Just seems like there's not enough light.

#1 Post by Greg Howald » Sat Nov 07, 2020 8:59 pm

My scope uses a 20w halogen bulb and it will not run a 30 watt bulb. Just doesn't seem to be enough light.
I would consider switching over to 5 watt led, but I don't have any idea where to get parts for that. I can get the bulb and socket without difficulty but where does one find a power supply?
Any suggestions?
Greg

DonSchaeffer
Posts: 3349
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Just seems like there's not enough light.

#2 Post by DonSchaeffer » Sat Nov 07, 2020 9:56 pm

Just get a flashlight.

hans
Posts: 1006
Joined: Thu May 28, 2020 11:10 pm
Location: Southern California

Re: Just seems like there's not enough light.

#3 Post by hans » Sat Nov 07, 2020 10:26 pm

I think some details on what type of illumination system the microscope has and how you are configuring it will be necessary if you want to troubleshoot.

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Just seems like there's not enough light.

#4 Post by Greg Howald » Sat Nov 07, 2020 10:50 pm

Seems more than likely that this problem may have to do with the lenses attached to my face.😃
Greg

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Just seems like there's not enough light.

#5 Post by Greg Howald » Sun Nov 08, 2020 3:57 am

Just found led replacement bulbs that will work in my scope. All I have to do is change the bulb for test purposes. Then I'll have an answer.
Greg

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Just seems like there's not enough light.

#6 Post by 75RR » Sun Nov 08, 2020 5:28 am

hans wrote:
Sat Nov 07, 2020 10:26 pm
I think some details on what type of illumination system the microscope has and how you are configuring it will be necessary if you want to troubleshoot.
Agree, but only because I have misplaced my crystal ball... :)
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

BramHuntingNematodes
Posts: 1546
Joined: Tue Jan 21, 2020 1:29 am
Location: Georgia, USA

Re: Just seems like there's not enough light.

#7 Post by BramHuntingNematodes » Sun Nov 08, 2020 6:18 am

Led bulbs have a lot of different configurations and some of them are extremely suboptimal for microscopy. Frequently light is directed where it's not useful or the light source is larger and more uneven than a filament.

The plus side is that some bulbs don't need an additional power supply. You just plug them straight into the wall.
1942 Bausch and Lomb Series T Dynoptic, Custom Illumination

hans
Posts: 1006
Joined: Thu May 28, 2020 11:10 pm
Location: Southern California

Re: Just seems like there's not enough light.

#8 Post by hans » Sun Nov 08, 2020 6:24 am

To give a point of reference my Reichert Microstar IV (which a "modified Koehler" system with 20 W halogen bulb) is blinding, painful to look at, in brightfield 4-40X and bright but tolerable at 100X if the room is brightly lit. The switchable ND filter must be used most of the time in brightfield to reduce the power.

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Just seems like there's not enough light.

#9 Post by Greg Howald » Sun Nov 08, 2020 1:43 pm

The scope is Amscope B660. 20w halogen, kholer, 4 to 100x, plain polarization in head, 1.25 condenser fully open. It has its difficulties. Condenser must be lowered at 4x to avoid interference from the Condenser on the outer portion of field of view. At 40x it dulls out. 100x is a little brighter. At no time is it as bright as my 20 watt halogen petrographic scope. The bulb has just been replaced with a new one with no improvement. I know led bulbs may not be the answer but I think it worth a try.
Greg

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Just seems like there's not enough light.

#10 Post by 75RR » Sun Nov 08, 2020 2:00 pm

.
Have a look through this interactive tutorial, it could be that your Köhler setup needs a little fine tuning.

In this tutorial you need to adjust the lamp filament first, something you would not probably have to do on your microscope.

http://zeiss-campus.magnet.fsu.edu/tuto ... flash.html
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Just seems like there's not enough light.

#11 Post by Greg Howald » Sun Nov 08, 2020 2:39 pm

75RR, I greatly appreciate this knowledge. Thank you.
Greg

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Just seems like there's not enough light.

#12 Post by Greg Howald » Sun Nov 08, 2020 3:54 pm

Studied tutorial. Centered lamp. Condenser 4mm below stage. Great improvement!
I appreciate you folks helping me out. You might have saved a $500 microscope from the dump.
😃😃😃 Greg