Re: Zeiss Standard 14

#31 Post by Glycolyse » Thu Jan 17, 2019 7:08 pm

75RR wrote:Could you post a photo of a larger subject (an alga for example) with the field diaphragm in view (just visible along the edge of the field of view) and sharply focused.
x6,3 obj:

Field diaphragm sharply focused (50% of the field of view): https://image.noelshack.com/fichiers/20 ... 195050.jpg
Field diaphragm sharply focused (90% of the field of view): https://image.noelshack.com/fichiers/20 ... 195432.jpg

Both pictures were taken with the condenser's diaphragm opened at 90%. Sorry for the very low quality...

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#32 Post by 75RR » Thu Jan 17, 2019 7:15 pm

It still looks rather uneven. Have you tried to focus the filament per my previous post + image on the condenser diaphragm? That might help.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#33 Post by Glycolyse » Thu Jan 17, 2019 7:32 pm

Thanks for the quick reply 75RR ! Indeed, there are patches of bright light in the background. I did not try yet to focus the filament, because I wanted to show you how it was before trying it. Here were the steps:

1) Remove any opaque filters that may be in the light path.
2) Close the condenser iris completely
3) Place a mirror or slide under the condenser, angle it so that you can see the closed iris
4) You should be able to see the filament coils reflected on the condenser iris. (This is the Front Focal Plane)
5) If no filament coils are visible or they do not look sharp you need to move the light source either in or out until the filament coil is sharply focused.


Actually, I don't know if there are an opaque filters, and how to remove them.
image on the condenser diaphragm
Do you want a picture of the condenser diaphragm through one removed eyepiece ?

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#34 Post by 75RR » Thu Jan 17, 2019 7:47 pm

Just had a look at a connector tube which is similar to the one you have, the diffuser/opaque filter is incorporated in it. It can be removed but not worth it at the moment.

What I am thinking is that the mirror or the lens which is part of the field housing is broken. You may have to take it off.

The Field Housing is held in by two screws.

1) Take anything that may fall off when you turn the microscope on its side off first.

2) Mark the Field Housing's position with a pencil so that you can place it in the same position again.

3) It comes out through the top of the base so if you are holding the microscope upside down or nearly (to get to the screws) be careful that it does not pop out and hit the floor.
Attachments
field-housing.jpg
field-housing.jpg (42.92 KiB) Viewed 14381 times
Last edited by 75RR on Thu Jan 17, 2019 8:23 pm, edited 2 times in total.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#35 Post by Glycolyse » Thu Jan 17, 2019 8:20 pm

I can't say I'm not struggling on this one... I don't know if what I see is the filament or something else. As you will see, the image is repeated quite a few times, compared to your's on the slide.

https://image.noelshack.com/fichiers/20 ... 211326.jpg

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#36 Post by Hobbyst46 » Thu Jan 17, 2019 9:11 pm

Glycolyse wrote:I can't say I'm not struggling on this one... I don't know if what I see is the filament or something else. As you will see, the image is repeated quite a few times, compared to your's on the slide.

https://image.noelshack.com/fichiers/20 ... 211326.jpg
A few months ago I struggled with the field aperture, the problem was that the mirror detached from its 45degrees holder plate and in addition, the mirror itself was worn out - it was (apparently by mistake) a second-surface glass mirror. And the reflecting surface had deteriorated. Aided by close support from 75RR, Apochronaut, MicroBob and others, I replaced it with a first surface metal mirror, quite cheaply and the illumination improved tremendously. So it can be solved.

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#37 Post by Glycolyse » Fri Jan 18, 2019 1:02 am

Oh my, guys. I had a good laugh :lol: . Indeed, there were patches of bright light in the background.

But do you know why ? Well, here's why: https://image.noelshack.com/fichiers/20 ... 013043.jpg

Yeah... There were these small things from the packaging beneath the microscope. Some were between the light path (from the halogen bulb to the mirror). No wonder why the light wasn't evenly distributed :lol:

Now, the light is brighter and evenly distributed. The quality of the image increased too !

Moss, 6,3x obj: https://image.noelshack.com/fichiers/20 ... 013549.jpg

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#38 Post by 75RR » Fri Jan 18, 2019 6:25 pm

I like simple solutions :) Well done!
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#39 Post by Glycolyse » Fri Jan 18, 2019 7:19 pm

Thank you 75RR !

I took some green marine algae samples today. Here are some photos.

obj x16 - Sample 1 - Focus stacking of 20 images.

Image

obj x40 - Sample 2 - Focus stacking of 20 images.

Image

It's not perfect and everything, but I'm fine with the result. What do you think ?

By the way, the x16 obj and the x40 bj seem to be blurred, especially the x40. I think they need to be cleaned (as well as the oculars probably). I found on one thread of this forum a pdf of Zeiss named The Clean Microscope - There are a lot of informations, so I will take time on this one. Every tip for cleaning my objectives and oculars are welcomed !

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#40 Post by 75RR » Fri Jan 18, 2019 7:45 pm

Looking much better. Nice oblique!

I wonder if you can notice a difference in the brightness if you move the bulb holder in and out a bit.

If you can notice a difference try for a bright yet even illumination.


As to cleaning optics 'The Clean Microscope' is an excellent resource.

Always good to clean used objectives (and eyepieces if dirty) when you first get them - that way you know where you stand.

Do not clean them too often though, make sure you cover the microscope and use a blower to remove dust.

The exception of course are oiled objectives which you must clean every time.

I use absorbent paper to remove most of the oil first (tap do not wipe) then use cotton swabs (see page 7) with a drop or two of isopropanol.

Do not use swabs more than once. Discard and use a new one after a single use. Use as many as you need. Breathe on lens with last one, use dry.
Last edited by 75RR on Fri Jan 18, 2019 8:53 pm, edited 1 time in total.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#41 Post by Glycolyse » Fri Jan 18, 2019 8:42 pm

Thank you 75RR !

I think I achieved Phase contrast on both x16 and the x40 objectives:

x16 obj: https://image.noelshack.com/fichiers/20 ... 212629.jpg
x40 obj: https://image.noelshack.com/fichiers/20 ... 213105.jpg

There is a bright halo on every subject. It's a bit messy since I don't have a Phase telescope, and not correctly focused, but wanted to know if Phase contrast looked like this.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#42 Post by 75RR » Fri Jan 18, 2019 8:55 pm

That looks about right.

Did you order a Phase telescope yet?
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#43 Post by Hobbyst46 » Fri Jan 18, 2019 8:57 pm

For phase contrast, a different specimen would be better. Cheek cells, or protozoans (slowed down to ease the viewing) - a truely thin and transparent specimen would be better indicatives of the quality of phase contrast than the algae.
In my opinion, the alignment of the condenser is not optimal.

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#44 Post by Glycolyse » Fri Jan 18, 2019 9:15 pm

75RR wrote:Did you order a Phase telescope yet?
I won't be able to order anything before the end of February. During this time, I'll check for a few things on Ebay/other websites:

- Zeiss trinocular head
- Zeiss oculars
- Phase telescope
- Zeiss Lens Cleaning, Isopropanol
- Zeiss Phase Contrast x60 dry objective (160)
- 8 Form test plate of diatoms from Mr.Klaus D. Kemp, which 75RR suggested in another thread
- (...)
Hobbyst46 wrote:In my opinion, the alignment of the condenser is not optimal.
Thank you Hobbyst, I definitely need a Phase telescope I guess.

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#45 Post by Hobbyst46 » Fri Jan 18, 2019 9:51 pm

1. Phase telescopes have been recently posted "for sale" by forum member Scarodactyl.
2. Dry 60X phase objective: There is a Zeiss Planachromat 63X0.90 ph3 objective, its working distance is very small (see the brochure "Zeiss Optical Systems" p. 53), might be difficult to use.
3. Zeiss occulars - the WF, high eyepoint (for glass wearers) version are much better than the non-WF version.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#46 Post by 75RR » Sat Jan 19, 2019 5:17 am

If you can borrow 25 euros from a kindly soul there is this one in Poland. They do not come much cheaper than that. PZO is one of the very good brands.

https://www.ebay.co.uk/itm/23-2mm-Phase ... :rk:7:pf:0
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#47 Post by 75RR » Sat Jan 19, 2019 5:31 am

You might also want to call in several xmas + birthday gifts in advance:

https://www.ebay.co.uk/itm/Zeiss-Micros ... :rk:2:pf:0
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#48 Post by Glycolyse » Fri Feb 15, 2019 11:04 pm

Hi everyone,

Is this item good enough to clean my objectives and oculars ? https://www.amazon.fr/Alcool-Isopropyli ... sopropanol

Image

Otherwise, where do you get yours ?

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#49 Post by Hobbyst46 » Fri Feb 15, 2019 11:58 pm

Glycolyse wrote:Hi everyone,

Is this item good enough to clean my objectives and oculars ? https://www.amazon.fr/Alcool-Isopropyli ... sopropanol

Image

Otherwise, where do you get yours ?
1. This is good enough for your purpose; 99.5% or 99% are likewise good enough. Have not tried lower grades; 70% alcohol may contain too much water.
2. I got mine from a store that caters scientific experiment equipment to schools. It is available from any chemical supply house, but probably not a small quantity.
3. In view of the hazards of the stuff, I would strongly suggest :
a) purchase and store at home only a small amount - say 0.5 liter or 16Oz. This will suffice for a very very long time.
b) store in a well-closed bottle in a ventilated place, preferably a cabinet that does not get too warm. Ordinary house temperatures are OK.
c) the storage place should be out of the reach of children.
d) for each session of use for microscope work, pour only a small amount - say, 10-20ml or less - of the alcohol into a small flask or container, so the
main quantity stays in the closed container all the time.
e) avoid pouring or using near open fire, cooking gas flames, etc, or electric sparks.
f) use in a reasonably ventilated place - not in an all-closed space.
g) as a backup safety measure against spillage, put a glass/metal/resistant/ plastic wide flat container (tray with raised rims, a backing plate etc) directly under the alcohol bottle (the bottle stands inside the container), so if the bottle is cracked or leaking, at the worst, the liquid will be contained and not spread around.
4. The alcohol is fulluy miscible with water.

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#50 Post by Glycolyse » Sat Feb 16, 2019 12:50 pm

Thanks for the tips Hobbyst. Is there anything to add in order to clean my microscope's objectives & oculars ?

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#51 Post by Hobbyst46 » Sat Feb 16, 2019 1:15 pm

Most welcome!
Microscope parts can suffer from over-cleaning of the glass surfaces. Blow away the dust first, inspect surfaces through an inverted eyepiece that serves as magnifier, against sideways light. If further cleaning is needed, exhale on the glass to cause water condensation, then gently wipe with lens cleaning tissue or lens cloth (can be obtained from eyeglasses shops). One-way strokes only, not back and forth movements. Never rub a dry glass surface.
Oil and grease are removable with the isopropanol or petrol-ether (gasoline, octane etc) - but use sparingly. Apply with a Q-tip or a moistened lens cleaning tissue. In very difficult cases xylene can help, but apply it very very carefully, very sparingly, since it can dissolve old optical cements.

ChrisR
Posts: 138
Joined: Tue Jun 14, 2016 11:01 pm
Location: Surrey, UK

Re: Zeiss Standard 14

#52 Post by ChrisR » Sat Feb 16, 2019 2:26 pm

I find small spray-bottles (where you finger provides the propulsion) are great for convenient quantities of IPA, acetone, xylene, methanol etc. I use a weaker solution of IPA with some aqueous antistatic agent for my spectacles!

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#53 Post by Glycolyse » Sun Feb 17, 2019 6:35 pm

Thanks both.

In this video, Maintenance is showed: https://www.youtube.com/watch?v=jM2lVHa ... x=4&t=337s

It looks about right for me. Yet, he didn't show the cleaning of the objectives. Is it the same process ?

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#54 Post by Glycolyse » Thu Mar 14, 2019 9:05 pm

Hi everyone,

On my Zeiss condenser (46 52 70), I noticed there were 3 big lenses, which I think are slots for rings, such as D-stop etc. Ideally, I'd like to purchase some filters to see the differences between them.

Image Image

The issue is the filters have a precise diameter (30, 31, 35mm etc.), and I can't tell what diameter my condenser's slots are. I can't find anything by typing the the number of the condenser on google. How can I know the diameter then ?

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#55 Post by Glycolyse » Fri Mar 15, 2019 12:37 pm

Update with pictures of the said-slots:

One with a lens:

Image

The other one, above this one, without a lens:

Image

Anyone has an idea ?

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#56 Post by 75RR » Fri Mar 15, 2019 12:43 pm

Zeiss filters are 32mm
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#57 Post by Glycolyse » Fri Mar 15, 2019 1:34 pm

Thanks ! I didn't know Zeiss condensers' filters were standardized.

I just received Isopropanol solution and a dust blower, gonna clean everything for the first time ;)

Hobbyst46
Posts: 4294
Joined: Mon Aug 21, 2017 9:02 pm

Re: Zeiss Standard 14

#58 Post by Hobbyst46 » Fri Mar 15, 2019 4:48 pm

Glycolyse wrote:Thanks ! I didn't know Zeiss condensers' filters were standardized.
Of the two auxiliary swing-out parts that you show, the top one contains a lens that is used to fill the low magnification objectives with light. That lens is fixed, not easily removeable. The bottom one is a filter carrier. I hope that Rheinberg and other special effect filters in that carrier will work for you. On my setup it did not.

I had thought that by "empty slots" you meant positions within the condenser turret itself. These might serve better for Rheinberg and other special effect filters, since they are at almost the same plane as the iris aperture. However, their diameter is not 32mm as far as I recall.

User avatar
Glycolyse
Posts: 129
Joined: Sat Dec 22, 2018 10:07 pm

Re: Zeiss Standard 14

#59 Post by Glycolyse » Sat Mar 16, 2019 5:48 pm

I didn't know that, thanks for the clarification Hobbyst !

I have one question. When looking through the x40 objective, the subject seems blurry, at any focus point given. As I received my isopropanol and the dust blower, I could clean my eyepieces, objectives and condenser correctly. However, the blurry aspect still exists. I remember while looking at a Motic BA210E-LED microscope, it was not the case.
- Was it because the subject was "bigger" ? (Elodea sp. compared to observing diatoms)
- My eyepieces are not the original ones (Chinese)
- Halogen light ?
- Wrong cleaning process ?
- Wrong setup of the microscope (Condenser height, etc.) ?
- ...

While I look at diatoms pictures from other members (such as 75RR, but it's DIC so I can't really relate with this technique),it looks like the subject shows good details, when mine doesn't.

I'd like to fix this "issue", if possible, because I bet it comes from me.

Maybe I should post a photo of a subject observed from a x40 objectif.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: Zeiss Standard 14

#60 Post by 75RR » Sat Mar 16, 2019 6:29 pm

You should be able to focus all your objectives sharply.

If the problem is just with the 40x then the objective itself is most likely the cause.

Show us a photo taken with it and also a photo of the objective itself, get as close an image of its lens as possible.