The improper use of a condenser with swing out lens. An observation.

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Post Reply
Message
Author
The QCC
Posts: 397
Joined: Sat Oct 11, 2014 10:13 pm

The improper use of a condenser with swing out lens. An observation.

#1 Post by The QCC » Thu Jun 25, 2015 1:43 am

The linked presentation shows the relationship between the condenser, resolution and contrast when the condenser is improperly used.
The microscope is a Labomed polarizing unit with swing out condenser and field diaphragm.
The condenser with the swing out lens in place has a N.A. of 1.15 and a N.A. of 0.9 when swung out.

The subject, a rock thin section, was photographed using a 10x and 40x obj.
The camera / microscope / objectives were set up with Kohler illumination, the top condenser lens in place and the condenser aperture stopped down approximately 50% to achieve good, but not excessive contrast. The standard setup for higher power obj.

If no settings are changed except the top condenser lens is swung out, the contrast increases sharply and resolution falls like a stone.
The result is similar to stopping down the condenser aperture too much with a condenser without a swing out lens.
Click the image to see the photo presentation.
Click the image to see the photo presentation.
Ruble_13_5MLX_40x_04_NoTopCon.jpg (85.86 KiB) Viewed 4894 times
[/url]
__________________
Edit: The titled was edited by the author.

User avatar
Crater Eddie
Posts: 1858
Joined: Wed Nov 12, 2014 4:39 pm
Location: Illinois USA

Re: The improper use of a condenser with swing out lens. An observation.

#2 Post by Crater Eddie » Thu Jun 25, 2015 2:07 pm

Very interesting presentation. The difference in the image is obvious when presented in this way.
CE
Olympus BH-2 / BHTU
LOMO BIOLAM L-2-2
LOMO POLAM L-213 / BIOLAM L-211 hybrid
LOMO Multiscope (Biolam)
Cameras: Canon T3i, Olympus E-P1 MFT, Amscope 3mp USB

billbillt
Posts: 2895
Joined: Mon Nov 10, 2014 10:01 pm

Re: The improper use of a condenser with swing out lens. An observation.

#3 Post by billbillt » Thu Jun 25, 2015 3:34 pm

Yes, I am glad someone did this... Makes a person think about it while in use...

jwsmith
Posts: 70
Joined: Mon Jun 01, 2015 4:15 pm
Location: Berkeley, CA, USA

Re: The improper use of a condenser with swing out lens. An observation.

#4 Post by jwsmith » Thu Jun 25, 2015 5:01 pm

I have an American Optical scope with swing-out condenser lens and no field diaphram.
The light source has a swing-out lens.

The scope condenser does have a diaphram.

I am (have always been) uncertain why my scope is so-configured.
Experimentation (combined/single use of these features) has not led me to any clear conclusions as to use/advantage.

Comment?

Judd

The QCC
Posts: 397
Joined: Sat Oct 11, 2014 10:13 pm

Re: The improper use of a condenser with swing out lens. An observation.

#5 Post by The QCC » Thu Jun 25, 2015 6:08 pm

Judd:
I am not familiar with AO microscopes with light source swing out lens, but I can tell you what the condenser and its aperture do in a not very technical way.

When you are using a low power objective, one less than 10x, the condenser lens should be OUT. when using a higher power obj, 10x and above, the condenser lens should be IN.
The condenser creates a light cone that illuminates the specimen and the obj. field of view. The swing out lens adjusts the light cone to match the higher power objects narrower field of view.
The photos in my presentation show what happens when the swing out lens is in the correct and wrong position.

Without a field diaphragm in the light source, the condenser aperture is also used to adjust the light illuminating the specimen and obj. The aperture also has the effect of increasing contrast and resolution. The latter can be overdone resulting in loss of resolution.

Edit: Added sample images
4x obj. Top condenser out, aperture set to N.A. .5
Top condenser lens  OUT
Top condenser lens OUT
SilconTaconite_5MLX_4x_01_NA.5TCO.jpg (219.4 KiB) Viewed 4856 times
4x obj. Top condenser lens IN N.A. .1
Top condenser lens  IN
Top condenser lens IN
SilconTaconite_5MLX_4x_02_NA.1TCI.jpg (258.98 KiB) Viewed 4856 times
10x obj. Top condenser lens IN, aperture set to N.A. .3
Top condenser lens IN
Top condenser lens IN
SilconTaconite_5MLX_10x_04_NA.3TCI.jpg (174.66 KiB) Viewed 4856 times
10x ob., Top condenser lens OUT. aperture set to N.A. .9
Top condenser OUT
Top condenser OUT
SilconTaconite_5MLX_10x_05_NA.9TCO.jpg (251.97 KiB) Viewed 4856 times