

#12 Post by Wes » Thu May 07, 2020 5:12 pm

75RR wrote:
Thu May 07, 2020 4:08 pm
MicroBob wrote:
Thu May 07, 2020 3:24 pm
Hi Glen,
do I see a broken solder joint here?

Bob
Well spotted! Yes I did. Turned out it was cracked but still connected. Soldered it anyway. Does not look as neat now :( I am going to have to learn to solder!

Double checked all the connections - all solid.

Tried the eraser trick to clean the track - worked very well. Still no luck.

Then I had a light bulb moment ... literally! Bulb was on its last legs.

Power supply now (again) solid as a rock!!!

Many thanks to all, especially Charles - who was the previous (very careful owner) :)


Feel sad about the demise of the bulb, had had it for several years - being always careful to keep below the rated voltage.

That was until I started trying to obtain COL + Pol with the 63x

Will use a LED for that next time!
.
.
.
.
They still sell these bulbs, I bought a couple a few months ago.
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#13 Post by 75RR » Thu May 07, 2020 5:20 pm

They still sell these bulbs, I bought a couple a few months ago.
Can you remember what you paid? I had two spare, now I only have one!

Since we are on the subject of illumination, anyone have a broken Zeiss 100w halogen lamphouse?

I ask because I have a Zeiss 100w halogen lamphouse that is perfectly fine apart from the fact that it is missing the focusing lens. Shame not to be able to get it going.

This is what the lens looks like ...

.
Attachments
Mail Attachment.jpeg
Mail Attachment.jpeg (31.15 KiB) Viewed 5922 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Wes
Posts: 1027
Joined: Sat Mar 09, 2019 12:58 pm

Re: Power supply woes

#14 Post by Wes » Thu May 07, 2020 5:30 pm

https://www.proflamps.com/uk/en/Product ... 302843057/

The most I've paid for a bulb, ever.

Hopefully you can get the lens for the 100W lamp, its ideal for DIC at high magnifications.
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#15 Post by 75RR » Thu May 07, 2020 5:51 pm

I remember they used to be around $8, which I thought excessive even then!
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Power supply woes

#16 Post by MicroBob » Thu May 07, 2020 7:02 pm

The bulb filament can be a very attractive object for observation under (improvised) incident lighting - so no complete loss. :lol:

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Power supply woes

#17 Post by 75RR » Thu May 07, 2020 7:54 pm

Every cloud has a tungsten lining ;)
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Power supply woes

#18 Post by MichaelG. » Thu May 07, 2020 8:09 pm

75RR wrote:
Thu May 07, 2020 5:51 pm
I remember they used to be around $8, which I thought excessive even then!
I was amazed to see that bulb has an ordinary bayonet mount, not a tricky prefocus flange.

MichaelG.
Too many 'projects'

Post Reply