After long delays I had 2 more microscopes arrive this week!
The first is an Olympus CH2, this one I bought mostly for the trinocular head, the complete scope was cheaper than any trinocular head I could find on eBay! It was describe as "In good working condition" but it arrived in a less the working condition. The trinocular head was fine, if a bit dirty, but the focusing mechanism is completely jammed. I had a look and there are stripped plastic as well as brass gears and the fine focus shaft is bent. The objectives are also in a less than great condition. It is a good thing I only wanted the trinocular head! The scope was not well packed and might have picked up some of the damage during transit, but the gears have not been stripped recently.... It will become a spares unit, the seller did promise to ship me a new focus mechanism, will see.
The second scope, an Olympus BH-2, arrived well packed and in a nice clean condition. Unfortunately it also had focus issues. A plastic gear was broken on the fine focus side and the rack gear attached to the bit the stage fits on is also stripped! Fortunately I could turn the brass rack gear around as only the one side was in use. The plastic gear I replaced with one I printed and then machined a bit. It was obvious that someone had worked on the focus mechanism of the scope before, he did not even bother to tighten all the screws again! So much for "Perfect working condition"!
I am glad I bought my first scope new, if I received something like either of these as my first scope after a long wait I would have been seriously discouraged
The focus on the BH-2 is now working perfectly. The stage also had an issue with interaction between the 2 axis of movement, if you moved the one the other moved as well. A complete strip down showed old oil and grease, a clean, regrease and careful attention to adjusting tensions had the stage working like new again.
The BH-2 is quite nice, everything else seems like new, it came with a 5 piece quick change objective turret with 5 nice FL SPlan objectives including 2x 100's, one with an iris. The turret condenser (BF/DF and 3 x PL settings) is also nice.
While working on the BH-2 I noticed that the condenser is fitted using a dove-tail that looks a lot like the one on my Amscope T690C. A quick test proved that the condensers are perfectly interchangeable! I tried swapping them both ways and it works. The Amscope DF/dry and DF/oil I have actually give better results that the Olympus unit. The build quality on the Amscope condensers are also better, they seem to be machined from solid metal.
My Canon DSLR with 2x adapter also fitted right into the trinocular head and produced perfect images without any adjustment.
So now I have a nice BH-2 on the bench. I think I will mostly use is for phase work but time will tell. The optics seem really good.
The next delivery should be an Olympus BHMJL, I am looking forward to trying some reflected light work!
Rudi