#1 Post by MicroMan2 » Sun Nov 02, 2014 4:12 am

So I was trying to learn the physics of oblique illumintation so what is this illuminating wavelength and how do I find its value

User avatar
Oliver
Posts: 187
Joined: Sat Oct 11, 2014 12:57 pm
Location: Austria
Contact:

Re: What is the Illuminating Wavelength

#2 Post by Oliver » Sun Nov 02, 2014 7:13 am

Hello,
The illumination wavelength is the wavelwngth of the light used to illuminate the specimen. It is the wavelength of the microscope lamp. You need to know this wavelength in order to calculate the theoretically maximum resolution. The shorter the wavelength, the higher the resolution. Now, there is no single wavelength, however. The visual spectrum goes from about 450nm to 700nm. For doing calculations, the middle wavelength (around 550nm) is used. Certain light sources, such as LEDs, have a more narrow wavelength spectrum. You have to check the datasheet of the LED to figure out the stongest wavelength. Other lamps (halogen) will shift the wavelength based on intensity. So this is often of theoretical value only.
Oliver.
Image Oliver Kim - http://www.microbehunter.com - Microscopes: Olympus CH40 - Olympus CH-A - Breukhoven BMS student microscope - Euromex stereo - uSCOPE MXII

MicroMan2
Posts: 159
Joined: Tue Oct 14, 2014 4:20 pm
Location: Canada,Alberta

Re: What is the Illuminating Wavelength

#3 Post by MicroMan2 » Sun Nov 02, 2014 3:09 pm

Thanks now i did my calculations and
2(0.65)
---------
(600nm)
Equals to 0.00216666666
I know I'm wrong because the resolution limit is 0.250
So any help.
And thanks for replying

User avatar
Oliver
Posts: 187
Joined: Sat Oct 11, 2014 12:57 pm
Location: Austria
Contact:

Re: What is the Illuminating Wavelength

#4 Post by Oliver » Sun Nov 02, 2014 4:12 pm

You got the fraction the other way around.

Resolution = L / (2*NA)
= 600nm / (2*0.65) = 0.46micrometers

compare this to the values in Table 2, here:
http://zeiss-campus.magnet.fsu.edu/arti ... ution.html

Oliver.