Olympus U-CA

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Post Reply
Message
Author
biofilmer86
Posts: 31
Joined: Fri Dec 25, 2020 3:30 am

Olympus U-CA

#1 Post by biofilmer86 » Tue Mar 09, 2021 11:34 pm

Before I consider getting one I would like to know if the magnification changer causes any image quality problems. Also, if the numerical aperture has to be increased for the objective being used. If I get a 60x objective I can "boost" it up to 120x by setting the U-CA to 2x?

PeteM
Posts: 3013
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Olympus U-CA

#2 Post by PeteM » Wed Mar 10, 2021 12:45 am

The numerical aperture is determined by the objective. Rule of thumb is that if you multiply the objective's N.A. x 1000 . . . that will be the maximum magnification you can get without getting "empty" magnification (just a larger blur). The 1000x rule assumes you're using the right condenser, oiling it to the slide if needed at high magnifications, and so on.

So, for a .65 N.A. 40x objective you could be fine using a magnification changer at 1.5x since the 600x magnification would be less than that .65 na x 1000. At 2x you're into empty magnification.

I don't know of any dry 60x objective that could handle a 2x boost. There are some 60x Plan Apos oil type objectives that might respond OK.

biofilmer86
Posts: 31
Joined: Fri Dec 25, 2020 3:30 am

Re: Olympus U-CA

#3 Post by biofilmer86 » Wed Mar 10, 2021 3:56 am

Thanks for the small lesson on numerical aperture. Multiplying the N.A. number by 1000 makes perfect sense especially since the maximum limit of light microscopy is 1500x and the largest N.A. of any 100x oil objective I have seen so far is 1.45

https://www.spachoptics.com/NIKON-MRD01 ... ambda.htm

Something that is normally forgotten in microscopy is multiply the objective magnification by the eyepiece magnification for total / maximum magnification. Using the magnification changer on 2x is equivalent to using eyepieces with a value of 20x. Some microscope brands supply these eyepieces stating 2000x can be achieved which is true but useless since you exceed the 1500x limit.

User avatar
hkv
Posts: 1012
Joined: Sun Nov 02, 2014 8:57 pm
Location: Sweden
Contact:

Re: Olympus U-CA

#4 Post by hkv » Wed Mar 10, 2021 7:47 am

I am using the U-CA on my microscopes. The quality is fairly OK with no or little signs of aberration. However, every lens you add in the pipeline is a potential weak link causing quality degradation. I rarely use the 2X setting except for video, but frequently use 1.25X. The main advantage of the U-CA is to achieve a better composition of images, especially when recording video as it is more difficult to crop without losing too much pixels (unless you shoot 8K video...). As already mentioned, all it does is to add empty magnification which normally should be avoided. I prefer to use a camera with sufficient resolution to resolve the full NA instead of adding tube lenses to increase magnification (if possible). If you use the Olympus PE photo eyepieces to connect your camera, it is preferable to go from PE2.5 to PE3.3 instead of engaging the 1.25 mag, in the U-CA. That gives approximately the same effect, but with one less lens in the pipe.
Flickr: https://www.flickr.com/photos/micromundus
Instagram: https://www.instagram.com/micromundusphotography
Web: https://hakankvarnstrom.com
Olympus BX51 | Olympus CX23 | Olympus SZ40 | Carl ZEISS EVO LS 10 Lab6 | Carl Zeiss Jena Sedival

Post Reply