Leitz/Leica Labovert DIC/ICT

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Post Reply
Message
Author
Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Leitz/Leica Labovert DIC/ICT

#1 Post by Rorschach » Fri Apr 19, 2024 12:12 pm

Hello,

Does anyone know what exactly look like the DIC/ICT system parts for a Labovert? I am considering a Leitz/Leica (it has both labels on it, so early 90ies I guess) Labovert that seems to have two of the necessary sliders (that go in below the nosepiece), a polarizer slide of some kind that slots under the trinocular (I suppose) and the correct looking universal condenser which seems to have at least one prism. Please see the photos:
Labovert6.jpg
Labovert6.jpg (56.51 KiB) Viewed 940 times
Labovert7.jpg
Labovert7.jpg (72.42 KiB) Viewed 940 times
Labovert8.jpg
Labovert8.jpg (81.24 KiB) Viewed 940 times
Labovert9.jpg
Labovert9.jpg (95.56 KiB) Viewed 940 times
Labovert10.jpg
Labovert10.jpg (101.51 KiB) Viewed 940 times

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#2 Post by Rorschach » Fri Apr 19, 2024 12:13 pm

More pics here:
Labovert1.jpg
Labovert1.jpg (60.97 KiB) Viewed 939 times
Labovert2.jpg
Labovert2.jpg (46.02 KiB) Viewed 939 times
Labovert3.jpg
Labovert3.jpg (57.62 KiB) Viewed 939 times
Labovert4.jpg
Labovert4.jpg (137.13 KiB) Viewed 939 times
Labovert5.jpg
Labovert5.jpg (65.35 KiB) Viewed 939 times

wabutter
Posts: 202
Joined: Thu May 09, 2019 12:27 am

Re: Leitz/Leica Labovert DIC/ICT

#3 Post by wabutter » Fri Apr 19, 2024 11:26 pm

It appears you have the required componnents for DIC The Polarizer, Analyser and the objective prism for the 25X L on the slider. It appears the condenser has a matching prism for the 25x (S25) as well? If so, it is complete for DIC on that one objective.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#4 Post by Rorschach » Sat Apr 20, 2024 5:21 am

wabutter wrote:
Fri Apr 19, 2024 11:26 pm
It appears you have the required componnents for DIC The Polarizer, Analyser and the objective prism for the 25X L on the slider. It appears the condenser has a matching prism for the 25x (S25) as well? If so, it is complete for DIC on that one objective.
Thanks Wayne, helpful as ever :-) Yes, the condenser does seem to have the 25x prism. Strangely though, no prism for the 40x. The DIC for these was marketed as having objectives and prisms for 25x and 40x. Please see the attached pic on Labovert brochure text on ICT and polarization.
Labovert_brochure_text_on_ICT_and_polarization.jpg
Labovert_brochure_text_on_ICT_and_polarization.jpg (95.81 KiB) Viewed 873 times
I did see a photo of a Labovert DIC scope which had a rotable slider in the slot above the condenser. That one this scope does not seem to have but I am not sure what capability is lost in this case.

Also, in a Labovert brochure there was this strange condenser part that I don't recognise, please see the attached photo. Perhaps it is related to something else, not DIC, though.
Labovert_condenser_part.jpg
Labovert_condenser_part.jpg (42.78 KiB) Viewed 875 times
Despite furious searching, I have been unable to find the user manual for the Leitz or Leica Labovert. Only a brochure could be found :cry:

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#5 Post by Rorschach » Sat Apr 20, 2024 7:24 am

The seller has then provided many photos of what looks like maybe two different prisms in the condenser turret. However, I can't verify what exactly the other prism(s) are, the seller isn't 'in the know', and, outside the red 'S 25', there are no other stickers/markings on the outer condenser turret ring.

owl_capone
Posts: 2
Joined: Thu Dec 22, 2022 7:39 pm

Re: Leitz/Leica Labovert DIC/ICT

#6 Post by owl_capone » Sat Apr 20, 2024 1:15 pm

I have a Fluovert with DIC/ICT working on the L25 Fluotar and L40 Fluotar. Afaik in this version the condensor only needs the single S25 (Schnittweite 25mm - 25mm working distance) prism and then the objective-side prism sliders for each respective objective. In practice you can also just use either the L25 or the L40 objective-side prism for both, at least through the eyepieces there's no big difference - I guess the low aperture of the LWD objectives plays a role here.
Also Phaco/non-Phaco for the objectives doesn't matter, it works with both.
Unfortunately so far it also only works with thise two as they have comparable positions of the back focal plane matching the prism. At least I haven't found other objectives in my stash of Leitz lenses that give a DIC effect with that setup.
Once I'm back home on sunday I can take some pictures of my setup.
In short you need: a 'fat' polarizer slider for the illumination slot at the top, condnsor with the required prism(s) (one for 25mm WD, more for other versions afaik), 'certified' objective (Fluotar L40 and Fluotar L25), objective-side prism (L40 or L25) and the slim ICT /polarizer slider for the nosepiece filter slot.
That last one can be flipped for normal DIC or color DIC with a sandwiched lambda plate.

A long time ago a Labovert with DIC was sold in the german forum that, in addition to the L25 and L40 Fluotars had a 10x Fluotar with the build-in wollaston prisms that Leitz used for their ICT system up until the Aristoplan. From then on the prisms moved into a seperate revolver in the nosepiece. Not that dissimilar to how it was done for the Labovert...
That 'scope came from a seller that didn't just plug together a frankenscope and it was stated to work out optcially.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#7 Post by Rorschach » Sat Apr 20, 2024 1:57 pm

owl_capone wrote:
Sat Apr 20, 2024 1:15 pm
I have a Fluovert with DIC/ICT working on the L25 Fluotar and L40 Fluotar. Afaik in this version the condensor only needs the single S25 (Schnittweite 25mm - 25mm working distance) prism and then the objective-side prism sliders for each respective objective. In practice you can also just use either the L25 or the L40 objective-side prism for both, at least through the eyepieces there's no big difference - I guess the low aperture of the LWD objectives plays a role here.
Also Phaco/non-Phaco for the objectives doesn't matter, it works with both.
Unfortunately so far it also only works with thise two as they have comparable positions of the back focal plane matching the prism. At least I haven't found other objectives in my stash of Leitz lenses that give a DIC effect with that setup.
Once I'm back home on sunday I can take some pictures of my setup.
In short you need: a 'fat' polarizer slider for the illumination slot at the top, condnsor with the required prism(s) (one for 25mm WD, more for other versions afaik), 'certified' objective (Fluotar L40 and Fluotar L25), objective-side prism (L40 or L25) and the slim ICT /polarizer slider for the nosepiece filter slot.
That last one can be flipped for normal DIC or color DIC with a sandwiched lambda plate.

A long time ago a Labovert with DIC was sold in the german forum that, in addition to the L25 and L40 Fluotars had a 10x Fluotar with the build-in wollaston prisms that Leitz used for their ICT system up until the Aristoplan. From then on the prisms moved into a seperate revolver in the nosepiece. Not that dissimilar to how it was done for the Labovert...
That 'scope came from a seller that didn't just plug together a frankenscope and it was stated to work out optcially.

Hello 'Owl' :-)

Thanks a lot! This info is very reassuring! Perhaps I can get by without a L40 objective prism, at least for starters. And maybe not that much need to worry about whether the condenser turret has the 40x prism or not But does your info also mean, that I am missing one of the sliders?

owl_capone
Posts: 2
Joined: Thu Dec 22, 2022 7:39 pm

Re: Leitz/Leica Labovert DIC/ICT

#8 Post by owl_capone » Sat Apr 20, 2024 9:29 pm

Nope, maybe my phrasing was a bit too convoluted :D
The thing in your third picture is the polarizer for the illumination side. In the Labovert the upper arm carries the illumination pathway as well as the image going into the head.

You can see some of the internal layout in the patent https://patentimages.storage.googleapis ... 619503.pdf.
Nice and complicated light routing that as rumours have it was very time consuming to adjust at the factory. Unfortunately for Leitz the Labovert seemed to be very popular for ergonomic reasons - the sample is right under your nose same as with a traditional upright. So the model stuck around for quite some time hence why you see them both with Leitz and Leica branding.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#9 Post by Rorschach » Sun Apr 21, 2024 5:04 am

owl_capone wrote:
Sat Apr 20, 2024 9:29 pm
Nope, maybe my phrasing was a bit too convoluted :D
The thing in your third picture is the polarizer for the illumination side. In the Labovert the upper arm carries the illumination pathway as well as the image going into the head.

You can see some of the internal layout in the patent https://patentimages.storage.googleapis ... 619503.pdf.
Nice and complicated light routing that as rumours have it was very time consuming to adjust at the factory. Unfortunately for Leitz the Labovert seemed to be very popular for ergonomic reasons - the sample is right under your nose same as with a traditional upright. So the model stuck around for quite some time hence why you see them both with Leitz and Leica branding.
Ok, excellent! It seems the scope is set to go then! I've seen photos of Laboverts with an adjustable polarizer in the upper slot but maybe that doesn't have anything to do with DIC.

When you have the time, it'd be great to see pics of your setup and the details of it :)

apochronaut
Posts: 6394
Joined: Fri May 15, 2015 12:15 am

Re: Leitz/Leica Labovert DIC/ICT

#10 Post by apochronaut » Sun Apr 21, 2024 7:41 am

Rorschach wrote:
Sat Apr 20, 2024 5:21 am
wabutter wrote:
Fri Apr 19, 2024 11:26 pm
It appears you have the required componnents for DIC The Polarizer, Analyser and the objective prism for the 25X L on the slider. It appears the condenser has a matching prism for the 25x (S25) as well? If so, it is complete for DIC on that one objective.

Also, in a Labovert brochure there was this strange condenser part that I don't recognise, please see the attached photo. Perhaps it is related to something else, not DIC, though.

Labovert_condenser_part.jpg

Despite furious searching, I have been unable to find the user manual for the Leitz or Leica Labovert. Only a brochure could be found :cry:
I think, a supplemental condenser lens for very low power wide field objectives?

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#11 Post by Rorschach » Sun Apr 21, 2024 7:48 am

apochronaut wrote:
Sun Apr 21, 2024 7:41 am
Rorschach wrote:
Sat Apr 20, 2024 5:21 am
wabutter wrote:
Fri Apr 19, 2024 11:26 pm
It appears you have the required componnents for DIC The Polarizer, Analyser and the objective prism for the 25X L on the slider. It appears the condenser has a matching prism for the 25x (S25) as well? If so, it is complete for DIC on that one objective.

Also, in a Labovert brochure there was this strange condenser part that I don't recognise, please see the attached photo. Perhaps it is related to something else, not DIC, though.

Labovert_condenser_part.jpg

Despite furious searching, I have been unable to find the user manual for the Leitz or Leica Labovert. Only a brochure could be found :cry:
I think, a supplemental condenser lens for very low power wide field objectives?
Yeah, probably. This scope doesn't have it, though. Yet there is a 4x objective. Finding the supplemental lens on eevilbay or elsewhere may prove near impossible.

apochronaut
Posts: 6394
Joined: Fri May 15, 2015 12:15 am

Re: Leitz/Leica Labovert DIC/ICT

#12 Post by apochronaut » Sun Apr 21, 2024 12:57 pm

Supplemental condenser lenses lead double lives and end up in strange places. My best acquisition of a needed one was from ebay and came complete with it's 1.4 achromat aplanat condenser. I picked it up on a Sunday morning and had to wait a while because the seller who looked like Keith Moon, especially in haircut, was just returning from a gig. He was a drummer in a Who tribute band but moonlighted (literally) as a member of a dumpster diver co-op, which was the source for the condenser. They had a brick and mortar storefront too!

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#13 Post by Rorschach » Sun Apr 21, 2024 1:11 pm

apochronaut wrote:
Sun Apr 21, 2024 12:57 pm
Supplemental condenser lenses lead double lives and end up in strange places. My best acquisition of a needed one was from ebay and came complete with it's 1.4 achromat aplanat condenser. I picked it up on a Sunday morning and had to wait a while because the seller who looked like Keith Moon, especially in haircut, was just returning from a gig. He was a drummer in a Who tribute band but moonlighted (literally) as a member of a dumpster diver co-op, which was the source for the condenser. They had a brick and mortar storefront too!
What a story! No way I'll be able to one-up that :D

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#14 Post by Rorschach » Tue Apr 23, 2024 8:13 pm

The Labovert comes with a Leitz 10x/0.25 Plan star objective, which is DIC compatible according to what I have read. So, perhaps also this one will work with the one 25x DIC prism. This is of course just an extrapolation from owl capones experiences.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#15 Post by Rorschach » Thu May 23, 2024 5:50 pm

I now have the Labovert in hand. Every mechanical control operates very smoothly, including XY and (very importantly!) both coarse and fine focus. These scopes have a focus mechanism that is notoriously difficult to fix if stuck. This is the FS version meaning fixed stage, so the objective nosepiece is what moves on z axis when focusing. This was implemented to facilitate the use of micromanipulators.

However, when looking at the universal condenser, I am not sure what there is in the slots. Definitely there are no phase discs or darkfield spots. But I am not sure if there are any DIC prisms on any of them either. I may have to open the condenser turret up to check. There seems to be three crews on the top plate so I assume that's how it is opened. Is there anything to be concerned about regarding the operation?

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#16 Post by Rorschach » Fri May 24, 2024 4:52 am

I learned yesterday that this condenser is a ULU condenser, not the regular UK condenser that usually came with a Labovert. It is because of this one being the FS (fixed stage) variation. Not all DIC parts and condenser parts are compatible, some are specific for the condenser type. Working distance for the ULU is 25mm whereas for the UK it is 15mm and the prisms are not interchangeable.

I was unable to get the silver turret out from the condenser, due to the central screw being an extremely annoying special one with very wide but narrow groove. None of drivers are useful here.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#17 Post by Rorschach » Fri May 24, 2024 10:53 am

Progress! Made a tool for that nasty screw and opened up the condenser, removed the turret. There indeed is the S 25 DIC prism in one slot but all the others are empty.
Labovert_S25_DIC_condenser_prism.jpg
Labovert_S25_DIC_condenser_prism.jpg (59.84 KiB) Viewed 238 times
Above the universal condenser in the stand, there is what appears one of the couple options for a simple, low resolution/aperture condenser. This probably needs to be removed when using the ULU universal condenser.
Labovert_simple_condenser.jpg
Labovert_simple_condenser.jpg (49.73 KiB) Viewed 238 times
Edit: correction, the "simple condenser with the description SK/UK/S25 appears not to be a condenser, instead it's an illumination lens to be used with the UK condenser (ULU also, I guess). This info was in the Leitz Fluovert user manual.

Rorschach
Posts: 397
Joined: Sun Mar 10, 2019 11:44 am

Re: Leitz/Leica Labovert DIC/ICT

#18 Post by Rorschach » Tue May 28, 2024 6:47 pm

An update on this. The 25x NPL Fluotar produces an extremely nice DIC effect. This isn't very surprising, I suppose, as that was written on the tin. However, in my very quick tests, I noticed that both the 10x Plan star lens and the 6.3x PL Fluotar star lens produced DIC as well. Probably not as good as it would be with dedicated prisms but I'll take it :D It will be very useful for observations and ID work even if not good enough for high-quality photos.

I was going to try also the 16x PL Fluotar star lens but it is too long to be able to turn (when turning the revolver). It hits the underside of the stage. Perhaps a different stage or stage insert with a larger central opening would allow it to be used? At some point I will test it nevertheless by simply screwing it in directly from above through the central opening, just to see if DIC can be had or not.

One small problem that I noticed is that the condenser vertical drive is slipping slowly. Not all the time but sometimes so I need to fix that at some point.

Post Reply