Screen Problem

Do you have any microscopy questions, which you are afraid to ask? This is your place.
Post Reply
Message
Author
aysenustunay
Posts: 1
Joined: Thu Sep 05, 2019 2:34 pm

Screen Problem

#1 Post by aysenustunay » Thu Sep 05, 2019 2:39 pm

Hello,
I have for microscope OMAX 40X-2500X LED Digital Trinocular Lab Compound Microscope with 5MP Camera Omax 3550S.
On the screen of my laptop I can see only small part of what I see on microscope.

Do you have any idea how i can see all of the vision from the microscope?
Thank you very much.

Hobbyst46
Posts: 4287
Joined: Mon Aug 21, 2017 9:02 pm

Re: Screen Problem

#2 Post by Hobbyst46 » Thu Sep 05, 2019 3:34 pm

This is quite expected, for an eyepiece camera, which I believe is what yours is.
Have you tried to consult with OMax ? they might suggest a solution.

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: Screen Problem

#3 Post by Scarodactyl » Thu Sep 05, 2019 4:56 pm

It sounds like ypu need reducing optics. C mount or eyepiece cameras have small sensors, so without reducing optics they can only see a fraction of the FoV. Omax will probably have some available.

Post Reply