#1 Post by Sabatini » Wed Feb 26, 2020 12:15 am

Hi, everybody.
I was able to get my hands on a Mitutoyo finescope fs100.
I took it apart completely so I could clean it up and I found this beautiful mirror I think it's a pelicle.
It has a lot of mold and fungus in it, but I'm AFREID to just bathe it in distilled water.
Please, what should I do?
Would it be possible to make a recommendation to clean it without damaging it or should I leave it as it is.
I want to show the microscope to this great group of microscopists from whom I have learned a lot.
It came with the uwf 10x 30 eyepieces
objectives M plan apo 2x 5x 10x 20x.
Thank you in advance for imparting so much knowledge to me.
Attachments
IMG_20200225_184319.jpg
IMG_20200225_184319.jpg (115.69 KiB) Viewed 9999 times

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#2 Post by Sabatini » Wed Feb 26, 2020 1:21 am

Screenshot_2020-02-25-18-31-44-372_com.google.android.apps.photos.jpg
Screenshot_2020-02-25-18-31-44-372_com.google.android.apps.photos.jpg (55.02 KiB) Viewed 9991 times
Attachments
Screenshot_2020-02-25-18-32-02-320_com.google.android.apps.photos.jpg
Screenshot_2020-02-25-18-32-02-320_com.google.android.apps.photos.jpg (39.62 KiB) Viewed 9991 times

Hobbyst46
Posts: 2585
Joined: Mon Aug 21, 2017 9:02 pm

Re: Mitutoyo finescope fs100

#3 Post by Hobbyst46 » Wed Feb 26, 2020 10:35 am

Hi,
Searching the web about the fs100, I find that (if new) it has both top and bottom illumination. Makes sense, since it is primarily a measurement microscope.
So I guess that the mirror you show is actually a half-mirror, namely a 50/50 beamsplitter, that is part of the top illuminator. To reflect the light beam onto the sample and pass the reflected light back to the camera and eyepieces.
If that is correct, it is has been coated with a reflective layer. It seems to be damaged.
I think that the only remedy is to find a replacement. Beamsplitters are available from various optical part suppliers, like Newport, Edmund etc.
Hope this helps.
Zeiss Standard GFL+Canon EOS-M10, Olympus VMZ stereo

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#4 Post by Sabatini » Wed Feb 26, 2020 12:07 pm

Hobbyst46
Thank you for your kind cooperation.
When I got the microscope it was covered with a layer of dust and dirt, so I haven't tried it yet, what I did was to blow the dirt off with a hair dryer and started to disassemble it immediately, the electrical part works perfectly, the prisms of the head are in very good condition.
I think I'll leave the half mirror (beansplit) as it is, without touching it, I don't want to risk damaging it more.
I'll be commenting on its real performance once I finish and test it.
In your search for the fs100, didn't you find the user manual? I didn't find anything and I've been trying to find it for days, not even Mitutoyo's page.
Would you be so kind as to pass me the link

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#5 Post by Sabatini » Wed Feb 26, 2020 12:25 pm

IMG_20200226_071941.jpg
IMG_20200226_071941.jpg (82.16 KiB) Viewed 9944 times
IMG_20200226_071922.jpg
IMG_20200226_071922.jpg (82.83 KiB) Viewed 9944 times
IMG_20200226_071910.jpg
IMG_20200226_071910.jpg (85.91 KiB) Viewed 9944 times
IMG_20200226_071850.jpg
IMG_20200226_071850.jpg (84.5 KiB) Viewed 9944 times

Hobbyst46
Posts: 2585
Joined: Mon Aug 21, 2017 9:02 pm

Re: Mitutoyo finescope fs100

#6 Post by Hobbyst46 » Wed Feb 26, 2020 12:25 pm

Sabatini wrote:
Wed Feb 26, 2020 12:07 pm
Hobbyst46
Thank you for your kind cooperation.
When I got the microscope it was covered with a layer of dust and dirt, so I haven't tried it yet, what I did was to blow the dirt off with a hair dryer and started to disassemble it immediately, the electrical part works perfectly, the prisms of the head are in very good condition.
I think I'll leave the half mirror (beansplit) as it is, without touching it, I don't want to risk damaging it more.
I'll be commenting on its real performance once I finish and test it.
In your search for the fs100, didn't you find the user manual? I didn't find anything and I've been trying to find it for days, not even Mitutoyo's page.
Would you be so kind as to pass me the link
Regretfully, I did not find a manual. I visited the Mitutoyo site but they only speak of measuring devices, not microscopes. Sorry that I cannot help with that.

EDIT: here is a link to this scope in the other forum. Guys there are knowledgeable. Perhaps try to approach the fellow who bought it ?
https://photomacrography.net/forum/view ... 21a7981c32
Zeiss Standard GFL+Canon EOS-M10, Olympus VMZ stereo

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#7 Post by Sabatini » Wed Feb 26, 2020 2:24 pm

Hobbyst46

Very kind of you.
Do you think I should try to clean it up, or is it more likely to be damaged than improved?

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#8 Post by Sabatini » Wed Feb 26, 2020 2:29 pm

IMG_20200226_092455.jpg
IMG_20200226_092455.jpg (120.96 KiB) Viewed 9922 times
IMG_20200226_092734.jpg
IMG_20200226_092734.jpg (92.11 KiB) Viewed 9922 times

Hobbyst46
Posts: 2585
Joined: Mon Aug 21, 2017 9:02 pm

Re: Mitutoyo finescope fs100

#9 Post by Hobbyst46 » Wed Feb 26, 2020 3:39 pm

Sabatini wrote:
Wed Feb 26, 2020 2:24 pm
Do you think I should try to clean it up, or is it more likely to be damaged than improved?
Such coatings might be very sensitive. I would only blow off the dust, then try to use it as is.

A nice set of objectives! the /0 mark means that they are to be used on specimens without coverslips.
Zeiss Standard GFL+Canon EOS-M10, Olympus VMZ stereo

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#10 Post by Sabatini » Wed Feb 26, 2020 5:13 pm

Hobbyst46
I appreciate all your support
Thanksyou.

microb
Posts: 233
Joined: Wed Mar 27, 2019 6:39 am

Re: Mitutoyo finescope fs100

#11 Post by microb » Wed Feb 26, 2020 7:32 pm

https://www.newport.com/n/how-to-clean-optics

Buy Kimtech's kimwipes, fold one over and rip in half, then use that rough edge as a little brush with an alcohol/solvent, never brush dry. Don't use q-tips. Don't use microfiber clothes. They're too abrasive. Kimtech is supposed to not remove coatings. Microfiber does remove coatings. So those microfiber clothes grind eyeglasses of those LED screen coatings people paid extra for. Store the kimtech in a zip lock bag, or if it has been sitting there in the open for a while, throw away the first kimwipe as a sacrificial one to avoid the accumulated abrasive dust from the air.

Newport mentions "a mix of 60% acetone† and 40% methanol" and "acetone-impenetrable gloves" Some people worry about mounting cements with acetone. Haven't confirmed anything about it, other than any cements I wanted to remove on Leica parts ignored my efforts with acetone.

If you do use acetone, use latex gloves to avoid putting human oils into the solution. Avoid vinyl as it dissolves, very slowly, but still dissolves in acetone and technically pollutes the solution as well.

SIDE NOTE ABOUT SEARCHING THE INTERNETS: Some people like coin collectors who also want to remove organics off of surfaces will post about how dangerous acetone is, and not because it is flammable like alcohol but basically because they fear the smell. You can touch acetone. The human body makes it after all. The smell is pungent though. People on the Atkins diet have bad breath because of ketosis, which is the ketone of acetone in their breath. You have to be pretty intentional to find a way to O.D. on acetone.

Scarodactyl
Posts: 758
Joined: Sat Mar 03, 2018 9:09 pm

Re: Mitutoyo finescope fs100

#12 Post by Scarodactyl » Wed Feb 26, 2020 9:08 pm

I have not had any trouble with q-tips removing coatings, for the record. Plenty of people seem to use them without trouble, though kimwipes seem like they should be a better option if available.

MicroBob
Posts: 1987
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Mitutoyo finescope fs100

#13 Post by MicroBob » Wed Feb 26, 2020 9:10 pm

When it comes to the cleaning of very delicate surfaces the cleaning medium is often less of a problem than abrasive dirt on the surface is. It gets wiped over the surface and scratches it.
For this application the half silvered mirror in its momentary state might actually be good enough as it is only used for lighting purposes. I have recently bought a (smaller) half silvered mirror from Aliexpress.

@Scarodactyl: To make sure: Did or did they not remove coatings?

Bob

Scarodactyl
Posts: 758
Joined: Sat Mar 03, 2018 9:09 pm

Re: Mitutoyo finescope fs100

#14 Post by Scarodactyl » Wed Feb 26, 2020 11:35 pm

No removal of coatings (obviously I can't guarantee this in every case, I just haven't had trouble with it). You definitely have to be careful of abrasive grit or dust having settled on the surface you're cleaning as you say.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#15 Post by Sabatini » Wed Feb 26, 2020 11:58 pm

Well, I think these instructions encourage me to take the initiative to try to do something about cleaning the half mirror.
Although Hobbyst46 only recommends that I blow it out with air, because of the delicate coating.
I'm going to try to clean the edges first, a little bit on the edges only... and see what happens.
thank you allá for sharing your knowledge.

microb
Posts: 233
Joined: Wed Mar 27, 2019 6:39 am

Re: Mitutoyo finescope fs100

#16 Post by microb » Thu Feb 27, 2020 12:33 am

The Newport instructions give that as a first step to try (air blowing).

Scarodactyl
Posts: 758
Joined: Sat Mar 03, 2018 9:09 pm

Re: Mitutoyo finescope fs100

#17 Post by Scarodactyl » Thu Feb 27, 2020 3:28 am

Blowing should be a first step in this type of situation for sure. That mold will need something more, but this is an industrial scope and was once in an industrial environment, so all sorts of crazy grit could have gotten on it.

MicroBob
Posts: 1987
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Mitutoyo finescope fs100

#18 Post by MicroBob » Thu Feb 27, 2020 10:42 am

If the part itself can handle it washing under flowing water with a bit of dish cleaning agent is quite a mild method of cleaning.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#19 Post by Sabatini » Thu Feb 27, 2020 1:28 pm

Greetings,.
The option of giving him a bath of soap with water I think it would work, I'll try it like this and then I'll show you how it looks.
Thank you

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#20 Post by Sabatini » Thu Feb 27, 2020 2:52 pm

Well, thank you for your instructions, what I did was, dish soap water, shake it into the soap mixture, then rinse it with distilled water, 1 full bottle and dry it with a cotton facial cleanser, then blow it with the hair dryer in cold mode.
The fungus didn't come out all the way, I see that there is something left there, but I am not going to go further, I rubbed it almost nothing, just rinse it with the soap and for the moment you can see the mirror effect well, we will have to wait to try it.
Now I am going to focus on the focusing mechanism, because as Scarodactyl said, it feels full of thick residue.
IMG_20200227_091639.jpg
IMG_20200227_091639.jpg (78.25 KiB) Viewed 9806 times
IMG_20200227_085902.jpg
IMG_20200227_085902.jpg (59.17 KiB) Viewed 9806 times
IMG_20200227_085859.jpg
IMG_20200227_085859.jpg (68.61 KiB) Viewed 9806 times
IMG_20200227_085914 (1).jpg
IMG_20200227_085914 (1).jpg (78.96 KiB) Viewed 9806 times
Screenshot_2020-02-27-09-41-17-098_com.google.android.apps.photos.jpg
Screenshot_2020-02-27-09-41-17-098_com.google.android.apps.photos.jpg (74.05 KiB) Viewed 9806 times

User avatar
daruosha
Posts: 267
Joined: Wed Oct 23, 2019 7:10 am
Location: Tehran, Iran

Re: Mitutoyo finescope fs100

#21 Post by daruosha » Thu Feb 27, 2020 3:40 pm

You can try the solution of %5 Hydroxide Peroxide at your own risk. I had a Canon FD 50mm f/1.4 and it developed a massive fungus on the front lens while it was in my parents basement for more than 2 decades. I wondered that I cannot make it worst so let's do a bit of percussive maintenance and soak the front lens in H2O2 for a few hours and to my surprise, it worked quite well.

H2O2 can be a bit aggressive (specially high concentrations) on lens oatings, but it was an old lens (my favorite childhood object in the whole universe) and I suspect it didn't have any coating and no harm was done in the end.

%5 Hydroxide Peroxide is on the safe side and risk of damaging coatings is very low.
Daruosh.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#22 Post by Sabatini » Thu Feb 27, 2020 5:26 pm

MicroBob wrote:
Wed Feb 26, 2020 9:10 pm
When it comes to the cleaning of very delicate surfaces the cleaning medium is often less of a problem than abrasive dirt on the surface is. It gets wiped over the surface and scratches it.
For this application the half silvered mirror in its momentary state might actually be good enough as it is only used for lighting purposes. I have recently bought a (smaller) half silvered mirror from Aliexpress.

@Scarodactyl: To make sure: Did or did they not remove coatings?

Bob
Well thanks for the recommendation, but as microbob recommends not to be too demanding with this and since its purpose is only the lighting pass true, I do not see the excessive need to leave it completely clean, I would have liked to remove everything completely, but I am satisfied, we will see how it works.

User avatar
daruosha
Posts: 267
Joined: Wed Oct 23, 2019 7:10 am
Location: Tehran, Iran

Re: Mitutoyo finescope fs100

#23 Post by daruosha » Thu Feb 27, 2020 5:39 pm

Well, if you are happy then of course the best approach is : "if it ain't broke, don't fix it".

Unfortunately I am the textbook example of OCD and every single bit and piece of my equipments must be shiny. Sometimes i cannot sleep at nights thinking of a piece of dust on my condenser or a bad /cold solder joint on a PCB. It's madness.

I have a feeling that your devil inside is whispering the fungus existence on the mirror in your ears, every time you look at the microscope. :twisted:
Daruosh.

MicroBob
Posts: 1987
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Mitutoyo finescope fs100

#24 Post by MicroBob » Thu Feb 27, 2020 8:03 pm

I have experienced two versions of fungus:
1. One that forms a dust like layer that cleans off with no problem
2. One that has deeple etched (from acid residues of the fungi) that can't be removed as it has already damaged the glass

For me it is most important to make sure the fungus doesn't come back quickly. In case 2 a perfect repair isn't possible so I would prefer to do less damage by excessive cleaning. Generally with these half silvered mirrors the cleaning options are limited in comparison to a naked or coated glass surface.

To the result of this cleaning looks very good and I don't think that it will impact the image.

PeteM
Posts: 978
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Mitutoyo finescope fs100

#25 Post by PeteM » Thu Feb 27, 2020 9:50 pm

I've had good luck killing and removing lens fungus with a 50-50 solution of household ammonia and hydrogen peroxide. Note that both will already come diluted in their household and medical (maybe 3% H2O2) versions.

Don't know if they would make further inroads on the spot of fungus that seems to remain, if they'd damage the coating, or if the coating and glass has already been etched as Bob notes can often happen.

Main point - you want to be sure you kill any spores and then store the scope at low enough relative humidity that it doesn't come back.

Worst case, I'd think you could source another half mirror of an appropriate size and replace the one you have? Those are excellent objectives and you should end up with a top quality scope.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#26 Post by Sabatini » Fri Feb 28, 2020 2:55 am

Well ,for the moment I think I have to finish restoring what is left, now I am concentrating on the focusing system, then I will have to clean the electrical part, the capacitor and the eyepieces are finished. And then when everything is in place I will try how everything works, if it is too poor what I observed I will repair it as I find it, especially with the issue of alignment ....Of course we've all been whispered to by that little devil, that reminds us that something is not perfect in our microscopes, but my level of disturbance has led me to commit irreparable damage that I do not want to repeat.
Thank you for your support.
I will continue to show the progress, so I have noticed this microscope has some rough finishes, I imagine that being a destined for the industry lacks direct level of sophistication.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#27 Post by Sabatini » Fri Feb 28, 2020 3:27 am

MicroBob wrote:
Thu Feb 27, 2020 8:03 pm
I have experienced two versions of fungus:
1. One that forms a dust like layer that cleans off with no problem
2. One that has deeple etched (from acid residues of the fungi) that can't be removed as it has already damaged the glass

For me it is most important to make sure the fungus doesn't come back quickly. In case 2 a perfect repair isn't possible so I would prefer to do less damage by excessive cleaning. Generally with these half silvered mirrors the cleaning options are limited in comparison to a naked or coated glass surface.

To the result of this cleaning looks very good and I don't think that it will impact the image.
Thank you for the good advice. You always have very interesting views to share.
I have enjoyed your contributions and talks with the best disposition to acquire knowledge of your experience, and attentive and always waiting for a new topic of my interest.

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#28 Post by Sabatini » Sat Mar 07, 2020 2:24 am

[/quote]Regretfully, I did not find a manual. I visited the Mitutoyo site but they only speak of measuring devices, not microscopes. Sorry that I cannot help with that.
Hobbyst46

Cordial greetings.
Searching and searching , I could find these instructions, they are not the user's manual nor are they the same model, but they are very close, they are from the fs110
Fs110T code 378-121A
378-122A
I'm almost done putting all the parts together.
24373954.pdf
(125.45 KiB) Downloaded 150 times

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#29 Post by Sabatini » Mon Mar 30, 2020 6:07 pm

Cordial greetings.
I was finally able to finish assembling the industrial microscope
Mitutoyo fs100, thanks to all of you for your suggestions, your contributions helped me a lot to renew it and leave it as good as possible. It was necessary to clean all the optical elements, reason why I had a lot of extra work, especially the head with all the prisms and the collimation afterwards, the objectives were not disassembled for obvious reasons... All the optics are acceptable, what impressed me most was the great field of vision it has, very nice and relaxing.
I wanted to ask you to help me again with your suggestions.
As I have no experience with reflected light microscopes, I have no reference to compare with.
I notice that this microscope (in reflected light) produces images that are totally lacking in contrast, flat and of very poor quality...completely washed out and cloudy. Looking for the possible fault, I notice that there is a reflection of light that bounces off the back of the lens to the eyepieces when the light enters it. I also notice that the microscope has a filter missing right at the entrance of the reflected light. The filter can be moved with a lever that turns it 180 degrees or half a turn.
Now if only an explanation were possible that would help improve this. Since by placing two external lights on each side of the sample, the contrast is excellent and the colors very saturated, helped by the planapos objectives, bet
Anyway I am very satisfied with the knowledge and experience gained first hand and by the forum of microscopists here present who share their knowledge.
Thank you.[
IMG_20200328_195301.jpg
IMG_20200328_195301.jpg (65.7 KiB) Viewed 3315 times
IMG_20200328_195208.jpg
IMG_20200328_195208.jpg (59.38 KiB) Viewed 3315 times
IMG_20200328_195148.jpg
IMG_20200328_195148.jpg (63.82 KiB) Viewed 3315 times
IMG_20200328_195003.jpg
IMG_20200328_195003.jpg (64.23 KiB) Viewed 3315 times
attachment=4]IMG_20200328_195139.jpg[/attachment]
Attachments
IMG_20200328_195139.jpg
IMG_20200328_195139.jpg (59.28 KiB) Viewed 3315 times

Sabatini
Posts: 102
Joined: Tue Nov 12, 2019 3:09 am

Re: Mitutoyo finescope fs100

#30 Post by Sabatini » Mon Mar 30, 2020 6:10 pm

IMG_20200328_195258.jpg
IMG_20200328_195258.jpg (70.01 KiB) Viewed 3315 times
IMG_20200328_195051.jpg
IMG_20200328_195051.jpg (63.28 KiB) Viewed 3315 times
IMG_20200328_223123.jpg
IMG_20200328_223123.jpg (73.14 KiB) Viewed 3315 times
IMG_20200328_195232.jpg
IMG_20200328_195232.jpg (65.13 KiB) Viewed 3315 times
IMG_20200328_195245.jpg
IMG_20200328_195245.jpg (55.35 KiB) Viewed 3315 times

Post Reply