by apochronaut » Fri Dec 25, 2020 1:21 pm
It is hard to determine what will happen in a non native objective once a phase annular diaphragm image passes through the condenser and gets focused. Having messed around with them quite a bit in AO systems, it seems that in order to get a reasonable phase match , even if you start with the same size diaphragm, the condenser focus itself must be precise, the condenser mechanical distances must be precise and the optical pathway to the phase plate in the objective must be very close to the original design parameters. Part of that last step includes the magnification of the annular diaphragm image up to the point in the objective that the phase plate lies. I don't have any B & L 160mm phase objectives on hand but I do have 38 AO and Reichert phase objectives and 4 Balplan objectives plus a few odds and sods of others.
My intuition tells me the B & L 160mm phase objectives would be very close to AO 160mm phase objectives in terms of the location of the phase plate in the objective. The size and composition might vary a bit but not by much. AO was half way around the track by the time B & L got up to the starting gate on phase and never caught up, so dollars to doughnuts, they had a complete AO phase system in their secret room of competitor's microscopes and modelled their first system quite a bit on it. So did Nikon, so did Olympus, so there need be no shame involved.
So, I have a Balplan phase condenser in hand and several AO phase condensers. The first thing that is of note is that AO only used an abbe condenser on their very early systems. Although not marked as such, the series 4 and earlier condensers seems likely a 1.25 abbe condenser but one with a specific very short rear focal point that fixes on the annular diaphragm only a few mm below it. You cannot bring one of their standard 1.25 abbe condensers into the system and have it work. Abbe condensers aren't the best at working dry. A 1.25 abbe is designed to be oiled but phase works at sub 1.0 N.A. levels for all objectives, so for their next act, AO built a 3 element in two groups dry achromat phase condenser, # 1242 of probably an N.A. of .90 or higher. It's specifications are undisclosed but I have examined it extensively and compared it to the later, # 1201 which is marked as a .90 achromat/aplanat. When oiled, each work at about 1.3 but the point of focus changes.
With the condensers, the B & L design and the AO design part company. B & L continued with a 1.30 abbe for the Dynoptic series and then used a 1.25 abbe aspheric for the Balplan. One similarity though, is thst both B & L in the Balplan condenser and AO in the series 10 condenser house a refocus lens below the phase diaphragms, presumably to bring the illumination beam to a longer focus at the annular diaphragms, since the phase condenser must have a short rear focus. An odd difference is, the annular diaphragms in the AO systems are relatively consistent in their increase in diameters from 10X through to 100X and the Balplan diameters are very similar to those in the AO condenser, except the 100X, which takes a big leap, being almost twice the diameter of the AO equivalent in the 160mm condenser. In other words, there is a huge difference between the diameter of the 40X Balplan annulus and the 100X. Going back to my postulàtion that the 160mm Dynoptic system would fairly closely resemble the 160mm AO system, the 160mm 100X B & L annulus might be much smaller than the diameter of the Balplan annulus.
Looking at various objectives, the phase plate in them is not always at the rear focal plane. Usually, yes but I have one objective where it occurs deep in a plan fluorite design, near to the front of the objective. Looking at the Balplan 100X planachromat, the phase plate as well is deep inside the objective, close to the front of the objective. The phase plate by comparison for the 20X planachromat is deposited on the last surface before the image leaves the objective, above the objective shoulder! This is all complicated by the fact that unlike the 160mm B & L objectives, the Flat Field and Planachromats magnify only 1/5 of their face value. The 100X is actually a 20X and so forth.