The Magnification Issue

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Post Reply
Message
Author
DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

The Magnification Issue

#1 Post by DonSchaeffer » Sun Jul 12, 2020 6:19 pm

I know this is an absurd kind of question. We don't care about "power" calculations. I'm just curious though. Conventionally, if i use a 10x eyepiece with my 40x objective we say the microscope is 400x. I have an eyepiece camera that includes a .5x reduction lens. What would be the power equivalent.

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: The Magnification Issue

#2 Post by Scarodactyl » Sun Jul 12, 2020 6:38 pm

Optical magnification would be 20x, assuming the 0.5x is actually doing a 0.5x reduction. A generic 0.5x will not actually give exactly that on every system.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#3 Post by DonSchaeffer » Sun Jul 12, 2020 9:38 pm

You are saying that the camera is equivalent to a 20x eyepiece?

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: The Magnification Issue

#4 Post by Scarodactyl » Sun Jul 12, 2020 10:02 pm

No, what I was saying is that 40 x 0.5 = 20. The camera has nothing to do with what the optical magnification is. This is why optical magnification is not useful for comparing images taken across different systems, while field of view is.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#5 Post by DonSchaeffer » Sun Jul 12, 2020 10:20 pm

Ok I wont ask it any more. Lol.

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: The Magnification Issue

#6 Post by Scarodactyl » Mon Jul 13, 2020 12:38 am

It's not a bad question to ask, it just doesn't end up leading to the answer you really want, since we apply our own magnification to the final image when we view it on the screen. Field of view deals with all the issues of cropping and resizing by skipping it and looking at the final result.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#7 Post by DonSchaeffer » Mon Jul 13, 2020 2:04 pm

Ok. What I want to know is what eyepiece the camera is roughly equivalent to. I know it crops less than my old Omax camera.

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: The Magnification Issue

#8 Post by MichaelG. » Mon Jul 13, 2020 4:17 pm

DonSchaeffer wrote:
Mon Jul 13, 2020 2:04 pm
Ok. What I want to know is what eyepiece the camera is roughly equivalent to. I know it crops less than my old Omax camera.
You need to include the sensor sizes [physical dimensions, not pixel count] in your calculation to do that, Don.

MichaelG.
Too many 'projects'

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#9 Post by DonSchaeffer » Mon Jul 13, 2020 4:51 pm

ok. Thanks.

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: The Magnification Issue

#10 Post by Scarodactyl » Mon Jul 13, 2020 7:46 pm

I'd suggest you skip the calculations and measure directly. Put a ruler in focus, take a look at it through your 10x eyepieces and through your camera to directly measure the field of view in each and compare.

hans
Posts: 1006
Joined: Thu May 28, 2020 11:10 pm
Location: Southern California

Re: The Magnification Issue

#11 Post by hans » Mon Jul 13, 2020 7:54 pm

If the specs at http://www.aibome.com/product/36_146.html as posted in your other thread are correct then you can estimate:

According to Wikipedia a 1/2.5" sensor is 5.76 mm wide. Assume the objective and 10x eyepiece are normally working with a 20 mm diameter intermediate image. The 0.5x relay should project that to a 10 mm diameter image in the plane of the sensor, of which the sensor sees only 5.76 mm. So the horizontal field of view you see with the camera should be approximately half (5.76/10) of what you would see with a 10x eyepiece.

I think it is not necessarily straightforward to say what eyepiece magnification is equivalent since, from what I understand, eyepieces above 10x typically crop the intermediate image also. If you assume that an inexpensive 20x eyepiece maintains roughly the same apparent field of view then it would be cropping by a factor of 2 relative to the 10x and give a field of view of your specimen that is similar to your camera.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#12 Post by DonSchaeffer » Mon Jul 13, 2020 8:02 pm

Scarodactyl wrote:
Mon Jul 13, 2020 7:46 pm
I'd suggest you skip the calculations and measure directly. Put a ruler in focus, take a look at it through your 10x eyepieces and through your camera to directly measure the field of view in each and compare.
Yes that is probably the best. I did it with my other camera (you may recall). The camera did cut off a fair bit of the field. I think this one cuts off less. I will try again.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#13 Post by DonSchaeffer » Mon Jul 13, 2020 8:20 pm

Forgive me: at 80 years old my arithmetic (or algebra) is very rusty. Using my 60x objective .01 mm looks like 3.8 cm or 38 mm. What would be the magnification factor then? Is it only 380 power?

Lifephoto
Posts: 8
Joined: Sat Jul 11, 2020 6:09 am

Re: The Magnification Issue

#14 Post by Lifephoto » Tue Jul 14, 2020 1:31 am

Hello, there are many interesting issues in this thread, here are a few - - -

One is: If .01mm in a "ruler" becomes 38mm, then the following is true, 0.01 x =38, so x equals 3,800 as the magnification. Any known dimension such as 0.01mm or even 1mm measured and compared as above can yield a final magnification.

A second issue is where does a "calculated" magnification occur? If the objective is 40x, a 0.5x intermediate lens is present, and the ocular is 10x , this yields a calculated magnification of 200x. As I understand it, that 200x magnification occurs at a distance of 10 inches from the eye point of the ocular. If the projection distance is less than 10 inches, then the fraction of 10 inches present as the projection then needs to be also multiplied. So if a distance of 5 inches projection is used, the magnification at that point would be 100x. Since there are +/- errors in all of the above, such a calculated magnification is very much an approximation.

Back to the first example above, I photograph a known dimension ruled slide or ruler with a given setup along with the images of my subjects. That known dimension can then be carefully remeasured in the final output: such as a photographic print, on your computer monitor, or even on a projection screen to let one calculate the final magnification you have produced. This is very simple and direct.

Using appropriate proportions one can use measurements like the above to accurately calculate the actual size for objects photographed under a microscope. I do not know how to produce equations in this application, otherwise I would create a fully detailed example.

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#15 Post by DonSchaeffer » Tue Jul 14, 2020 2:42 am

I knew it. I got the whole thing over-simplifed and calculated wrong. Forgive me. I was going by an image on my monitor and I multiplied wrong. I'm staying out of this calculation business.

Scarodactyl
Posts: 2790
Joined: Sat Mar 03, 2018 9:09 pm

Re: The Magnification Issue

#16 Post by Scarodactyl » Tue Jul 14, 2020 5:17 am

Lifephoto wrote:
Tue Jul 14, 2020 1:31 am
A second issue is where does a "calculated" magnification occur? If the objective is 40x, a 0.5x intermediate lens is present, and the ocular is 10x , this yields a calculated magnification of 200x. As I understand it, that 200x magnification occurs at a distance of 10 inches from the eye point of the ocular. If the projection distance is less than 10 inches, then the fraction of 10 inches present as the projection then needs to be also multiplied. So if a distance of 5 inches projection is used, the magnification at that point would be 100x. Since there are +/- errors in all of the above, such a calculated magnification is very much an approximation.
There are some misconceptions here.
When Don is talking about a 0.5x lens, it is part of a camera mount. No 10x eyepiece is involved there.
A viewing ocular typically projects the image to infinity (which your eye lens then refocuses onto your retina which is a whole other can of worms in making calculations). If you have a high enough eyepoint eyepiece and move your eye towards and away from the eyepiece the image does not change in magnification.
DonSchaeffer wrote:
Mon Jul 13, 2020 8:20 pm
Forgive me: at 80 years old my arithmetic (or algebra) is very rusty. Using my 60x objective .01 mm looks like 3.8 cm or 38 mm. What would be the magnification factor then? Is it only 380 power?
I think you're making this harder on yourself than it needs to be. I don't know what it would mean for it to look like 3.8cm, don't worry about that. All you need to do is see how much of the ruler is in the field of view:
Image
(It should be a LOT less than 62mm in real life!)
Then take a picture of the ruler with your camera and compare the width.

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: The Magnification Issue

#17 Post by MichaelG. » Tue Jul 14, 2020 6:58 am

DonSchaeffer wrote:
Mon Jul 13, 2020 8:20 pm
Forgive me: at 80 years old my arithmetic (or algebra) is very rusty. Using my 60x objective .01 mm looks like 3.8 cm or 38 mm. What would be the magnification factor then? Is it only 380 power?
DonSchaeffer wrote:
Tue Jul 14, 2020 2:42 am
I knew it. I got the whole thing over-simplifed and calculated wrong. Forgive me. I was going by an image on my monitor and I multiplied wrong. I'm staying out of this calculation business.


38 ÷ 0.01 = 3800

Note that this is the overall magnification, on your screen
... which is a concept that you disparaged in a previous discussion

Ref. viewtopic.php?f=24&t=9109

MichaelG.
Too many 'projects'

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#18 Post by DonSchaeffer » Tue Jul 14, 2020 10:24 am

I know I'm making too much of a big deal about this. I guess it started with the notion that microscopes are advertised as 1,000x. They arrive at that number by simply multiplying objective times eyepiece. What an lovely formula that is? So my mind says, so why can't I multiply objective times my camera. Then I'm faced with, "ok what is my camera number?" It seems a simple enough notion. How do you arrive at it? It's certainly not worth the fuss I'm making. Even an estimate based on how it looks would make me happy.

viktor j nilsson
Posts: 761
Joined: Thu Aug 23, 2018 10:12 pm
Location: Lund, Sweden

Re: The Magnification Issue

#19 Post by viktor j nilsson » Tue Jul 14, 2020 10:47 am

It is very straightforward. How large is your subject? Let's say it's 1mm. How large is your subject when projected on your sensor? To find out, you take a photograph of the subject. You then bring this photo into a photo editing software with some sort of measuring capabilities. I use imageJ for this.

You need to know how large your sensor is (in mm), and how many pixels there are along one of its sides.

You divide the width of your sensor (in mm) by the number of pixels along it's wide side. This gives you a number which represents the width (in mm) of each pixel.

You then measure how many pixels your subject covers on the image.

You then multiply the width of your subject (in pixels) with the pixel width you calculated earlier.

This will tell you how large the subject is when projected on the sensor.

If you divide the width of your subject (on the sensor) by it's size (in real life), you get the magnification. If it was 10mm on sensor and 1mm in real life, your optical magnification is 10x.

So magnification on sensor is very straightforward. The way eyepiece magnification is calculated is actually more abstract.

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: The Magnification Issue

#20 Post by MichaelG. » Tue Jul 14, 2020 1:11 pm

DonSchaeffer wrote:
Tue Jul 14, 2020 10:24 am
[…]
Then I'm faced with, "ok what is my camera number?"
[…]
Even an estimate based on how it looks would make me happy.
.
You already have that, Don
... You did it with the check on the 60x objective

3800 ÷ 60 = 63.33333333333333 [ish]

MichaelG.
Too many 'projects'

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#21 Post by DonSchaeffer » Tue Jul 14, 2020 4:11 pm

ok I'm a happy camper. It's amazing how little I care about validity.

MichaelG.
Posts: 4026
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: The Magnification Issue

#22 Post by MichaelG. » Tue Jul 14, 2020 4:37 pm

DonSchaeffer wrote:
Tue Jul 14, 2020 4:11 pm
It's amazing how little I care about validity.
If I knew what you meant by that, I might know whether to be pleased or offended.

MichaelG.
Too many 'projects'

Zuul
Posts: 212
Joined: Fri May 01, 2020 9:01 pm
Location: California

Re: The Magnification Issue

#23 Post by Zuul » Tue Jul 14, 2020 6:43 pm

Maybe he was hoping somebody would say “42” ?? :D

User avatar
janvangastel
Posts: 533
Joined: Mon Dec 03, 2018 7:05 pm
Location: Huizen, Netherlands
Contact:

Re: The Magnification Issue

#24 Post by janvangastel » Wed Jul 15, 2020 5:20 am

I measured the FOV of a video and photograph by photographing/filming a grating having 500 lines/millimeter. For instance, with my 63x objective, the width of the FOV had 102 lines and the height 56 lines, which calculates to a FOV of 102/500 and 56/500, or 0.2x0.11 millimeter. If the size of an object is known, it is easy now to calculate the total magnification by just measuring the object size on the photograph relative to the FOV size (length of width) and divide by 0.2 or 0.11. And it's also easy to calulate the real size of an object if unknown,

DonSchaeffer
Posts: 3355
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: The Magnification Issue

#25 Post by DonSchaeffer » Wed Jul 15, 2020 11:43 am

No reflection on you guys. Thanks for this discussion. It's my muzzy mind. I just want a talking point.

Post Reply