The objective is the heart of the microscope. Everything else is basically there to help it do its job.
The eyepiece however is in a slightly different position, as it can be seen as an extension of the objective.
This is so because it not only magnifies the image produced by the objective but also incorporates the second stage of optical corrections; the first stage taking place in the objective.
This means that eyepieces from manufactures different to that of the objectives should not be used, as their use would degrade the final image.
As for Zeiss eyepieces:
"C-type eyepieces may be used in conjunction with simple types of objective.
Objectives of higher correction should, if possible, always be combined with Kpl eyepieces.
This applies above all to the Planachromats and Planapochromat." Zeiss Optical Systems
By simple they mean achromats (objectives without a designation are always achromats)
I would add Neofluars to the must use compensating i.e. Kpl eyepieces list.
Best then as zzffnn suggests ... go with 10x Kpl eyepieces as you will no doubt want to get some 'higher correction' objectives at some point.
Note: If you come across 12.5x Kpl eyepieces, they will do equally well.