Converting Brightfield Microscope To DIC

#1 Post by MicroscopyLearning » Sun Jun 27, 2021 6:31 pm

Hello. Could someone explain what components are needed to convert a brightfield microscope into a DIC microscope. I have found this article: scientifica.uk.com/learning-zone/differential-interference-contrast, but I did not understand what it meant by an objective specific prism. Is this an additional prism one must add, or is it talking about a prism a brightfield microscope would already have. The microscope I am looking to convert is: Swift SW350T. Could the analyzer be put in the place below the head?

apochronaut
Posts: 6309
Joined: Fri May 15, 2015 12:15 am

Re: Converting Brightfield Microscope To DIC

#2 Post by apochronaut » Sun Jun 27, 2021 6:48 pm

The extra components and critical alignments necessary to convert a microscope like yours to DIC are roughly equivalent to converting a basic television set into a 3-D home theater.

User avatar
Wes
Posts: 1027
Joined: Sat Mar 09, 2019 12:58 pm

Re: Converting Brightfield Microscope To DIC

#3 Post by Wes » Sun Jun 27, 2021 8:17 pm

apochronaut wrote:
Sun Jun 27, 2021 6:48 pm
The extra components and critical alignments necessary to convert a microscope like yours to DIC are roughly equivalent to converting a basic television set into a 3-D home theater.
I wonder if the promiscuous PZO DIC kit can fit his scope.
Zeiss Photomicroscope III BF/DF/Pol/Ph/DIC/FL/Jamin-Lebedeff
Youtube channel

apochronaut
Posts: 6309
Joined: Fri May 15, 2015 12:15 am

Re: Converting Brightfield Microscope To DIC

#4 Post by apochronaut » Sun Jun 27, 2021 9:04 pm

Putting the PZO interference head on it would be like trying to make a houseboat out of a disused hockey arena. I think it outweighs the microscope.
There is no condenser carrier, the illumination is a 1 watt led, the objectives are not strain free, the PZO dovetail is 43mm. PZO interference kits are north of $1000.00
these days.

Greg Howald
Posts: 1186
Joined: Tue Oct 20, 2020 6:44 am

Re: Converting Brightfield Microscope To DIC

#5 Post by Greg Howald » Mon Jun 28, 2021 12:27 am

I'm pretty sure you are stuck with "poor man's DIC".
You can do some amazing things with oblique filtering which will produce wonderful results even when using phase contrast, but without the items necessary for actual DIC You can not achieve it. DIC is costly. If you try to do it by buying parts to build it, DIC can cost you more in parts than the new scope.
I suppose you could achieve it with lots of patience and lots of time. I tried for a while.
Greg

BramHuntingNematodes
Posts: 1546
Joined: Tue Jan 21, 2020 1:29 am
Location: Georgia, USA

Re: Converting Brightfield Microscope To DIC

#6 Post by BramHuntingNematodes » Mon Jun 28, 2021 2:12 am

Yeah if I ever luck into some nice iceland spar at a swap meet and manage to cobble together a CNC faceting machine I got the barnesite all hooked up and ready to try out I don't know I guess the dozen or so configurations we would need to hone in on the right prism specifications ought to be p. straightforward
1942 Bausch and Lomb Series T Dynoptic, Custom Illumination

synnus
Posts: 72
Joined: Tue Feb 23, 2021 7:46 pm

Re: Converting Brightfield Microscope To DIC

#7 Post by synnus » Tue Jun 29, 2021 3:51 pm

MicroscopyLearning wrote:
Sun Jun 27, 2021 6:31 pm
Hello. Could someone explain what components are needed to convert a brightfield microscope into a DIC microscope. I have found this article: scientifica.uk.com/learning-zone/differential-interference-contrast, but I did not understand what it meant by an objective specific prism. Is this an additional prism one must add, or is it talking about a prism a brightfield microscope would already have. The microscope I am looking to convert is: Swift SW350T. Could the analyzer be put in the place below the head?
I have a PZO mpi5 for sell. If you are interested. please inform me.

MicroscopyLearning
Posts: 64
Joined: Wed Feb 10, 2021 2:50 pm

Re: Converting Brightfield Microscope To DIC

#8 Post by MicroscopyLearning » Thu Jul 01, 2021 5:02 am

Wes wrote:
Sun Jun 27, 2021 8:17 pm
apochronaut wrote:
Sun Jun 27, 2021 6:48 pm
The extra components and critical alignments necessary to convert a microscope like yours to DIC are roughly equivalent to converting a basic television set into a 3-D home theater.
I wonder if the promiscuous PZO DIC kit can fit his scope.
What does this kit contain? I have not been able to find much about it.

microb
Posts: 729
Joined: Wed Mar 27, 2019 6:39 am

Re: Converting Brightfield Microscope To DIC

#9 Post by microb » Thu Jul 01, 2021 7:23 pm

MicroscopyLearning wrote:
Sun Jun 27, 2021 6:31 pm
Swift SW350T
Can the condenser be removed?

Is there a phase contrast option that can be bought for the Swift SW350T?

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#10 Post by DonSchaeffer » Thu Jul 01, 2021 7:51 pm

Don't laugh but I think you can get pretty close to a DIC look by closing your condensor iris down in bright field, adjusting the contrast, and focusing carefully.

MicroBob
Posts: 3154
Joined: Sun Dec 25, 2016 9:11 am
Location: Northern Germany

Re: Converting Brightfield Microscope To DIC

#11 Post by MicroBob » Fri Jul 02, 2021 6:44 am

Hi Don,
can you post an image where your method has actually worked?

Bob

MicroscopyLearning
Posts: 64
Joined: Wed Feb 10, 2021 2:50 pm

Re: Converting Brightfield Microscope To DIC

#12 Post by MicroscopyLearning » Tue Jul 13, 2021 3:22 am

microb wrote:
Thu Jul 01, 2021 7:23 pm
MicroscopyLearning wrote:
Sun Jun 27, 2021 6:31 pm
Swift SW350T
Can the condenser be removed?

Is there a phase contrast option that can be bought for the Swift SW350T?
It appears that it can. It has screws that look like they can be taken out. I do not think there is an option for phase contrast on this microscope, but I will look.

MichaelG.
Posts: 4019
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Converting Brightfield Microscope To DIC

#13 Post by MichaelG. » Tue Jul 13, 2021 5:40 am

DonSchaeffer wrote:
Thu Jul 01, 2021 7:51 pm
Don't laugh but I think you can get pretty close to a DIC look by closing your condensor iris down in bright field, adjusting the contrast, and focusing carefully.
I think you will find that approach creates diffraction artefacts, Don
… The ‘look’ may be similar, and quite pleasing, but much of the ‘information’ is fake.

MichaelG.
Too many 'projects'

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#14 Post by DonSchaeffer » Tue Jul 13, 2021 2:41 pm

I alsays thought microscope optic relies on diffraction patterns. It's basically all fake.

BramHuntingNematodes
Posts: 1546
Joined: Tue Jan 21, 2020 1:29 am
Location: Georgia, USA

Re: Converting Brightfield Microscope To DIC

#15 Post by BramHuntingNematodes » Tue Jul 13, 2021 4:32 pm

But we do want our image to have some sort of correspondence with the observed object in a reliable, consistent fashion. Gray gives a good illustration of the spurious artifacts in his Handbook, showing that non-existent holes can be made to appear in torn foil with a sufficiently tamped-down NA.

The appearance of relief in DIC is also false, but It's false in a different way than oblique or field stop narrowing. There then is at least one difference in the techniques.
1942 Bausch and Lomb Series T Dynoptic, Custom Illumination

farnsy
Posts: 209
Joined: Mon Jun 17, 2019 9:03 pm
Location: Fort Worth, Texas

Re: Converting Brightfield Microscope To DIC

#16 Post by farnsy » Wed Jul 14, 2021 12:20 am

The components in a DIC system, from the bottom (light source) to the top are:

1. A linear polarizer (easy to get)
2. A lower Nomarski prism (has to be the right one for your objective)
3. A fluorite or apochromatic objective (Swift does not sell any of these but you may be able to find them on old microscopes)
4. An upper Nomarski prism (also specific to your objective)
5. An analyzer (that is, another linear polarizer)

To address the original question, the Nomarski prisms are not anything a regular microscope comes with. They are specialized equipment designed for DIC . On my scope, I have a lower prism for 10X, one that works for 20X and 40X, and one that works for 100X. One upper prism (slider) works for 10X-20X, another for 40X-100X. If you buy them direct from the Chinese factory, the kit costs just over $3,000 before shipping and tariffs with a 6 week lead time or so. I'm pretty sure this is rock bottom pricing for this equipment (new). That's why it probably doesn't make much sense to try and adapt it for a $250 microscope.

The lower Normarski prisms can likely be placed in a condenser designed for phase contrast (mine is). The upper prisms and analyzer can be in sliders. Of course, your scope has to have a slider slot and a slider that can accommodate what you need to put in it.

In general, it's not real easy to add DIC to a system that wasn't designed for it.

viktor j nilsson
Posts: 761
Joined: Thu Aug 23, 2018 10:12 pm
Location: Lund, Sweden

Re: Converting Brightfield Microscope To DIC

#17 Post by viktor j nilsson » Wed Jul 14, 2021 6:26 am

You don't need fluorite or apochromatic objectives for DIC. For best performance, you want the objective the system was designed for. In many cases that will be the better corrected objectives, but sometimes it is not. The BH2 era olympus DIC systems, for example, was designed for the SPlan Achromats. But the more highly corrected objectives also work quite well.

My older Olympus Vanox AH DIC system was designed to use the short-barrel plan Achromats, and don't work as well with the contemporary PlanApo's. So it's not a guarantee that Fluorites and PlanApo's give better DIC, even from the same maker and era.

farnsy
Posts: 209
Joined: Mon Jun 17, 2019 9:03 pm
Location: Fort Worth, Texas

Re: Converting Brightfield Microscope To DIC

#18 Post by farnsy » Wed Jul 14, 2021 6:36 am

viktor j nilsson wrote:
Wed Jul 14, 2021 6:26 am
You don't need fluorite or apochromatic objectives for DIC.
That is interesting; not what I had heard. The folks who sold me my system said you need fluor or better and the only test I've been able to perform indicates that my 60X achromat doesn't work with my 40X/100X DIC components. There could be other reasons for that, though. I guess I should never say never.

viktor j nilsson
Posts: 761
Joined: Thu Aug 23, 2018 10:12 pm
Location: Lund, Sweden

Re: Converting Brightfield Microscope To DIC

#19 Post by viktor j nilsson » Wed Jul 14, 2021 7:00 am

farnsy wrote:
Wed Jul 14, 2021 6:36 am
viktor j nilsson wrote:
Wed Jul 14, 2021 6:26 am
You don't need fluorite or apochromatic objectives for DIC.
That is interesting; not what I had heard. The folks who sold me my system said you need fluor or better and the only test I've been able to perform indicates that my 60X achromat doesn't work with my 40X/100X DIC components. There could be other reasons for that, though. I guess I should never say never.
The most likely explanation is that the prisms were designed for those objectives. Like I said, you want the objectives that the system was designed for. For modern systems, it is typically the better ones.

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#20 Post by DonSchaeffer » Thu Jul 15, 2021 9:33 pm

MicroBob wrote:
Fri Jul 02, 2021 6:44 am
Hi Don,
can you post an image where your method has actually worked?

Bob
Watch my videos about microbes.

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#21 Post by DonSchaeffer » Thu Jul 15, 2021 9:48 pm

MicroBob wrote:
Fri Jul 02, 2021 6:44 am
Hi Don,
can you post an image where your method has actually worked?

Bob
These are screen prints of a video. I usually erase the stills after I upload to YouTube.
four.jpg
four.jpg (30.27 KiB) Viewed 7928 times
five.jpg
five.jpg (69.47 KiB) Viewed 7928 times
six.jpg
six.jpg (109.6 KiB) Viewed 7927 times

PeteM
Posts: 3005
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Converting Brightfield Microscope To DIC

#22 Post by PeteM » Fri Jul 16, 2021 3:02 am

For comparison, some actual DIC video images. Go in about 10 seconds to get an idea of the detail and contrast possible with DIC:

https://www.youtube.com/watch?v=UFErSrAr5xM

PeteM
Posts: 3005
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Converting Brightfield Microscope To DIC

#23 Post by PeteM » Fri Jul 16, 2021 3:05 am


DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#24 Post by DonSchaeffer » Fri Jul 16, 2021 5:43 am

Wow! That's the end of that argument. I think I was confusing phase contrast with DIC. What I was imitating was phase contrast!

PeteM
Posts: 3005
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Converting Brightfield Microscope To DIC

#25 Post by PeteM » Fri Jul 16, 2021 4:09 pm

No problem. I just didn't want others to be confused. Closing the condenser iris down does increase contrast and depth of field, but at the expensive of resolution.

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#26 Post by DonSchaeffer » Fri Jul 16, 2021 7:52 pm

The thing is, you can sometimes see things that will not be seen when the diaphragm is open.

MichaelG.
Posts: 4019
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Converting Brightfield Microscope To DIC

#27 Post by MichaelG. » Fri Jul 16, 2021 8:57 pm

DonSchaeffer wrote:
Fri Jul 16, 2021 7:52 pm
The thing is, you can sometimes see things that will not be seen when the diaphragm is open.
Some of which don’t actually exist … That was my point at post #13

MichaelG.
Too many 'projects'

DonSchaeffer
Posts: 3329
Joined: Sun Mar 22, 2020 10:06 am
Location: Winnipeg, Manitoba, Canada

Re: Converting Brightfield Microscope To DIC

#28 Post by DonSchaeffer » Fri Jul 16, 2021 9:26 pm

You mean that some of those entities in my last image, the moving strings of bacteria, don't really exist? Those amoebas don't exist?

Dubious
Posts: 426
Joined: Sun May 09, 2021 7:55 pm

Re: Converting Brightfield Microscope To DIC

#29 Post by Dubious » Fri Jul 16, 2021 10:17 pm

If they are moving around, they probably do exist--for some suspected bacteria and amoeba, that's about the best I can do :)

The tradeoff between reduced resolution and more contrast as you close the iris down that last bit of range should be noticeable if you focus on a fixed object (ie, not moving) that has tiny detail.

DIC (which I would like to have someday) is impressive for many things but introduces its own artifacts, namely the hills/valleys/contours that don't actually exist, which can leave some waterlife looking like it was sculpted out of mud. I was looking at some traveling rotifers scoot along the other day using Hoffman Modulation Contrast (similar to DIC but probably even worse in this respect) and remember thinking it made them look like they had alligator hide.

MichaelG.
Posts: 4019
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Converting Brightfield Microscope To DIC

#30 Post by MichaelG. » Fri Jul 16, 2021 10:49 pm

DonSchaeffer wrote:
Fri Jul 16, 2021 9:26 pm
You mean that some of those entities in my last image, the moving strings of bacteria, don't really exist? Those amoebas don't exist?
No, I mean some of the ‘detail’ within them doesn’t exist.

MichaelG.
.

Edit: __ Please see p8 of the excellent paper by Michael W. Davidson and Mortimer Abramowitz

Available here: https://cw.fel.cvut.cz/b182/_media/cour ... oscopy.pdf
When the aperture diaphragm is erroneously closed too far, resulting diffraction artifacts cause visible fringes, banding, and/or pattern formation in photomicrographs. Other problems, such as refraction phenomena, can also produce apparent structures in the image that are not real (21).
Last edited by MichaelG. on Sat Jul 17, 2021 6:08 am, edited 2 times in total.