How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#1 Post by Hobbyst46 » Sun Apr 14, 2019 6:13 pm

I am trying to refurbish and bring life to a very old Zeiss GFL stand, that was so far stored in a remote cupboard. There are two issues with it. One, regarding the field aperture assembly, will be described in a future post. The other challenge is the fine focus mechanism. The fine focus control functions but is too stiff and requires much more force than the normal. The coarse focus is fine. I suspect that the problem is due to old hardened grease.
Some time ago, a kind forum member has uploaded a link to a draft of a concise record, including sketches, written by his father, of disassembly and repair of the fine focus mechanism. I apologize for having forgotten the name of the poster - hopefully he will read this post and reappear! I followed the first steps as described, dismantled and separated the focus mechanism from the microscope arm, as shown below. Identifications and notations of parts are cited from the above record as far as I understand them.

The focusing gear moves the thin push-rod that in turn pushes the block against a stopper spring (photo(1), marked "Spring", and photo(3), left side of the block). The gear is actually hidden from sight by the block. The red circles symbolize invisible rollers that are caged within Vee-shaped grooves. The end of the push-rod itself is connected with a thin wire spring (photo (3), right side of the block, hardly visible). In the end face near that end of the push-rod, there is a retaining ring (photo (4)). It is locked with a side screw (not directly visible). That end face is touching the base plate of the whole assembled microscope. The face opposite the block is the female dovetail of the coarse focus, not related to the problem.

Complete disassembly of the mechanism appeared tricky. The heads of six screws are "locked" with red Loctite, against tinkering by amateurs (me). I tried to go around it. Yet, drops of light machine oil into the ends of the Vee-shaped grooves, from the outside, made no difference. Removal of the end plate, cleaning and re-greasing the stopper spring (photo(3))did not help. Tinkering with the two bolts was futile - hopefully it did not worsen the situation. I did not try WD-40, nor heat: this gizmo is a combination of brass and steel parts, and each has its own expansion coefficients.

Further dismantling might require a custom pin-spanner to remove the retaining ring without deburring. But I am not sure that it will expose the problematic hard grease area.
So, what would you suggest to do next? thanks in advance for ideas! perhaps links to available maintenance manuals?
Attachments
fine focus mechanism (1).jpg
fine focus mechanism (1).jpg (115.22 KiB) Viewed 9237 times
fine focus mechanism (2).jpg
fine focus mechanism (2).jpg (172.3 KiB) Viewed 9237 times
fine focus mechanism (3).jpg
fine focus mechanism (3).jpg (173.71 KiB) Viewed 9237 times
fine focus mechanism (4) .jpg
fine focus mechanism (4) .jpg (112.69 KiB) Viewed 9237 times

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#2 Post by 75RR » Sun Apr 14, 2019 6:56 pm

My apologies if you know this, but in both the GFL and WL fine focus is tightened and loosened by turning the fine focus knobs against or away from each other.
Attachments
fine-focus-GFL.jpg
fine-focus-GFL.jpg (94.2 KiB) Viewed 9219 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#3 Post by MicroBob » Sun Apr 14, 2019 7:35 pm

75RR wrote:My apologies if you know this, but in both the GFL and WL fine focus is tightened and loosened by turning the fine focus knobs against or away from each other.
Are you sure about this? It is been a while since I had one of these apart but I think to remember a fine focus shaft that runs from one knob to the other. I might be wrong though.

@Doron: I think I have made a special tool for this job. I have a look tomorrow, I might even have a Zeiss Winkel GFL in the right state of disassembly and can maybe give a hint how to get into it. As far as I remember it was straightforward to disassemble apart from the one special tool.
To loosen screws it is good to apply some fine oil and then put the screwdriver in the slot and apply some knocks with a small hammer. This loosens whatever is in the thread and allows the oil to penetrate much better. The screwdriver should fit the slot in thickness and lenght and be ground well.

When reapplying lubricants it has to be taken into account that there is much surface in contact and grease at the wrong points gives too much damping for the fine focus. I would suggest to grease the ways and oil the shaft.

Bob

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#4 Post by Hobbyst46 » Sun Apr 14, 2019 9:41 pm

75RR wrote:My apologies if you know this, but in both the GFL and WL fine focus is tightened and loosened by turning the fine focus knobs against or away from each other.
MicroBob wrote:Are you sure about this? It is been a while since I had one of these apart but I think to remember a fine focus shaft that runs from one knob to the other. I might be wrong though.
Thanks 75RR and Bob,
Funny ! as occasionally mentioned on this forum, not all Zeiss Standard microscopes of the same model are entirely identical to each other!
I own two GFL scopes. One is in use, the other, which I believe is somewhat older, is under reconstruction. On both of them, the coarse focusing behaves exactly as described in the manuals: a special single control knob, coaxial with the coarse focusing knob, and fitted between the focusing knob and the arm, controls the tension of the coarse focusing movement.

The fine focus knobs are NOT the same. On the scope in use, the rightside and leftside knobs can be rotated in mutually opposing directions (from the operator's point of view). Such manipulation seems to affect the tension, yet since the fine focus works nicely I will not test it too much. Also, the knobs are uniformly knurled, no small screws on them.
In the scope under reconstruction, the one that bothers me, I cannot rotate them in opposing directions (I will not attempt with tools, though). Each knob has a small set screw that apparently locks it to the shaft. Whether this screw and its function are related to the tension of the focusing movement - I doubt but cannot say.
Bob, thanks for the tips. When I figure out what to do they will certainly help!

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#5 Post by 75RR » Mon Apr 15, 2019 5:37 am

In the scope under reconstruction, the one that bothers me, I cannot rotate them in opposing directions (I will not attempt with tools, though). Each knob has a small set screw that apparently locks it to the shaft.
If it is fixed, firm measured finger pressure will not break it. If it is not fixed then it might do the trick.

As it is a little too tight, you want to turn the fine focus knobs away from each other. As you face the front of the microscope turn the right one towards you - turn left one away.

Just a few mm will do.
Attachments
fine focus knob-.jpg
fine focus knob-.jpg (148.82 KiB) Viewed 9167 times
Last edited by 75RR on Mon Apr 15, 2019 7:42 am, edited 1 time in total.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#6 Post by Hobbyst46 » Mon Apr 15, 2019 7:32 am

75RR wrote:
In the scope under reconstruction, the one that bothers me, I cannot rotate them in opposing directions (I will not attempt with tools, though). Each knob has a small set screw that apparently locks it to the shaft.
If it is fixed, firm measured finger pressure will not break it. If it is not fixed then it might do the trick.

As it is a little too tight, you want to turn the fine focus knobs away from each other. As you face the front of the microscope turn the right one towards you - turn left one away.

Just a few mm will do.
I think that it makes sense, that thd user can simply adjust thd friction on the fine focus movement. Unforetunately, in this specific case, turning the knobs against each other is impossible. Perhaps some previous ownef tried and succeeded to block it, or it is blocked by a mass of old grease or dirt...

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#7 Post by 75RR » Mon Apr 15, 2019 7:41 am

... turning the knobs against each other is impossible ...
Not against ... away from each other - as in untighten
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#8 Post by Hobbyst46 » Mon Apr 15, 2019 8:10 am

Yes, this is what I have been trying to do. The right knob towards me, the left knob away. In other terminology: the right knob CCW, the left knob CW when viewed from the right side of the microscope. Does not move at all.

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#9 Post by MicroBob » Mon Apr 15, 2019 8:27 am

Hi 75RR,
I didn't see the atattchment at first because I read the thread on my smartphone. Thank you for posting it! Do you know if the Universal and Photomikroskop are supposed to work this way too? I thought that the WL and the big stands would share the same focus block design.

I had apart 3 GFL, most or all were Zeiss Winkel GFLs, so quite early ones. There clearly were differences, most visible was the nosepiece revolver. The oldest version was more rounded and had a matte surface. The design of the later nosepiece revolver was a lot better.

I atattch some pictures: Old Zeiss Winkel GFL and tool for removal of holding ring for the fine adjustment gears. This holding ring needs a lot of torque to loosen so you will need a strong tool. A machinst will be able to make one with my pictures and the microscope in front of him. The steel bar is just glued in with epoxy.
Before turning the holding ring a small setscrew has to be reomoved!

My fine focus shaft knobs have a set screw on one of them.

Zeiss
Attachments
ZW Feintrieb und Haltering.jpg
ZW Feintrieb und Haltering.jpg (450.16 KiB) Viewed 9157 times
Zeiss Winkel GFL Triebblock 1.jpg
Zeiss Winkel GFL Triebblock 1.jpg (486.31 KiB) Viewed 9157 times
Werkzeug Ausbau Feintrieb 3.jpg
Werkzeug Ausbau Feintrieb 3.jpg (482.49 KiB) Viewed 9157 times
Werkzeug Ausbau Feintrieb 2.jpg
Werkzeug Ausbau Feintrieb 2.jpg (455.13 KiB) Viewed 9157 times
Werkzeug Ausbau Feintrieb 1.jpg
Werkzeug Ausbau Feintrieb 1.jpg (328.14 KiB) Viewed 9157 times

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#10 Post by 75RR » Mon Apr 15, 2019 8:28 am

Oh well ... would have been nice if it had worked on your GFL
Do you know if the Universal and Photomikroskop are supposed to work this way too? I thought that the WL and the big stands would share the same focus block design.
As far as I know the WL shares the same focusing mechanism (block) with the Universal and the Photomicroscope, it also shares the same quick release stage and condenser holders.
Last edited by 75RR on Mon Apr 15, 2019 8:36 am, edited 2 times in total.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#11 Post by MicroBob » Mon Apr 15, 2019 8:28 am

more pics
Attachments
Halteschräubchen Haltering Feintrieb.jpg
Halteschräubchen Haltering Feintrieb.jpg (387.48 KiB) Viewed 9157 times
Klemmschrauben Feintrieb.jpg
Klemmschrauben Feintrieb.jpg (437.28 KiB) Viewed 9157 times

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#12 Post by Hobbyst46 » Mon Apr 15, 2019 9:22 am

Hi Bob,
Questions:
I see complete disassembly of the Winkel block. Including the roller bearings ?
What part of the mechanism adjusts the friction of movement ?
Is there a way to adjust without disassembly ?

Notes:
The Winkel mechanism in your pics is somewhat different from mine, but maybe is based on the same principles.
The special tool you fabricated for the ring is indeed the best design. I have used a similar tool in another mechanical task. On top of the cylinder, there was a 12mm(L)x12mm(W)x8mm(H) square bump or protrusion, for gripping with a socket wrench or an adjustable spanner, for tightening a large ring.
Last edited by Hobbyst46 on Mon Apr 15, 2019 12:11 pm, edited 1 time in total.

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#13 Post by Hobbyst46 » Mon Apr 15, 2019 11:58 am

75RR wrote:As far as I know the WL shares the same focusing mechanism (block) with the Universal and the Photomicroscope, it also shares the same quick release stage and condenser holders.
On a post from 1915 in Photomacrography forum, a post by KurtM, forum member Eddie has stated that:
"The two WL (Black and White) you can see in the second picture above. They are considered the small research scope since it has the same focus block as the large research stands the Universal/Photomicroscopes and they have the quick release/quick mount stage and condenser which is interchangeable with the large research stand. Below is the black WL alongside a couple of Photomicroscopes and a Universal. The large research scopes are quite large and take up a lot of room." On my GFL, the stage carrier (and condenser rack) are fixed to the arm with screws, not a quick-detach knob, in contrast to the larger Zeiss Standards.
(http://www.photomacrography.net/forum/v ... hlight=gfl)

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#14 Post by MicroBob » Mon Apr 15, 2019 12:06 pm

No roller bearings on this Zeiss Winkel GFL! The two guides are both brass dovetails.
The coarse focus is adjusted with the separate collar. I don't know whether there is a fine focus adjustment at all. To me it seems a bit like an emergency-solution to adjust the fine focus this way. The brass guides should always provide enough friction to keep the stage at height.

The earlier Standard Juniors also had brass guides, the later versions "Chromleiste" hat roller bearings like the later Standard RA and 14-18.
These early post war Zeiss West instruments were not yet as bullet proof as the later instruments. The sold vast amounts and improved the design soon.

I have a Phomi, first version, second type from 1961 on, but I haven't had time to do anything with it yet. The fine focus has a bit more friction than I would like so maybe the knob-twist-trick is just what I need to get it going.

User avatar
75RR
Posts: 8207
Joined: Sun Oct 12, 2014 2:34 am
Location: Estepona, Spain

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#15 Post by 75RR » Mon Apr 15, 2019 1:46 pm

Some more indirect information (not a Zeiss publication) on the fine focus adjustment on the Zeiss Universal (I imagine that includes the Photomicroscope as well)

See particularly Part 1 #1

http://asiimaging.com/docs/zeiss_universal_zdrive
Attachments
FFK.jpg
FFK.jpg (31.14 KiB) Viewed 9126 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#16 Post by Hobbyst46 » Mon Apr 15, 2019 2:03 pm

Hobbyst46 wrote:I am trying to refurbish and bring life to a very old Zeiss GFL stand, that was so far stored in a remote cupboard. There are two issues with it. One, regarding the field aperture assembly, will be described in a future post. The other challenge is the fine focus mechanism. The fine focus control functions but is too stiff and requires much more force than the normal. The coarse focus is fine. I suspect that the problem is due to old hardened grease.
Some time ago, a kind forum member has uploaded a link to a draft of a concise record, including sketches, written by his father, of disassembly and repair of the fine focus mechanism. I apologize for having forgotten the name of the poster - hopefully he will read this post and reappear! I followed the first steps as described, dismantled and separated the focus mechanism from the microscope arm, as shown below. Identifications and notations of parts are cited from the above record as far as I understand them.

The focusing gear moves the thin push-rod that in turn pushes the block against a stopper spring (photo(1), marked "Spring", and photo(3), left side of the block). The gear is actually hidden from sight by the block. The red circles symbolize invisible rollers that are caged within Vee-shaped grooves. The end of the push-rod itself is connected with a thin wire spring (photo (3), right side of the block, hardly visible). In the end face near that end of the push-rod, there is a retaining ring (photo (4)). It is locked with a side screw (not directly visible). That end face is touching the base plate of the whole assembled microscope. The face opposite the block is the female dovetail of the coarse focus, not related to the problem.

Complete disassembly of the mechanism appeared tricky. The heads of six screws are "locked" with red Loctite, against tinkering by amateurs (me). I tried to go around it. Yet, drops of light machine oil into the ends of the Vee-shaped grooves, from the outside, made no difference. Removal of the end plate, cleaning and re-greasing the stopper spring (photo(3))did not help. Tinkering with the two bolts was futile - hopefully it did not worsen the situation. I did not try WD-40, nor heat: this gizmo is a combination of brass and steel parts, and each has its own expansion coefficients.

Further dismantling might require a custom pin-spanner to remove the retaining ring without deburring. But I am not sure that it will expose the problematic hard grease area.
So, what would you suggest to do next? thanks in advance for ideas! perhaps links to available maintenance manuals?
EDIT: the above mentioned draft record of the repair of a Zeiss GFL fine focus was written by C.J. Matthews and was later posted by Graham Matthews, can be downloaded from MICROMAGUS.NET August 2006.
Last edited by Hobbyst46 on Tue Apr 16, 2019 6:50 am, edited 1 time in total.

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#17 Post by Hobbyst46 » Mon Apr 15, 2019 2:13 pm

75RR wrote:Some more indirect information (not a Zeiss publication) on the fine focus adjustment on the Zeiss Universal (I imagine that includes the Photomicroscope as well)...See particularly Part 1 #1
Thanks 75RR! so, there is a chance that loosening the left knob set screw will free the CW motion of that knob, such that the two-knob maneuvre can be accomplished... Actually I have already tried it (in vain) on the set screw of the RIGHT knob - simply because it was easier to do than the left knob set screw... both heads of those screws are already wounded by previous screwdrivers... We'll see!

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#18 Post by Hobbyst46 » Mon Apr 15, 2019 4:02 pm

So, I filed a watchmaker's screwdriver until it fits inside the slot of the focus knob set-screw, and loosened the left knob set screw - but the knob itself cannot be turned when the right focus knob is static and secured. Loosened both set screws - still no change. This becomes a separate riddle: what is the function of these knob set screws ? Perhaps pressing with a lever rod against the larger knob (the coarse focus knob) can help to push the fine focus knob outside, if the knob is slipped on the shaft rather than screwed on it.(?) :?
Bottom line
The above questions and discussion are now academic. The roller bearings can be serviced, but the fine focus gear itself is very delicate, the teeth on the lever that raises and lowers the block, as well as the teeth on the shaft that directly rotates that lever, are badly worn or broke. My findings are essentially the same as those of C. Matthews on his microscope. Until some miracle happens, this will be a purely coarse focus microscope...

PeteM
Posts: 2982
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#19 Post by PeteM » Thu Dec 02, 2021 8:24 pm

This thread proved useful to me and I thought I'd update it with my recent experience. A black Zeiss WL came to me with the focus block stuck. Initially trying to loosen things up with a bit of solvent and then oil (hope springs eternal) didn't work, so I started working my way inside. My version has a different fine focus mechanism. Same general idea of 2mm total movement and a spring-loaded pin lifting the stage - but an earlier version. What amazed (disheartened) me was the condition of the fine focus guide rails and the 3mm ball bearings (near 100 of them lined up in four brass retaining strips and guided between eight round rails). The whole thing was locked solid as if in cement, with a lumpy brown-green congealed grease the consistency of tar and coarse sand surrounding everything. FWIW, the coarse focus mechanism was stiff and some seriously oxidized grease, but pristine compared to this.

I've been inside several very old focus blocks and never seen something this bad. Don't know if it was the grease Zeiss used, several generations of people adding incompatible lubes, or what that caused this? However, seems it might be a common problem among 70 year old Zeiss blocks of this design.

The rails eventually cleaned up by soaking in solvent, picking away the coarse grease, wire brushing, polishing etc. Some of the ball bearings were degraded - but a hundred 3mm ball bearings are cheap to buy. It should go back together OK once those bearings arrive.

The main reason for posting is that if you come across one of these WL/GLF/Universal focus blocks with congealed grease, you might want to strip it down and clean it rather than try to force it and strip the fine focus gears. It looks as if mine will be OK through some combination of dumb luck, persistence, and a Zeiss design that looks as if the fine focus raising pin might be able to spring out of the way a bit if forced.

Hobbyst46
Posts: 4277
Joined: Mon Aug 21, 2017 9:02 pm

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#20 Post by Hobbyst46 » Thu Dec 02, 2021 10:15 pm

Hi PeteM, I am glad that the thread served some purpose.
What I did after that frustrating episode with the focus block was to secure a backup microscope stand.
And regretted altogether that I hung on to Zeiss - until I read about mechanical problems in Optiphots made 30 years later...

PeteM
Posts: 2982
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#21 Post by PeteM » Thu Dec 02, 2021 11:05 pm

. . . or a two-year-old scope made to sell for, say, $700 and already beginning to show its mechanical deficiencies two years in?

I'm grateful some of these old scopes can be returned to their German-Swiss-Japanese-English-US-Austrian mechanical and (often) optical excellence.

User avatar
ImperatorRex
Posts: 571
Joined: Fri Mar 30, 2018 4:12 pm
Location: Germany
Contact:

Re: How to remove stiffness from the fine focusing gear of Zeiss GFL ?

#22 Post by ImperatorRex » Fri Dec 03, 2021 4:29 pm

To be honest ... nothing more exciting than to touch and turn a well (re-) greased Zeiss Standard WL Fine focus knob (the nice looking metal knobs, not plastic). :-)