Last spring I purchased a Zeiss Standard scope with phase contrast condenser, due to family and work issues, I couldn't test it for correctness. I had difficulties getting Kohler illumination, not matter what I did, there was too much light "bleed" where the iris never got sharp or had a halo around it. After spending many hours, it seems like the the lamp source was not aligned correctly(per Fig 13 in the document below), I only see one one coil when testing the light source never the mirror image. It appears like the concave mirror aligned at the factory is out of collimation, is this something which can be fixed or should I buy a new lamphouse setup.
http://www.science-info.net/docs/zeiss/ ... tor100.pdf
http://www.science-info.net/docs/zeiss/ ... tor100.pdf
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
I think you are describing two different problems, which although related are adjusted independently. If you follow the directions in the manual you should be able to adjust the mirror so that the two filament images are visible. The screws for the adjustments are on the back side of the lamp house and are marked in the instructions. Once you have aligned the mirror you can adjust the Kohler illumination. In my experience, the halo around the image of the iris is due to some issue within the condenser.
If you are having problems adjusting the mirror it may be necessary to check the motion of the mirror as you manipulate the adjusting screws.
These steps work for me, but it is difficult to diagnose without seeing the problem.
If you are having problems adjusting the mirror it may be necessary to check the motion of the mirror as you manipulate the adjusting screws.
These steps work for me, but it is difficult to diagnose without seeing the problem.
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
The initial setup is done off-microscope with the lamp about 2 - 3 meters from a wall as it suggests in the manual you posted.
Once the basics are correctly setup you can then make the final adjustments with it on the microscope.
Additionally, as part of the overall setup make sure that both the lamp and the condenser are centered.
http://zeiss-campus.magnet.fsu.edu/tuto ... flash.html
Start the interactive tutorial in the link above by centering the element.

Once the basics are correctly setup you can then make the final adjustments with it on the microscope.
Additionally, as part of the overall setup make sure that both the lamp and the condenser are centered.
http://zeiss-campus.magnet.fsu.edu/tuto ... flash.html
Start the interactive tutorial in the link above by centering the element.

Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
Hvae tried those processes, here is what I see turning the light beam from the lamp two meters.
- Attachments
-
- File_000 (1).jpeg (33.44 KiB) Viewed 2791 times
-
- File_000.jpeg (63.23 KiB) Viewed 2791 times
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
thanks, I agree that the two issues could be unrelated. First things first is there some detailed instructions on how to align the lamphouse mirror.lorez wrote:I think you are describing two different problems, which although related are adjusted independently. If you follow the directions in the manual you should be able to adjust the mirror so that the two filament images are visible. The screws for the adjustments are on the back side of the lamp house and are marked in the instructions. Once you have aligned the mirror you can adjust the Kohler illumination. In my experience, the halo around the image of the iris is due to some issue within the condenser.
If you are having problems adjusting the mirror it may be necessary to check the motion of the mirror as you manipulate the adjusting screws.
These steps work for me, but it is difficult to diagnose without seeing the problem.
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
Have you checked inside the lamp house to determine that there is a mirror ? I have seen a few that have no reflecting mirror and the single image is all that you get. A single filament image does not fill the field as well, but does work.
Once you have solved the lamp filament issue you can begin to align the Kohler.
Once you have solved the lamp filament issue you can begin to align the Kohler.
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
lorez wrote:Have you checked inside the lamp house to determine that there is a mirror ? I have seen a few that have no reflecting mirror and the single image is all that you get. A single filament image does not fill the field as well, but does work.
Once you have solved the lamp filament issue you can begin to align the Kohler.
yes there is a concave mirror behind the lamp and collector lenses in front of the lamp.
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
.
Last edited by Yvan on Sun Dec 04, 2016 4:50 pm, edited 1 time in total.
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
Picture was taken without the frosted glass diffuser, but through the collector system.Yvan wrote:I don't have experience with the Zeiss "Leuchte 100" (I'm looking for one, but they're not cheap, these days...), but I alligned lots of Reichert Illuminators (Lux US, Lux FNI) for all kinds of Reichert scopes (Zetopan, Diapan, MeF, ...).pdxhiker wrote:thanks, I agree that the two issues could be unrelated. First things first is there some detailed instructions on how to align the lamphouse mirror.lorez wrote:I think you are describing two different problems, which although related are adjusted independently. If you follow the directions in the manual you should be able to adjust the mirror so that the two filament images are visible. The screws for the adjustments are on the back side of the lamp house and are marked in the instructions. Once you have aligned the mirror you can adjust the Kohler illumination. In my experience, the halo around the image of the iris is due to some issue within the condenser.
If you are having problems adjusting the mirror it may be necessary to check the motion of the mirror as you manipulate the adjusting screws.
These steps work for me, but it is difficult to diagnose without seeing the problem.
Alligning the concave or parabolic mirror behind the bulb was never a problem. It's rather straight forward, but it takes some patience. Well, a lot of patience, actually.
In most Reichert illuminators, the mirror is factory aligned and mount with only one (FNI) or two (Lux US) screws. I suppose Zeiss is something similar.
Try again, but put the illuminator some 50 cm from a wall...
Is the picture from the bulb filament taken trough the collector system of the illuminator?
Some Zeiss illuminators have a non removable frosted glass somewhere in the system, or even worse: a frosted collector.
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
I tried fixing the lamphouse and this is what It looks like right now.
- Attachments
-
- File_000 (2).jpeg (42.57 KiB) Viewed 2732 times
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
Whatever you did, you were successful. Onward and upward.
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
Looking much better. Can you get them any closer?
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: Fixing Zeiss 100 Illuminator Concave mirror focus
I fully dismantled the lamphouse and noticed the concave mirror was not fully seated in the two bronze metal holders gripping it, now the question I'll get to them closer and post another picture.
Compound
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube
PH Condenser Ph2,3, BF, DF | 6.3x Plan, 16x,25x, 40x,100x(Oil),Neofular Ph2 & Ph3, 40x PlanAPO Iris | 10x Eyepiece
Standard 14/15/18 (not sure) | Illuminator 100 with 12V PS| Trinocular tube