#1 Post by Sauerkraut » Tue Nov 19, 2019 1:28 am

My Optiphot has a phase contrast condenser with a dark field stop. The regular plan objectives give decent dark field through 20x with that single DF setting. Today I tried the setup with a 20x plan apo objective and discovered that it would not give DF. Instead it gave oblique illumination that looked more like DIC. Around the edges the image looked somewhat more like DF. Moving the condenser up and down did not remedy the situation, nor did cranking up the light intensity.

I assume plan apo objectives can give decent DF, so maybe the plan apo objective needs a different size DF stop than supplied in the phase condenser?

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Plan apo and dark field

#2 Post by 75RR » Tue Nov 19, 2019 5:07 am

I assume plan apo objectives can give decent DF, so maybe the plan apo objective needs a different size DF stop than supplied in the phase condenser?
One needs to match the darkfield stop to the NA of the objective rather than the magnification or the type. Larger stop should do it.

You may need to DIY or obtain a dedicated darkfield condenser
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Chris Dee
Posts: 216
Joined: Mon Aug 12, 2019 1:02 pm

Re: Plan apo and dark field

#3 Post by Chris Dee » Tue Nov 19, 2019 8:39 am

I also encountered this with my 25x 0.65 NA objective. As 75RR states you'll need a larger custom patch. On my microscope I noticed a difference of just 1mm in radius had a large effect, and the margin of error for good contrast and dark background was narrow. Careful condenser height adjustment was needed to get even illumination, and can be a fiddly to get right.

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Plan apo and dark field

#4 Post by Hobbyst46 » Tue Nov 19, 2019 8:59 am

I would suggest you try oiling the condenser to the slide (although the objective is used dry) and see if this improves the DF for NA's of 0.4-0.7.

Sauerkraut
Posts: 211
Joined: Tue Apr 16, 2019 2:07 am
Location: Oregon, USA

Re: Plan apo and dark field

#5 Post by Sauerkraut » Tue Nov 19, 2019 4:21 pm

Thanks, all. I'll try a larger stop, probably in a dedicated DF condenser.

Oil would be a good idea too, but due to the clean up required, I've shied away from using it much.