My Zeiss Microscopes
Re: My Zeiss Microscopes
Exactly! Nice drawing too!
If you rotate the polarizer in this setup would you see light diminution indicative of polarization.
If you rotate the polarizer in this setup would you see light diminution indicative of polarization.
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel
Youtube channel
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
I am very proud to present my second "non Zeiss" member in the microscope family: An Olympus BX60 microscope. This should be really my very very last scope:
Packing was taken seriously ... 114lb/52kg with wooden crate - air freight from oversea, has some costs but it safely and well arrived.
Packing was taken seriously ... 114lb/52kg with wooden crate - air freight from oversea, has some costs but it safely and well arrived.
Re: My Zeiss Microscopes
Congratulations! Very nice looking. Is that the DIC slider I see above the nosepiece?
As you say - well packed
As you say - well packed
They do say never say never ...This should be really my very very last scope
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: My Zeiss Microscopes
Very cool! Is it new, sure looks like it. What lenses and filters does it come with?
Must be quite exciting at your place with such a beautiful array of different scopes
Must be quite exciting at your place with such a beautiful array of different scopes
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel
Youtube channel
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hi 75RR,
thank you - yes it has DIK. Took quite long until I decided to buy this microscope - so hardly a day or night without dreaming about this microscope. Finally...dreams become true.
Dear Wes,
still an old scope, must be something like 20 or more years old I guess. But like new after fixing few smaller things & cleaning (re-greasing the noise piece and one eyepiece).
It has the older UIS objectives fitted UPLFL (10,20,40 + 100) and Fluorescense cubes for UV, Blue and Green erection are in the turret.
Main reason to get this nice scope was the possiblity to directly capture the full corrected intermediate image directly on the camera sensor (without okular/projective or camera objective).
thank you - yes it has DIK. Took quite long until I decided to buy this microscope - so hardly a day or night without dreaming about this microscope. Finally...dreams become true.
Dear Wes,
still an old scope, must be something like 20 or more years old I guess. But like new after fixing few smaller things & cleaning (re-greasing the noise piece and one eyepiece).
It has the older UIS objectives fitted UPLFL (10,20,40 + 100) and Fluorescense cubes for UV, Blue and Green erection are in the turret.
Main reason to get this nice scope was the possiblity to directly capture the full corrected intermediate image directly on the camera sensor (without okular/projective or camera objective).
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
A quick comparision of the Photo-Adaptions:
a) Olypus BX60 - UPlanFL40/0.75 PH2 - Canon EOS 600 (with flash)
Focus on Pleurosigma angulatum:
b) Zeiss IM35 - Plan Neofluar 40/0,90 Imm - SKpl W10/20 - YN35mm F2 - Canon EOS 650 (with flash)
Focus on Pleurosigma angulatum:
c) Olypus BX60 - UPlanFL40/0.75 PH2 - Canon EOS 600 (with flash)
Focus on Stauroneis phoenicenteron:
d) Zeiss IM35 - Plan Neofluar 40/0,90 Imm - SKpl W10/20 - YN35mm F2 - Canon EOS 650 (with flash)
Focus on Stauroneis phoenicenteron:
a) Olypus BX60 - UPlanFL40/0.75 PH2 - Canon EOS 600 (with flash)
Focus on Pleurosigma angulatum:
b) Zeiss IM35 - Plan Neofluar 40/0,90 Imm - SKpl W10/20 - YN35mm F2 - Canon EOS 650 (with flash)
Focus on Pleurosigma angulatum:
c) Olypus BX60 - UPlanFL40/0.75 PH2 - Canon EOS 600 (with flash)
Focus on Stauroneis phoenicenteron:
d) Zeiss IM35 - Plan Neofluar 40/0,90 Imm - SKpl W10/20 - YN35mm F2 - Canon EOS 650 (with flash)
Focus on Stauroneis phoenicenteron:
Last edited by ImperatorRex on Wed Sep 25, 2019 10:47 am, edited 2 times in total.
Re: My Zeiss Microscopes
All being equal and calculated by thumb, is that a 50% difference in magnification between the photo adapters?
Be interesting if you increased the magnification of the UPlanFL40/0.75 PH2 by that amount to compare the objectives.
Be interesting if you increased the magnification of the UPlanFL40/0.75 PH2 by that amount to compare the objectives.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hi 75RR,
think the important difference is the Field of View - the adaption with the Olympus allows to capture the full FOV 22+, while the Zeiss finite has less FOV and crops a rectangle out of the FOW of 20.
Direct image of the intermediate image to the sensor of the camera also mean that is is less sensible to dirt allocated in between.
think the important difference is the Field of View - the adaption with the Olympus allows to capture the full FOV 22+, while the Zeiss finite has less FOV and crops a rectangle out of the FOW of 20.
Direct image of the intermediate image to the sensor of the camera also mean that is is less sensible to dirt allocated in between.
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Some corpped images:
Olympus:
Zeiss:
Obviously the Zeiss has some advantage due to higher n.A. and the water immersion. Also cover glass of the testplate has only thickness of 0.13, so the correction collar of the Zeiss Objective makes some difference to abopt the S.A.
Olympus:
Zeiss:
Obviously the Zeiss has some advantage due to higher n.A. and the water immersion. Also cover glass of the testplate has only thickness of 0.13, so the correction collar of the Zeiss Objective makes some difference to abopt the S.A.
Re: My Zeiss Microscopes
You are quite right of course. That was what you were showing us.ImperatorRex wrote: ↑Wed Sep 25, 2019 10:54 amHi 75RR,
think the important difference is the Field of View - the adaption with the Olympus allows to capture the full FOV 22+, while the Zeiss finite has less FOV and crops a rectangle out of the FOW of 20.
Direct image of the intermediate image to the sensor of the camera also mean that is is less sensible to dirt allocated in between.
Thanks for the comparison (which I have placed side by side and reduced a bit, hope that is alright.) I think the Olympus holds up very well!
Note: So that both images would be bias positive I rotated the Zeiss image 180 - that might account for some perceived differences.
- Attachments
-
- Olympus-Zeiss-2.jpg (103.47 KiB) Viewed 23165 times
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Well done 75RR, much better to compare. Also think the Olympus is doing quite well, also considering it is a phase contrast objective.
Re: My Zeiss Microscopes
Indeed. Phase and 0.75NA, doing very well!ImperatorRex wrote: ↑Wed Sep 25, 2019 8:03 pmWell done 75RR, much better to compare. Also think the Olympus is doing quite well, also considering it is a phase contrast objective.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: My Zeiss Microscopes
Did you use oil or other immersion medium with the Neofluar objective?
You get very nice quality with both objective, it'd be interesting to see some images you take with your new microscope, it looks very promising
You get very nice quality with both objective, it'd be interesting to see some images you take with your new microscope, it looks very promising
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel
Youtube channel
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hi Wes,
the Zeiss Plan Neofluar 40/0,90 is a multi-immersion objective, so the collar allows to use the immersion medium of your choice: oil, water or glycerin. Water is very convenient to use, its very easy to clean. I usually prefer the the dry objective Zeiss Planapo 40/0,95 but it has a bit a gradient in the DIC background, so I was using the Plan Neofluar instead to give the Zeiss a good chance against the Olympus
Yes, will do some first photos. The flash setup at the Olympus is a bit provisionally, but seems to be good enough to start with photos. I have not really good pond samples, so probabely will go for some Haematococcus pluvialis algae...they are quick to re-activate from dried aplanospores. Maybe some direct comparisons then between the DIC from Olympus against the old Zeiss. What I have seen is that the Olympus has a very nice & strong contour contrast, were the Zeiss is much softer. Lets see how this looks on the photos.
the Zeiss Plan Neofluar 40/0,90 is a multi-immersion objective, so the collar allows to use the immersion medium of your choice: oil, water or glycerin. Water is very convenient to use, its very easy to clean. I usually prefer the the dry objective Zeiss Planapo 40/0,95 but it has a bit a gradient in the DIC background, so I was using the Plan Neofluar instead to give the Zeiss a good chance against the Olympus
Yes, will do some first photos. The flash setup at the Olympus is a bit provisionally, but seems to be good enough to start with photos. I have not really good pond samples, so probabely will go for some Haematococcus pluvialis algae...they are quick to re-activate from dried aplanospores. Maybe some direct comparisons then between the DIC from Olympus against the old Zeiss. What I have seen is that the Olympus has a very nice & strong contour contrast, were the Zeiss is much softer. Lets see how this looks on the photos.
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
My Zeiss Microscopes (Comparison photo adaption Zeiss Standard versus Olympus BX)
Hi,
did a comparison of my photo adaptions I use currently:
green rectangle: Olympus BX with Canon EOS600D (the intermediate image is directly projected to the camera sensor)
red rectangle: Zeiss Standard with large field optovar 0,8x - KPL SKPL 10x/20 (Ocular) - YN EF35mm (Camera objective) - Canon EOS650D
blue rectangle: Zeiss Standard - KPL SKPL 10x/20 (Ocular) - YN EF35mm (Camera objective) - Canon EOS650D
The BX photo adaption does even capture a bit more width of the FOV than is visible in the eyepieces (FOV22).
Below the single images taken with the 10x objectives:
Olympus BX UPlanFL 10x/0,30 PH1
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Zeiss Standard PlanApo 10/0,32 (with Optovar 1,0)
Some other photos
Pleurosigma angulatum (in air):
Olympus BX UPlanFL 10x/0,30 PH1:
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Olympus BX UPlanFL 40x/0,75 PH2
Zeiss Standard PlanApo 40/0,95 (with Optovar 0,80)
Zeiss Standard PlanApo 40/0,95 (with Optovar 1,0)
Olympus BX UPlanFL 10x/0,30 PH1:
(Scyllium young Shark)
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Zeiss Standard PlanApo 10/0,32 (with Optovar 1,0)
did a comparison of my photo adaptions I use currently:
green rectangle: Olympus BX with Canon EOS600D (the intermediate image is directly projected to the camera sensor)
red rectangle: Zeiss Standard with large field optovar 0,8x - KPL SKPL 10x/20 (Ocular) - YN EF35mm (Camera objective) - Canon EOS650D
blue rectangle: Zeiss Standard - KPL SKPL 10x/20 (Ocular) - YN EF35mm (Camera objective) - Canon EOS650D
The BX photo adaption does even capture a bit more width of the FOV than is visible in the eyepieces (FOV22).
Below the single images taken with the 10x objectives:
Olympus BX UPlanFL 10x/0,30 PH1
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Zeiss Standard PlanApo 10/0,32 (with Optovar 1,0)
Some other photos
Pleurosigma angulatum (in air):
Olympus BX UPlanFL 10x/0,30 PH1:
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Olympus BX UPlanFL 40x/0,75 PH2
Zeiss Standard PlanApo 40/0,95 (with Optovar 0,80)
Zeiss Standard PlanApo 40/0,95 (with Optovar 1,0)
Olympus BX UPlanFL 10x/0,30 PH1:
(Scyllium young Shark)
Zeiss Standard PlanApo 10/0,32 (with Optovar 0,80)
Zeiss Standard PlanApo 10/0,32 (with Optovar 1,0)
Re: My Zeiss Microscopes
Interesting comparison. What I personally take from it:
1. The S-KPL is a very good eyepiece for photography (as previously shown by others) - no wonder it is rare and very expensive
2. In spite of the 20-30 years age difference and the infinity optics of the BX, and of course the huge price difference, the Zeiss Standard yields about the same excellent performance as the BX (for brightfield !).
And a question please. Why a 35mm camera lens and not a 40mm lens ? considering the 10X eyepiece and APS-C sensor ?
1. The S-KPL is a very good eyepiece for photography (as previously shown by others) - no wonder it is rare and very expensive
2. In spite of the 20-30 years age difference and the infinity optics of the BX, and of course the huge price difference, the Zeiss Standard yields about the same excellent performance as the BX (for brightfield !).
And a question please. Why a 35mm camera lens and not a 40mm lens ? considering the 10X eyepiece and APS-C sensor ?
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hi Hobbyst46,
definitely the Zeiss Standard still performes very well, respectively if equipped with Planapos - and yes it will cost only a fraction of the invinite systems.
From my experience I do not think that the SKPL 10x/20 makes a much better performance compared to a KPL 10x/20. So it is a nice to have but not really a must.
A cheap Photoadaption with good results is also a KPL 8x/18 with a 50mm Camera objective. Of coarse a slightly smaller FOV.
definitely the Zeiss Standard still performes very well, respectively if equipped with Planapos - and yes it will cost only a fraction of the invinite systems.
From my experience I do not think that the SKPL 10x/20 makes a much better performance compared to a KPL 10x/20. So it is a nice to have but not really a must.
A cheap Photoadaption with good results is also a KPL 8x/18 with a 50mm Camera objective. Of coarse a slightly smaller FOV.
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Olympus BX Photo Adaption
Would like to share some details for a very compact and simple photo adaption for the Olympus BX:
- Photo adapter length is just 60mm / 2,4 inches (distance from the trino flange to the Canon EOS Camera flange)
- The intermediate picture of the microscope is directly projected on the sensor of the Camera. No optic like photo ocular PE or camera objective is required
- The photo image is even a bit larger than the image visible in the eyepieces
Re: My Zeiss Microscopes
Ideal solution for an infinity system.
Made to order phototube I suppose, is it adjustable?
What is the camera sensor size?
Made to order phototube I suppose, is it adjustable?
What is the camera sensor size?
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hi 75RR,
you are right, the tube is not commerically availble, but there is a guy in the Austrian micro forum who fabricates such tubes. Anyway it is not rocket-science, one side connects to the tube head an the other side has the thread for the T2 adapter for Canon bajonet. The phototube itself is not adjustable, the parfocal setup is made by adjustable eyepieces. This works quite well, so I do not depend on life-view of the camera.
you are right, the tube is not commerically availble, but there is a guy in the Austrian micro forum who fabricates such tubes. Anyway it is not rocket-science, one side connects to the tube head an the other side has the thread for the T2 adapter for Canon bajonet. The phototube itself is not adjustable, the parfocal setup is made by adjustable eyepieces. This works quite well, so I do not depend on life-view of the camera.
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Zeiss IM35 - Adapter for Petri dishes
There are various different table inserts available for the vertical Zeiss IM35. Recently I had the chance to get another table insert, it is another variant of a holder for petri-dishes, the Zeiss item number is 47 17 38. The holder is adjustable, the claws of the holder allows the use of different Petri dish diameters.
I find it very handy because it is also possible to fit other slides, cover glasses and other carriers that I have in use.
Below some fotos that I would like to share:
I find it very handy because it is also possible to fit other slides, cover glasses and other carriers that I have in use.
Below some fotos that I would like to share:
Re: My Zeiss Microscopes
Nice design features in that table insert - well made, well thought out.
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Olympus BX - DIY - LED retrofit with integrated flash
Simple conversion of the 100W BX Lamp housing to a LED solution with option to fit a flash bulb. The BX lamp housing can be easily converted back into its original status if such is required. Similiar solution may be applicable also for other microscopes?
Reasons for the conversion:
So far I used a beam splitter (50/50) for the flash. Howerever with DIK and objectives 60x and 100x (ISO 200 at the camera) the flash power was reaching the limit.
A simple construction - but I have to admit it may look a bit strange:
Costs are rather limited:
- Alu disc („ALU Ronde Ø60x1,5mm“ - 2,60 Euro).
- 3 metric screws M3x40mm with 6 washers.
- a bit cable to extent the flash bulb.
- 1x LED CREE XM-L2 U4 + Heattransfer and gluing pad (16,50 Euro).
- Flash - very cheap chinese clones for Canon flushs... etc.
Reasons for the conversion:
So far I used a beam splitter (50/50) for the flash. Howerever with DIK and objectives 60x and 100x (ISO 200 at the camera) the flash power was reaching the limit.
A simple construction - but I have to admit it may look a bit strange:
Costs are rather limited:
- Alu disc („ALU Ronde Ø60x1,5mm“ - 2,60 Euro).
- 3 metric screws M3x40mm with 6 washers.
- a bit cable to extent the flash bulb.
- 1x LED CREE XM-L2 U4 + Heattransfer and gluing pad (16,50 Euro).
- Flash - very cheap chinese clones for Canon flushs... etc.
Last edited by ImperatorRex on Fri May 22, 2020 1:56 pm, edited 1 time in total.
Re: My Zeiss Microscopes
Nice!
Is the LED sitting directly behind or just a bit above/below the flash tube?
Is the LED sitting directly behind or just a bit above/below the flash tube?
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Olympus BX - DIY - LED retrofit with integrated flash
The flash tube is as close as possible, just a little bit above the LED.
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Olympus BX - DIY - LED retrofit with integrated flash
Flash is working very well
Something you may consider if you apply similiar solution:
I have been told that xenon tubes emit quite some portion of UV light:
http://www.allaboutsymbian.com/features ... mtech3.jpg
I had some doubt if the short flash light duration may ever cause some issue? But seems such UV rays can even damage pol filters. Note that it was not a HBO lamp that damaged the pol-filter, damage was caused by a xenon flash: https://www.mikroskopie-forum.de/index. ... #msg201628
Therefore I now consider to mount a UV filter. Will install after the lamp collector where previously the heat filter glas was mounted (dia. 44mm).
Something you may consider if you apply similiar solution:
I have been told that xenon tubes emit quite some portion of UV light:
http://www.allaboutsymbian.com/features ... mtech3.jpg
I had some doubt if the short flash light duration may ever cause some issue? But seems such UV rays can even damage pol filters. Note that it was not a HBO lamp that damaged the pol-filter, damage was caused by a xenon flash: https://www.mikroskopie-forum.de/index. ... #msg201628
Therefore I now consider to mount a UV filter. Will install after the lamp collector where previously the heat filter glas was mounted (dia. 44mm).
Re: Olympus BX - DIY - LED retrofit with integrated flash
Gives new meaning to the expression "blinded by the flash"ImperatorRex wrote: ↑Fri May 22, 2020 1:55 pmI had some doubt if the short flash light duration may ever cause some issue? But seems such UV rays can even damage pol filters. Note that it was not a HBO lamp that damaged the pol-filter, damage was caused by a xenon flash: https://www.mikroskopie-forum.de/index. ... #msg201628
Zeiss Standard WL (somewhat fashion challenged) & Wild M8
Olympus E-P2 (Micro Four Thirds Camera)
Olympus E-P2 (Micro Four Thirds Camera)
Re: Olympus BX - DIY - LED retrofit with integrated flash
A very good safety measure. Xenon lamps do emit UV. Not only in microscope illuminators but in flash units as well.ImperatorRex wrote: ↑Fri May 22, 2020 1:55 pmFlash is working very well
Something you may consider if you apply similiar solution:
I have been told that xenon tubes emit quite some portion of UV light:
http://www.allaboutsymbian.com/features ... mtech3.jpg
I had some doubt if the short flash light duration may ever cause some issue? But seems such UV rays can even damage pol filters. Note that it was not a HBO lamp that damaged the pol-filter, damage was caused by a xenon flash: https://www.mikroskopie-forum.de/index. ... #msg201628
Therefore I now consider to mount a UV filter. Will install after the lamp collector where previously the heat filter glas was mounted (dia. 44mm).
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Re: My Zeiss Microscopes
Hobbyst46/75RR,
right...Polfilter may be replaced, but the eyelight not. Safety first!
right...Polfilter may be replaced, but the eyelight not. Safety first!
- ImperatorRex
- Posts: 571
- Joined: Fri Mar 30, 2018 4:12 pm
- Location: Germany
- Contact:
Olympus BX - DIY LED replacement for the HBO 100W high pressure Mercury burner
Hi,
once more another simple DYI LED solution for the Olympus BX. This time it is intended to replace the HBO high pressure Hg 100W burner used for incident light fluorescence. Maybe you can apply a similiar solution also on other microscope stands?
My plan was to fit an UV LED toghether with a normal CREE LED, so no disassemble required to switch between UV or VIS LED.
Some photos first that explain the principal:
The LED lighting has a blue plastic collar to protect from stray light:
Two seperate connectors, one plug for the UV LED power supply and the other one for the "normal" LED
The UV LED is right in the center of the Alu-Disc:
...a small shift of the cover centers the other LED into center of the collector optic:
First testing and results:
UV-excitation filter:
Blue light excitation filter:
Green light excitation filter:
First I had a different plan: I installed the UV LED exactly in the center focal point of the collector lense. The other LED has some little offset. I just cutted off some off the LEDs cupper board since intention was to keep offset as small as possible. My expectation was that it will still provide sufficient beam power even if it remains de-centered.
However the result was disappointing, the beam power of the out of center LED was very low, not usable to achive blue or green fluoreszenz excitation! I should have tried before cutting the LEDs!
But there is always a plan "B"
Remove two of the stud screws to make the cover disc moveable - so it is not a big thing to quickly turn the disc a bit to center the other LED. Actually it is rather convinient to do, so finally mission acomplished
Ende gut, alles gut
FYI: I have installed the following, rather cheap LEDs:
- SSC VIOSYS UV 365nm CUN66A1B
- CREE XM-L2 U4
once more another simple DYI LED solution for the Olympus BX. This time it is intended to replace the HBO high pressure Hg 100W burner used for incident light fluorescence. Maybe you can apply a similiar solution also on other microscope stands?
My plan was to fit an UV LED toghether with a normal CREE LED, so no disassemble required to switch between UV or VIS LED.
Some photos first that explain the principal:
The LED lighting has a blue plastic collar to protect from stray light:
Two seperate connectors, one plug for the UV LED power supply and the other one for the "normal" LED
The UV LED is right in the center of the Alu-Disc:
...a small shift of the cover centers the other LED into center of the collector optic:
First testing and results:
UV-excitation filter:
Blue light excitation filter:
Green light excitation filter:
First I had a different plan: I installed the UV LED exactly in the center focal point of the collector lense. The other LED has some little offset. I just cutted off some off the LEDs cupper board since intention was to keep offset as small as possible. My expectation was that it will still provide sufficient beam power even if it remains de-centered.
However the result was disappointing, the beam power of the out of center LED was very low, not usable to achive blue or green fluoreszenz excitation! I should have tried before cutting the LEDs!
But there is always a plan "B"
Remove two of the stud screws to make the cover disc moveable - so it is not a big thing to quickly turn the disc a bit to center the other LED. Actually it is rather convinient to do, so finally mission acomplished
Ende gut, alles gut
FYI: I have installed the following, rather cheap LEDs:
- SSC VIOSYS UV 365nm CUN66A1B
- CREE XM-L2 U4
Last edited by ImperatorRex on Sat May 23, 2020 5:52 pm, edited 1 time in total.