Oblique darkfield illumination

Here you can discuss different microscopic techniques and illumination methods, such as Brightfield, Darkfield, Phase Contrast, DIC, Oblique illumination, etc.
Post Reply
Message
Author
User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Oblique darkfield illumination

#1 Post by gekko » Fri Jul 01, 2016 6:19 pm

A while ago, Rod Nabbolz ("rnabholz") posted a darkfield image that I erroneously thought was illuminated off-center. That gave me the idea of trying to use an off-center stop to obtain oblique darkfield illumination. More recently, John B ("mrsonchus") posted what he called "Dark-field of a sort" ( viewtopic.php?f=6&t=3300 ) that I think, from his description and from the appearance of the image, corresponds to what I've called oblique darkfield illumination. Below are image pairs taken from student-grade commercial slides comparing darkfield images with those taken using oblique darkfield illumination. The oblique darkfield images were obtained by de-centering the darkfield stop which was large enough to still block the direct illumination, as represented in the diagram below:
                  Image
where the green circle represents the darkfield stop**, and the condenser iris** is shown fully open.
______________________________
** as seen at the back focal plane of the objective.

Another way may be to use a centered darkfield stop, but block the illumination from one side (or, using a darkfield condenser, block the illumination reaching the condenser from one side). I've not tried this yet, but I think that should work as well (except perhaps would give less light).

Descriptions were copied from the slide label. The left image was taken using regular oblique illumination, and the right oblique darkfield. All images were taken with a 10x objective, and the DF position of the Nikon Phase Contrast condenser was used as a darkfield stop (and shifted slightly to produce oblique darkfield).

__________________________ Darkfield ____________________ ___________________ Oblique Darkfield _________________

"Silver berry scaly hair"
Image

"Angora rabbit hair"
Image

"Frog blood"
Image

"Nymphaea of aqustio stem"
Image

"Diatom"
Image
Last edited by gekko on Fri Jul 01, 2016 11:37 pm, edited 1 time in total.

User avatar
vasselle
Posts: 2763
Joined: Sat Oct 25, 2014 5:32 pm
Location: France

Re: Oblique darkfield illumination

#2 Post by vasselle » Fri Jul 01, 2016 6:41 pm

Bonjour gekko
Des fois pour certain sujets je fait comme vous, fond noir décentré et je trouve que c'st plus spectaculaire et très facile à mètre en oeuvre.
Car ça donne du bon relief à image.
Voici une photo avec la même technique que vous utiliser.
1.jpg
1.jpg (349.43 KiB) Viewed 10119 times
Cordialement seb
Microscope Leitz Laborlux k
Boitier EOS 1200D + EOS 1100D

User avatar
zzffnn
Posts: 3200
Joined: Sat Jun 20, 2015 3:57 am
Location: Houston, Texas, USA
Contact:

Re: Oblique darkfield illumination

#3 Post by zzffnn » Fri Jul 01, 2016 6:54 pm

Great work, gekko.

I think you meant the left is regular darkfield.

I am guessing for "oblique darkfield" to work well, subject would have to be thick, translucent and have many planes of focus depth. And probably objective NA at 0.4 or less.

From your photos, I personally prefer oblique darkfield for #1 (due to enhanced depth) and regular darkfield for #2, #4 and #5 (due to more even lighting). My eyes are not good enough to see differrnce in #3.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Oblique darkfield illumination

#4 Post by 75RR » Fri Jul 01, 2016 7:11 pm

Works very well. Thanks for posting this.

vasselle: your example is excellent.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
rnabholz
Posts: 3086
Joined: Thu Jan 08, 2015 10:11 pm
Location: Iowa USA
Contact:

Re: Oblique darkfield illumination

#5 Post by rnabholz » Fri Jul 01, 2016 7:31 pm

Very nice technique, very well explained and demonstrated.

The effect on the Diatom is quite pronounced and interesting.

I guess I will be sticking more things into my condenser again...

;)

Rod

JimT
Posts: 3247
Joined: Fri Oct 24, 2014 1:57 pm

Re: Oblique darkfield illumination

#6 Post by JimT » Fri Jul 01, 2016 8:03 pm

Thanks. Something else to experiment with. Good explanation and examples.

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#7 Post by gekko » Fri Jul 01, 2016 11:02 pm

seb, thank you very much for contributing your comment (which I totally agree with) and, expecially, your superb example (much better than my attempts).

zzffnn, thank you! Yes, I meant the left to be "regular" (concentric) darkfield. I think you are right: I think the same criteria that apply to objects useful for "regular" oblique illumination also apply to oblique darkfield. I may be wrong, but I don't think special NA limitations apply; I think whenever one can get concentric darkfield, one can get oblique darkfield (either by extending the darkfield stop in one direction or by blocking the light coming from one direction. I would be interested if you would like to try covering about half the bottom of your darkfield condenser and confirm this or prove my statement wrong. I'm not 100% sure it would work as I've not tried it, but I can't think of why it wouldn't. It would reduce the light by about one half, but apart from that I think it should work.

75RR, many thanks!

Rod, thank you very much. I chose to use a diatom as an example on purpose :) . I look forward to see how well this would work with your diatoms and darkfield condenser.

JimT, many thanks: yes I thought you would like to experiment with stuff like that :).

Thanks to all.

User avatar
zzffnn
Posts: 3200
Joined: Sat Jun 20, 2015 3:57 am
Location: Houston, Texas, USA
Contact:

Re: Oblique darkfield illumination

#8 Post by zzffnn » Sat Jul 02, 2016 12:41 am

gekko wrote: zzffnn, thank you! Yes, I meant the left to be "regular" (concentric) darkfield. I think you are right: I think the same criteria that apply to objects useful for "regular" oblique illumination also apply to oblique darkfield. I may be wrong, but I don't think special NA limitations apply; I think whenever one can get concentric darkfield, one can get oblique darkfield (either by extending the darkfield stop in one direction or by blocking the light coming from one direction. I would be interested if you would like to try covering about half the bottom of your darkfield condenser and confirm this or prove my statement wrong. I'm not 100% sure it would work as I've not tried it, but I can't think of why it wouldn't. It would reduce the light by about one half, but apart from that I think it should work.
gekko,

I did try covering up half of my darkfield condensers, three times, with your B&L paraboloid, an AO 214F and a LOMO OI-13 cardioid. None worked for most high magnification subjects that I tried. Part of the subjects just went dark, even though my 4000 lumen LED was used at max. For very low magnification, it may work, since light is ample.

I guess in practice, high NA high mag darkfield needs lots of light, very precise alignment and optical correction, which may not go well with the way oblique darkfield works. I would be interested to see if anyone can get good oblique darkfield at 40x NA 0.65 or above. I couldn't for sure.

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#9 Post by gekko » Sat Jul 02, 2016 12:59 am

zzffnn, thank you so much for taking the time and trouble to do the test so quickly. So that seems to rule out high magnification oblique darkfield. I have a 40x/.70 objective with which I can barely get darkfield, so I will test it and see what will happen. I may not be able to get to try it until Sunday, but I'll report what I find. Thank you again.

User avatar
zzffnn
Posts: 3200
Joined: Sat Jun 20, 2015 3:57 am
Location: Houston, Texas, USA
Contact:

Re: Oblique darkfield illumination

#10 Post by zzffnn » Sat Jul 02, 2016 1:06 am

I did those experiments a few months ago. Phil holds the opinion that "oblique darkfield" generally does not work well. I did not believe him first, but then after doing my own experiments, I think it is indeed the case at or above 40x NA 0.65.

At NA 0.4 or lower, the darkfield illumination cone lies far outside of objective edge NA, so one would have some room for error/playing around.

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#11 Post by gekko » Mon Jul 04, 2016 12:05 pm

I just found out that there there is an article by Thomas L Webster published in Micscape Magazine in 2005 entitled "Oblique Darkfield Illumination",
http://www.microscopy-uk.org.uk/mag/ind ... lique.html
with the same idea. Instead of a darkfield stop, Webster uses a condenser phase annulus but the idea is the same. I was not aware of that paper when I posted this thread, and I apologize for not citing his work at the time.
Last edited by gekko on Mon Jul 04, 2016 12:34 pm, edited 1 time in total.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Oblique darkfield illumination

#12 Post by 75RR » Mon Jul 04, 2016 12:19 pm

I was not aware of that paper when I posted this thread, and I apologize for not citing his work at the time.
You should be a little easier on gekko. Bit unfair to expect him to be all knowing! ;)

i.e. Don't be so hard on yourself.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#13 Post by gekko » Mon Jul 04, 2016 12:54 pm

Thank you, 75RR, but I was only giving credit where credit is due.

billbillt
Posts: 2895
Joined: Mon Nov 10, 2014 10:01 pm

Re: Oblique darkfield illumination

#14 Post by billbillt » Mon Jul 04, 2016 3:37 pm

Hi Gekko,

I just stumbled upon this.. Your photos are superb!!... Some of the best darkfield illumination I have ever seen...

BillT

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#15 Post by gekko » Mon Jul 04, 2016 4:24 pm

Thank you, BillT for your much-too-kind comment. I think what I've done was to use the easiest route to DF, low magnification and prepared slides. The really challenging DF is what Rod and others do, with 100x objectives!
Despite zzffnn's warning, I was trying this morning to get some darkfield and oblique darkfield images with my 40x/0.8-1.3 oil-immersion objective just as a challenge (I don't even dare try the 100x), but I haven't looked at the pictures yet to see if I have succeeded or not (hard to tell by looking through the eyepiece) I have acquired new respect for the difficulties involved in darkfield imaging with a darkfield condenser where you are working literally in the dark and see nothing until everything is adjusted (so easy to poke the slide with the objective or condenser). And that is not even with a 100x objective...

billbillt
Posts: 2895
Joined: Mon Nov 10, 2014 10:01 pm

Re: Oblique darkfield illumination

#16 Post by billbillt » Mon Jul 04, 2016 5:10 pm

Well, I think you did a great job!...
BillT

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#17 Post by gekko » Mon Jul 04, 2016 6:45 pm

Thank you, BillT. I will post my attempts at DF and Oblique DF using my 40x objective and hope to get some critical comments (positive or negative).

Rodney
Posts: 748
Joined: Thu Feb 19, 2015 2:52 am
Location: Southern Georgia, USA

Re: Oblique darkfield illumination

#18 Post by Rodney » Thu Jul 07, 2016 3:37 am

Those really stood out Gekko, nice work, trying to get my eyes adjusted now.

Rodney

User avatar
gekko
Posts: 4701
Joined: Sun Oct 12, 2014 7:38 am
Location: Durham, NC, USA.

Re: Oblique darkfield illumination

#19 Post by gekko » Thu Jul 07, 2016 11:26 am

Rodney wrote:Those really stood out Gekko, nice work, trying to get my eyes adjusted now.
Rodney
Thanks Rodney. And sorry if the images got your eyes out of adjustment :) .

Post Reply