Reichert Polyvar

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Message
Author
wabutter
Posts: 201
Joined: Thu May 09, 2019 12:27 am

Re: Reichert Polyvar

#151 Post by wabutter » Thu May 02, 2024 5:55 pm

For those interested, that are a number of interation of the Polyvar. the link below is a summary photo list. Some are extremely rare, like the Infrapol. A Polarized light microscope that is only presenting near IR light to the sample. No need for a bino tube. Only a IR sensitive camera. Pretty cool that Silicon is transparent to IR light.
the Poly Spec was am attachable microspectrophotometer. work at the blazing speed of a Z80 process all the way up to 2mhz.

The Polyvar Pol. A TL/IL polarized light microscope for qualitative and quantative microscopy in petrography and forensic. By the way, all optics produced in Vienna were conditinoally strain free for use in Polarized light. the component that had a np disignation were strain free to the quality required in a petrographic scope.

linkn below
https://drive.google.com/file/d/1B2o-OZ ... sp=sharing

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#152 Post by Rorschach » Thu May 02, 2024 6:10 pm

Thanks Wayne! So that is when the Plan Fluor-Apo objectives came out, for the Polyvar 2. Interesting!

The InfraPol and the PolySpec sound intriguing!
Last edited by Rorschach on Thu May 02, 2024 6:42 pm, edited 1 time in total.

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#153 Post by Rorschach » Thu May 02, 2024 6:40 pm

wabutter wrote:
Thu May 02, 2024 4:44 pm
I rescanned the order guide with the missing pages 4-5.
Here is a link

https://drive.google.com/file/d/1HhM2Hy ... drive_link

This link will take you to the Polyvar 2 4 page brochure, it is the only version I have
https://drive.google.com/file/d/1hFPQUz ... drive_link

Thre was only one version of the top element for the Universal Condenser. It mention a conservative spec with the acrhromatic reference. I didn'f find a Apochromatic refernce in my lit. However, with 5 elements, I am confident it is a Aochromatic/Aplanatic condenser.What made is somewhat unique was it could be used dry or with oil. Also with the aux lens below the condenser mount, it could be used with a 4x objective

When the Diastar was introduced, that top element could also fit in the 1201 condenser mount. Very good option when using the Reichert produced Plan Fluor objectives.
There is one thing that confuses me. If there was just one version of the top element for the Universal Condenser, how come I have three different tops in each of my thee Universal Condensers?
A=1.3 np
A=1.3 P
A=0.9 P
Could it be then, that the latter two tops are from another Reichert condenser, another scope model?

Edit: Based on what you wrote couple messages above, I assume the 'np' in the first of the condenser tops means that it is of a grade sufficient for a petrographic scope, in terms of being strain-free? Does the 'P' then mean that those are only good enough for simple polarization?

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#154 Post by apochronaut » Thu May 02, 2024 9:11 pm

Rorschach, could you measure across the thread of one of your condenser barrels?.

wabutter
Posts: 201
Joined: Thu May 09, 2019 12:27 am

Re: Reichert Polyvar

#155 Post by wabutter » Thu May 02, 2024 11:35 pm

Rorschach wrote:
Thu May 02, 2024 6:40 pm


There is one thing that confuses me. If there was just one version of the top element for the Universal Condenser, how come I have three different tops in each of my thee Universal Condensers?
A=1.3 np
A=1.3 P
A=0.9 P
Could it be then, that the latter two tops are from another Reichert condenser, another scope model?

Edit: Based on what you wrote couple messages above, I assume the 'np' in the first of the condenser tops means that it is of a grade sufficient for a petrographic scope, in terms of being strain-free? Does the 'P' then mean that those are only good enough for simple polarization?
As the universal condenser was always supplied with DIC prism, the P was also strain free for use in simple POL or DIC. As I mentioned in an earlier post, all objectives and condenser were designed and built to a strain free standard, Those used on the PLM were selected as the best of the best.

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#156 Post by Rorschach » Sat May 04, 2024 9:16 am

apochronaut wrote:
Thu May 02, 2024 9:11 pm
Rorschach, could you measure across the thread of one of your condenser barrels?.
Sure: it is 33mm.

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#157 Post by apochronaut » Sat May 04, 2024 1:46 pm

Rorschach wrote:
Sat May 04, 2024 9:16 am
apochronaut wrote:
Thu May 02, 2024 9:11 pm
Rorschach, could you measure across the thread of one of your condenser barrels?.
Sure: it is 33mm.
Thanks.

Well, what a surprise. Not sure where that last sentence in post 150 came from.
A Reichert Austria condenser won't fit the AO/Reichert U.S.A. condenser housings, none of them and certainly not the standard one that the #1201 condenser would have been threaded into. Almost all of the AO fittings were SAE, not metric and the condenser thread was 1 1/4" X 32 t.p.i. or just over 31mm, since the late 60's. If the Reichert Austria 1.30 condenser did fit the 1201 condenser housing that would have meant that they were made in Buffalo.

The Diastar did use a 100X 1.30 Plan Fluor objective and therefore needed a higher order condenser than the standard cat.# 1970 1.25 oil aspheric or cat.# 1201 .90 achromat aplanat but they already had a thread in 1.30 oil achromat aplanat since at least the mid. 1970s and released a very similar 1.40 achromat aplanat along with the Diastar introduction, used primarily in the Diastar.
I have used both of them. Nice , precision well corrected condensers. They are 24mm

Here is the patent for the AO cat. # 2110 1.30 achromat aplanat. https://patents.google.com/patent/US4060306A/en. This patent shows as expired after 1994 but then it shows Leica Microsystems as the current assignee???
Leica also is the current assignee for the over the top 11 element in 8 group achromat aplanat patented by George Aklin and Richard Seidenberg at B & L a little earlier. I don't know if it ever went into production but it was likely aimed at the high N.A. objectives designed for the Research Balplan.

There is a bit of a back story to the Reichert Plan Fluor Apos too, if you are interested.
Last edited by apochronaut on Sat May 04, 2024 2:24 pm, edited 3 times in total.

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#158 Post by Rorschach » Sat May 04, 2024 2:12 pm

I am always interested in anything that has to do with the Polyvar lineage, optics etc. :)

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#159 Post by apochronaut » Sat May 04, 2024 2:25 pm

Here is the patent for your 1.30 condenser. Note the position of the exit pupil.

https://patents.google.com/patent/US391 ... +condenser

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#160 Post by apochronaut » Sun May 05, 2024 2:52 pm

Rorschach wrote:
Sat May 04, 2024 2:12 pm
I am always interested in anything that has to do with the Polyvar lineage, optics etc. :)
I am not sure when the Polyvar 2 was introduced but I am pretty sure it was sometime after 1985. The Reichert U.S.A. Diastar was introduced in 1985 or 1986. The standard complement of objectives were planachros from 2.5X, 4X, 10X, 20X, 40X and 2- 100X oil and a short series of Neoplans, 2 variations of 10X , 40X, 50X oil and 100X.
In the 1986 brochure there is also listed 4 Plan Fluor objectives. 10X, 40X, and 2-100X. These were made in Austria and it is highly likely that the production of them went hand in hand with the intended production of Plan Fluors for the Polyvar, dubbed Plan Fl Apos.
y
AO had recently produced a short set of 34mm parfocal Plan Apos which could have been put into longer barrels and offered for the Microstar IV/ Diastar. There was a 10X .30 and 40X .80 and I can find a patent for a 100X which I don't think ever came to market but probably by the time they were tweaked to go into 45mm barrels by 1986, they would have been pretty expensive, especially with U.S. labour costs. The 40X was already $800.00 in it's 34mm incarnation. It probably made more sense economically to tweak the existing 4 , 45mm parfocal Austrian Plan Apos into Diastar barrels but then they would have been too expensive as well. The market would have been limited, and the Plan Achros very good already.

It seems that the two factories likely decided to consolidate production on a Plan Fluorite line. Both needed a mid-priced , well colour corrected option.

In using and comparing all of the existing options that overlap between the AO/Reichert Plan Fluors and Reichert Plan Fl Apos , they appear to be the same objectives outside of the tiny over correction that was deliberately installed in the Plan Fluor versions.. The reason for this, remains obscure but it may have more to do with marketing than design.
In comparing the 40X examples for instance, both are .70, have the same working distance, front surface diameter and curvature, back surface diameter and curvature, coating, optical stack length, visual aperture, and optical performance. Only the physical barrel is different and presumably an element somewhere made of a different glass , likely in the back group that provides the needed overcorrection.

Personally, I think the decision out of Austria to call those objectives Plan Fl Apos a poor one. I can directly compare the Plan Achromat, Plan Fluor/Plan Fl Apo and Plan Apo objectives for a number of magnifications in BF , DF and in most cases phase as well, although that is a little trickier due to diaphragm diameter and condenser alterations that have to be observed. Those then are 10X with .25 .30 and .32 apertures, 40X with ..66, .70, .75 and 1.0 apertures and 100X with 1.25, 1.30 and 1.32 apertures , The difference between the achromat offerings and the fluorite offerings is fairly obvious, especially in terms of colour correction but not so much in resolution. There is not so much difference in phase and the working distance of the comparable objectives similar. The biggest performers of the Plan Fluor/ Plan Fluor Apo set are the 2.5X .07 ..and the 40X .70 , which seems every bit a .85 objective but wth a more forgiving coverslip tolerance. It is a great high quality deep field objective for quick and fairly thick fresh mounts where oversize protists and some debris might be.
However, none of the Plan Fluors are Apos. and in dubbing them as such, it gives a false impression of what those objectives are capable of but it may have provided an opportunity to elevate the price some. The difference is particularly noticeable in phase between the 100X 1.30 Plan Fl Apo and the 100X 1.32 Plan Apo, and this is with a .90 phase condenser, so the Plan Apo might be at some slight disadvantage. Stained blood corpuscles for instance, have a shaded , defined 3 dimensionality to them with the Plan Apo. You can see them in depth, their curvatures and angle of presentation distinct and almost lifelike. With the Plan Fl Apo the same corpuscles, although more vivid and defined than when viewed through a Plan Achro, are flat, lacking in many of the visual cues that flesh out the details through the Plan Apo. The 40X .70 performs the best of the phase options and I would have to say is the best of the group, the closest approaching Apochromat performance.
It works well with 15 or 16X eyepieces to yield a clean 600 to 640X magnification too. Both factories were pretty deficient in their supply of 60-63X objectives. There were non from the Buffalo side , although they left a space for one in their cat. numbering scheme, so did intend to trott one out eventually. Austria produced a 63X 1.0 Plan Achro in both glycerin and oil immersion . Very uncommon to find one. I have seen exactly one infinity version in years and years of looking, although the 160mm finite 63X are common. Fortunately all of the 40X objectives take 15X or 16X eyepieces with ease. Both the 10X snd 16X Reichert eyepieces have a 51° view, so the apparent field is identical at 400X or 640X. In the Diastar, the 10X eyepiece has a 41° view and the 15X a 51° view, so the apparent field in the Diastar at 600X is 20% wider than at 400X and in fact the same as that of the Polyvar.

There was a 63X .80 34mm parfocal semi- planachro made in Austria for the Neovar 2 and used briefly towards the mid. 80's by AO for the series 100 microscopes. There are examples of that objective around, fitted with an 11 mm extension at the factory. It thus becomes parfocal with the Poly objectives, so could be used as a dry 63X if need be. It is barely non-plan with a 20mm f.o.v. but would be moreso at a 24mm field.

Both factories also made a 50X oil immersion objective. The Reichert Austria Plan ( achromat) was made very early in the program, should be iris equipped based on Reichert's one objective for all contrast methods policy but shows in the catalogue as not being, and I have never seen one to verify. By 1983 it seems to have been dropped. The U.S.A. objective is a 45mm parfocal 50X .80 derived from the 34mm parfocal # 1016 50X .80 Plan Achro w./ iris but this time issued without an iris and as a Neoplan. All were intended as lower magnification partners to iris equipped 100X objectives for principally, DF.

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#161 Post by Rorschach » Sun May 12, 2024 4:34 pm

That is interesting stuff!

The Polyvar in our unit, which I assembled today, has five different 40x objectives. Three came with the scope and two in a parts package.

This lead me to think that maybe having two nosepieces for transmitted would be a good idea. One would be for dry objectives and the other for oil immersion. Which leads to the question that did Reichert ever produce an oil immersion25x for this system, PlanApo or Plan?

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#162 Post by apochronaut » Sun May 12, 2024 6:22 pm

Just the two dry 25X. A Plan ( Achro) .45 and Plan Apo .65 Any literature that I have seen says neither were IK but I know the .45 Plan existed as IK.The 25X Plan Apo is not a common objective.
Certainly, using the 40X 1.0 oil planapo w./iris and 100X 1.32 oil planapo w./iris is a dream pair for DF. The 40X will easily handle your 16X oculars too, so plus the mag. changer , in any application where you are trying to optimize the use of oil immersion , you could theoretically have a quite broad magnification range getting down to 320X, immersed in oil. Not too far above the 250X a 25X would provide but with the mag. changer a 25X will duplicate a 20X as well, or going the other way, a 10X can duplicate a 20X with the mag. changer. See the other thread. viewtopic.php?f=5&t=13173

At some point, they veered from the stated policy of installing iris diaphragms in all objectives over .75 N.A., so that an installed objective set could be used for all possible contrast methods. I know this because the relatively recently made( painted barrel) 63X 1.0 glycerin immersion objective I have does not have an iris. In order to use it for DF however, I discovered that the threaded return spring retaining rear diaphragm of several objectives that I looked at had the same thread. Replacing the 63X diaphragm with the very slightly smaller diapragm from a 100X 1.25 Plan Achro reduced the aperture down to the .90 or .85 required for DF.
If you ever come across a similar 63X oil, the diaphragm swap is a solution then, as an oil complement for DF but you don't want to go too small : just enough.
That one is the only over .75X objective outside of phase objectives I have seen, that has not had an iris. Perhaps it were made for the Diavar 2, which was a similar stand to the Diastar but probably not marketed in North America.
My objective came from a former Buffalo area dealer that worked closely with the Buffalo factory. I think he did some market testing for them and had several prototype designs. It's possible that 63X was such, since it was new. AO prototypes I saw though, were marked as such. They used the term, " submission piece". and had a typed strip of tape attached with a submission date and in the case of an example I own others I saw, had "approved" marked on the label too. B & L prototypes he had also did not have a label.

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#163 Post by Rorschach » Sat May 18, 2024 6:55 am

Thanks again, good info!

As the unit's Polyvar will be used quite a lot in teaching, we will definitely use a dry 40x during those times. Having students mess around with immersion is just dumb, asking for trouble. So no, that will be reserved for other, more experienced users.

I have now scanned all three original Polyvar user manuals that I have in my possession: the English, German and Finnish language versions. Here's a link to the Dropbox folder where they can be downloaded:
https://www.dropbox.com/scl/fo/5aqk6b9m ... 7gbox&dl=0

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#164 Post by apochronaut » Sat May 18, 2024 8:29 am

The scans are all very helpfull to fill in the gaps. Since you have 5 - 40X objectives, I am hoping you have the 40X .70 Plan Fl Apo . Since your plan is to avoid the use of immersion with your students I have found it one of the most perfect objectives for general use. Despite having only a modest N.A. at .70, whereas the Plan Achro is .75, it's superior colour correction elevates it well above the achromat plus it has almost twice the working distance. The Plan Achro is .29mm and the Plan Fl is .48, so there is a better margin of safety plus some allowance for thicker specimens. It tolerates cover slip variances more.
There is also a welcome safety feature on any of the tighter tolerance dry objectives that I have looked at closely. It may be on all of them, oil too. There is a barrier ring around the front lens that is just a very tiny distance extended beyond the front lens, guarding against the possibility of driving the lens into a cover slip. The guard guarantees that the cover slip may be sacrificed, not the front lens. A welcome feature where students are concerned.

I have not found any literature that covers the Plan Fl Apo objective specs. From the 1986 Diastar brochure that does profile them, only labelled Plan Fluor, here are the ones listed. I know of one other. A 2.5X Plan Fl Apo
10X .30 N.A. W.D. 4.27mm
40X .70 N.A. W.D. .48mm
100X 1.30 N.A. W.D. .15mm

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#165 Post by Rorschach » Sat May 18, 2024 11:16 am

Since it seems that a lot of people didn't have these user manuals, just the one for the Polyvar Met (which is quite a bit different and does not cover many important aspects at all), I thought scanning them in good quality would be the obvious service for the community to do.

Also, I have never found any of the three language versions on the web, only the same circulating scan of a Polyvar Met user manual. These three original printed manuals for the transmitted/biological Polyvar are in fact the only ones in print that I have ever seen...and you can believe I've looked high and low for them! I hope the manuals will be of use to people.

Yes, that 40X .70 Plan Fl Apo is one the five 40x lenses. Good to know that it is of superior quality!

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#166 Post by Rorschach » Sun May 19, 2024 5:49 am

Here's images of the objectives on the unit's Polyvar.

First of the 40x lenses has badly worn markings, but 'IK' is visible:
20240512_113911.jpg
20240512_113911.jpg (78.51 KiB) Viewed 77 times
20240512_113922.jpg
20240512_113922.jpg (78.96 KiB) Viewed 77 times
Next is a Plan Iris IK 40x:
20240512_114024.jpg
20240512_114024.jpg (98.79 KiB) Viewed 77 times
20240512_114002.jpg
20240512_114002.jpg (102 KiB) Viewed 77 times

Then the 40x Plan phase:
20240512_114111.jpg
20240512_114111.jpg (94.97 KiB) Viewed 77 times

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#167 Post by Rorschach » Sun May 19, 2024 5:57 am

A second photo of the 40x Plan phase:
20240512_114141.jpg
20240512_114141.jpg (113.16 KiB) Viewed 74 times
Then the Plan Apo Oel Iris 40x (no mention of 'IK' on this one):
20240512_114248.jpg
20240512_114248.jpg (97.06 KiB) Viewed 74 times
20240512_114301.jpg
20240512_114301.jpg (101.76 KiB) Viewed 74 times
20240512_114309.jpg
20240512_114309.jpg (110.58 KiB) Viewed 74 times

Rorschach
Posts: 386
Joined: Sun Mar 10, 2019 11:44 am

Re: Reichert Polyvar

#168 Post by Rorschach » Sun May 19, 2024 6:01 am

And two photos of the Plan Fluor 40x (no mention of 'IK', also no mention of 'Apo') and the Plan Fl Apo IK 10x:
20240512_120057.jpg
20240512_120057.jpg (68 KiB) Viewed 74 times
20240512_120028.jpg
20240512_120028.jpg (72.41 KiB) Viewed 74 times
Two photos of the polarizers. The condenser is very heavy for it's size:
20240509_145105.jpg
20240509_145105.jpg (57.81 KiB) Viewed 73 times
20240509_145330.jpg
20240509_145330.jpg (69.6 KiB) Viewed 73 times

apochronaut
Posts: 6376
Joined: Fri May 15, 2015 12:15 am

Re: Reichert Polyvar

#169 Post by apochronaut » Sun May 19, 2024 12:44 pm

The 40X .70 Plan Fluor is based on a third generation system designed for the series 400 microscopes. The Plan Achro and Plan Fluor objectives in that program were all for interference contrast. Examples of working DIC Diastars for instance are fitted with standard Plan Achros and Plan Fluors. It seems they were all made strain free according to literature but not marked as such because it was universal. Those objectives came out in 1985 or 86.

A 40X .70 Plan Fl Apo for the Austrian made instruments is the same objective as the Plan Fluor as far as performance goes, in it's chosen system, and in fact the same optical design, except for it being engineered deliberately to have a greater level of lateral colour. The IK designation might not have been actually necessary because the objectives were probably made strain free anyway but obviously Reichert in Austria felt they wanted to verify that on the barrel.
I know the " greater level of lateral colour" doesn't make any sense because it is called an Apo but it is really just a Fluorite., and a good one too.

Here is the explanation for the increased level of lateral colour that is fully compensated for by the time the image comes out of the eyepieces.

When the Univar/Polyvar optical system was designed 55 years ago , it was based on the early 60's AO patent covering the infinity corrected design. Ten years at least had elapsed since that design initiative, which was pretty basic and involved what today would essentially be student objectives. The Univar needed a much wider field, a broader contrast arsenal and superior corrections, so in order to accomodate the more all encompassing system that AO envisioned, improvements were made to the infinity system , primarily to flatten the much wider field and plan apochromats were designed. I just linked some optics patents to several threads in the last couple of days. When it came to apochromatic performance, it was not possible at the time to correct for all aberrations in the objective, or at least not practical in a 45mm barrel. The correction for each aberration or distortion is unique, so when spherical aberration is corrected for, chromatic aberration is introduced and visa versa. The same is true to different degrees for coma, astigmatism, field curvature etc. Thus , it makes sense to fully correct in the objective for all that is practical and leave further corrections for downstream optics. Thus , that early AO/Reichert system corrected for everything in the objective but in doing so, installed 1% residual axial colour. This residual colour was compensated for , .6% in the tube lens and a further .4% in the eyepieces. All AO or Reichert infinity optical systems designed prior to 1985 adhere to this essential design but the 34mm optical system compensates more in the tube lens and less in the eyepiece. That is why you can put the newest Polyvar objective in the oldest AO infinity stand and it will work well. You cannot put the oldest AO infinity objective in a Polyvar and have it work well , simply because of the imperfect nature of the objective but many of the post 1970 AO objectives with an F.N. of 24 or greater will work in a Polyvar, especially if you install an 11 mm collar on them to get them parfocal.
The infinity space turned out to be a problem in the pre-1985 system.. Putting an 11 mm spacer on an objective, yes that amount of lengthening is o.k., or a polarizer , or slider but it turned out that due to the compensation required in the tube lens, when stretching that infinity space beyond a certain point, aberrations would infiltrate the system. So, stacking components or very thick components such as the AO 10/20 Bertrand lens assemhly maxed out or exceeded the corrective capacity of the system. According to the designers of the third generation system, and they were by and large responsible for the 1 and 2 as well , once the infinity space was lengthened more than 25mm, further compensation of some type had to be applied to the system.

So, when it came time to design the series 400 or the third generation, there was a desire to remove the compensation out of the tube lens, so the infinity space was free to be stretched more. The first improvement was to reduce the residual axial colour to .6%. The second was to remove the compensation from the tube lens and install it in the head. They appear to possibly have put a small amount in the eyepieces too but I guess that doesn't really matter.

The upshot of all this is that you can install the older Reichert Austria 45mm objectives in a Diastar, as I do, and they work fine as long as you also use eyepieces with .4% compensation. It turns out that the cat.# 145 eyepieces designed for the StereoStar 580 fit the bill and provide a slightly greater than 20mm f.o.v. Once I get it done, I wiil install a set of Polyvar eyepieces in the Diastar and have a 24mm f.o.v. but only when using older Reichert 45mm objectives or extended AO 34mm objectives.

However, installing the series 400 objectives in a Polyvar, will likely lead to some problems of over compensation and yield slight residual ca. Since .4% compensation has been installed in the Reichert eyepieces that will be where the extra compensation will occur, since the .6% is already compensated for in the tube lens. Thus a neutral set of 30mm eyepieces would probably work well with the series 400 40X Plan Fluor but since it is essentially the same objective as the Plan Fl Apo, probably eventually sourcing one of those would allow it's full ability to be used with other Polyvar objectives.

That 10X .30 Plan Fl. Apo by the way should have a 4.3 mm w.d. , so it should coordinate well with your oil immersion nosepiece and even give you 20X with the mag. changer.