Hello,
I have for microscope OMAX 40X-2500X LED Digital Trinocular Lab Compound Microscope with 5MP Camera Omax 3550S.
On the screen of my laptop I can see only small part of what I see on microscope.
Do you have any idea how i can see all of the vision from the microscope?
Thank you very much.
Screen Problem
Re: Screen Problem
This is quite expected, for an eyepiece camera, which I believe is what yours is.
Have you tried to consult with OMax ? they might suggest a solution.
Have you tried to consult with OMax ? they might suggest a solution.
-
- Posts: 2790
- Joined: Sat Mar 03, 2018 9:09 pm
Re: Screen Problem
It sounds like ypu need reducing optics. C mount or eyepiece cameras have small sensors, so without reducing optics they can only see a fraction of the FoV. Omax will probably have some available.