Leica DM IL and PHACO 2 And 160/0.17
Leica DM IL and PHACO 2 And 160/0.17
Hi,
I've been mostly focused on metallurgical and metrology microscopes, so I have not used inverted. So with an inverted microscope like the Leica/Leitz DM IL (pre-LED model), what PHACO 2 objective is valid to use? The only ones I'm finding are not 160/- but 160/.017's and that would mean a cover glass of .17mm. How does a cover glass factor into an inverted set up.
And does anyone know what microscope models go with a Leitz Wetzlar 170/0.17 PL APO 16/0.40? I'm assuming a 160 set up would not work.
Thanks,
Ted
I've been mostly focused on metallurgical and metrology microscopes, so I have not used inverted. So with an inverted microscope like the Leica/Leitz DM IL (pre-LED model), what PHACO 2 objective is valid to use? The only ones I'm finding are not 160/- but 160/.017's and that would mean a cover glass of .17mm. How does a cover glass factor into an inverted set up.
And does anyone know what microscope models go with a Leitz Wetzlar 170/0.17 PL APO 16/0.40? I'm assuming a 160 set up would not work.
Thanks,
Ted
Re: Leica DM IL and PHACO 2 And 160/0.17
See page 14: http://www.microscopy-uk.org.uk/mag/img ... es7rev.pdfAnd does anyone know what microscope models go with a Leitz Wetzlar 170/0.17 PL APO 16/0.40? I'm assuming a 160 set up would not work.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: Leica DM IL and PHACO 2 And 160/0.17
Modeled the PHACO2 S55 annulus. Will try printing out a PH1 based on some data. I have a S55 PH2 here to check things against.
Will then try to test the microscope with a PHACO1 objective.
Will then try to test the microscope with a PHACO1 objective.
- Attachments
-
- 34.PNG (113.61 KiB) Viewed 7851 times
Re: Leica DM IL and PHACO 2 And 160/0.17
So I'm remeasuring again.
Thanks,
Ted
Thanks,
Ted
Last edited by microb on Sun Mar 08, 2020 4:44 am, edited 2 times in total.
Re: Leica DM IL and PHACO 2 And 160/0.17
Ted, With an inverted scope you're typically viewing THROUGH something like a Petri dish. An objective spec'd for .17mm cover slips could be used with the slide upside down. Most objectives for inverted scopes are spec'd at something like 1, 1.2, 1.4mm etc. Some will have a correction collar that goes from something like 0 (no cover) to 2mm. At low magnifications (say, 10x) this won't matter so much. Once you get up to and beyond 20x, you want the right corrections.microb wrote: ↑Sat Mar 07, 2020 7:08 pmHi, I've been mostly focused on metallurgical and metrology microscopes, so I have not used inverted. So with an inverted microscope like the Leica/Leitz DM IL (pre-LED model), what PHACO 2 objective is valid to use? The only ones I'm finding are not 160/- but 160/.017's and that would mean a cover glass of .17mm. How does a cover glass factor into an inverted set up. . . .
Thanks, Ted
As far as phase condensers and annuli go, inverted scopes often have a long working distance condenser of somewhat lower numerical aperture than the 1.25+ you might be used to in an upright compound scope.
Re: Leica DM IL and PHACO 2 And 160/0.17
Hi Pete,
So what you are stating about the thickness of a petri dish is what led me to posting my question. The .17mm thickness is not applicable to an inverted. So what do people use on a Leica DM IL? It has PH2 and PH3 (or at least people are selling the annulus slide loaded with such), but all those higher objectives are 160/.17 not 160/-.
So are people selling DM IL's with anything other than PH1 and DF selling an invalid configuration?
Thanks,
Ted
So what you are stating about the thickness of a petri dish is what led me to posting my question. The .17mm thickness is not applicable to an inverted. So what do people use on a Leica DM IL? It has PH2 and PH3 (or at least people are selling the annulus slide loaded with such), but all those higher objectives are 160/.17 not 160/-.
So are people selling DM IL's with anything other than PH1 and DF selling an invalid configuration?
Thanks,
Ted
Re: Leica DM IL and PHACO 2 And 160/0.17
Fixed the mask dimensions for PH2. Inner diameter is 5mm. Outer is 5 + .42 + .42. Web fingers are .6mm wide.
- Attachments
-
- 5_.42mm.PNG (134.89 KiB) Viewed 7844 times
Re: Leica DM IL and PHACO 2 And 160/0.17
So for PH1, 3mm inner, 3_.38_.38 outer. Fingers .53.
- Attachments
-
- PH1_3_.38.PNG (137.14 KiB) Viewed 7843 times
Re: Leica DM IL and PHACO 2 And 160/0.17
If it says "-" the objective is relatively insensitive to how much glass is between the objective and specimen. Usually lower power objectives.
If it says "0.17" then it expects not much more than a thin glass cover slip (.17mm thick) between the objective and speciment.
Here's an example of a Leitz objective that adjusts to different thicknesses of glass (Petri dish bottom, slide thickness, etc.) between the objective and specimen in an inverted scope: https://www.ebay.com/itm/Leitz-Microsco ... SwjXBbv7Bb
Here's a LWD (Long Working Distance) type that would be suitable for an inverted scope: https://www.ebay.com/itm/Leitz-20-0-32- ... Sw8~1eV~nk
If it says "0.17" then it expects not much more than a thin glass cover slip (.17mm thick) between the objective and speciment.
Here's an example of a Leitz objective that adjusts to different thicknesses of glass (Petri dish bottom, slide thickness, etc.) between the objective and specimen in an inverted scope: https://www.ebay.com/itm/Leitz-Microsco ... SwjXBbv7Bb
Here's a LWD (Long Working Distance) type that would be suitable for an inverted scope: https://www.ebay.com/itm/Leitz-20-0-32- ... Sw8~1eV~nk
Re: Leica DM IL and PHACO 2 And 160/0.17
There are petri dishes with .17 bottoms: https://www.tedpella.com/section_html/p ... dishes.htm
Some microscopists make their own: http://www.microscopy-uk.org.uk/mag/ind ... vert2.html
Some microscopists make their own: http://www.microscopy-uk.org.uk/mag/ind ... vert2.html
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: Leica DM IL and PHACO 2 And 160/0.17
Thanks.
160/0.6 to 1.6
160/0.6 to 1.6
Re: Leica DM IL and PHACO 2 And 160/0.17
From personal knowledge - commercial, rather than DIY-ed, dishes are recommended. They can be rinsed with water, dilute caustic, ethanol, and recycled. Avoid strong organic solvents like acetone or other ketones.75RR wrote: ↑Sun Mar 08, 2020 7:57 amThere are petri dishes with .17 bottoms: https://www.tedpella.com/section_html/p ... dishes.htm
Some microscopists make their own: http://www.microscopy-uk.org.uk/mag/ind ... vert2.html