No power to light sources on Leica DMRE

Everything relating to microscopy hardware: Objectives, eyepieces, lamps and more.
Post Reply
Message
Author
lxlotl
Posts: 3
Joined: Sat Dec 09, 2023 1:13 am

No power to light sources on Leica DMRE

#1 Post by lxlotl » Thu Jan 04, 2024 8:19 pm

Hello all,

This is my first post on the forum, I'm a complete noob when it comes to microscopy.

I recently acquired my first microscope, an old huge Leica DMRE that was destined for the trash.
ImageLeicaDMRE by Linaria Larkin, on Flickr

Presently, the light sources do not work and I don't know why. I was wondering if you all would be able to help me at all diagnose what the problem is.
Probing with a multimeter after turning the light power on, there is no power getting to either light sources.
The microscope does get power, it has a front display for manipulating the z-height that does light up. But I also do not know much of anything about electronics, so I don't know where to even begin!

Thank you all in advance for any advice/help!
You must not fool yourself, and you are the easiest person to fool - Richard Feynman

MichaelG.
Posts: 4030
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: No power to light sources on Leica DMRE

#2 Post by MichaelG. » Thu Jan 04, 2024 10:23 pm

I’m afraid this won’t help [or make you happy] : https://www.ebay.co.uk/itm/256205241887

Hopefully, in amongst that lot … you might just possibly find a blown fuse !

… if not, it could be a long investigation.

MichaelG
Too many 'projects'

Hobbyst46
Posts: 4288
Joined: Mon Aug 21, 2017 9:02 pm

Re: No power to light sources on Leica DMRE

#3 Post by Hobbyst46 » Thu Jan 04, 2024 10:25 pm

How about both bulbs are burnt out, normal end of life ?

apochronaut
Posts: 6327
Joined: Fri May 15, 2015 12:15 am

Re: No power to light sources on Leica DMRE

#4 Post by apochronaut » Thu Jan 04, 2024 11:15 pm

If you smell around close to the back of the microscope, below where the louvers on top are : do you smell a pungent sort of burnt aroma?

PeteM
Posts: 3014
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: No power to light sources on Leica DMRE

#5 Post by PeteM » Fri Jan 05, 2024 3:35 am

Your DMRE is potentially a marvelously capable microscope.

And, the worst case for the lamps isn't so bad. If the problem turns out to be more than a burnt lamp or fuse, a loose wire, or just not understanding which switch switches what -- you could simply get an external 100 watt power supply (often available from $50 to $100) and wire it directly to the lamp.

If it's a bad power supply, you also likely have enough room in the base of the scope to get a modern switching power supply and do a replacement if the original power supply board is burnt out or missing.

apochronaut
Posts: 6327
Joined: Fri May 15, 2015 12:15 am

Re: No power to light sources on Leica DMRE

#6 Post by apochronaut » Fri Jan 05, 2024 1:32 pm

I would go to 150 watts for that but I would rule out the obvious usual culprits first. Fuse, capacitors, switching.Keep the in house p.s. if possible. Remote is a bit cumbersome and matching the microscope's power needs might be a bit tricky.
If the microscope sat for a while without being powered up, electrolytic capacitors can leak or dry up. They tend to fail more unused. A quick check by an electonics guy could find the problem fast. It doesn't have to be a microscope technician. It's just a power supply. No fan in that?

lxlotl
Posts: 3
Joined: Sat Dec 09, 2023 1:13 am

Re: No power to light sources on Leica DMRE

#7 Post by lxlotl » Fri Jan 05, 2024 2:18 pm

Hobbyst46 wrote:
Thu Jan 04, 2024 10:25 pm
How about both bulbs are burnt out, normal end of life ?
I wasn't able to directly probe the plug on the microscope for the light sources with a multimeter, but I was able to probe the connection within one of them, and there wasn't any voltage across them. Not super experienced with a multimeter, so I could've done something wrong, so I suppose it would be worth trying new bulbs.
You must not fool yourself, and you are the easiest person to fool - Richard Feynman

lxlotl
Posts: 3
Joined: Sat Dec 09, 2023 1:13 am

Re: No power to light sources on Leica DMRE

#8 Post by lxlotl » Fri Jan 05, 2024 2:21 pm

PeteM wrote:
Fri Jan 05, 2024 3:35 am
Your DMRE is potentially a marvelously capable microscope.

And, the worst case for the lamps isn't so bad. If the problem turns out to be more than a burnt lamp or fuse, a loose wire, or just not understanding which switch switches what -- you could simply get an external 100 watt power supply (often available from $50 to $100) and wire it directly to the lamp.

If it's a bad power supply, you also likely have enough room in the base of the scope to get a modern switching power supply and do a replacement if the original power supply board is burnt out or missing.
That is not a bad idea, for a worst case scenario! I do have a couple power supplies scavenged from other workplace instruments.
Reading through the manual there was quite a lot of words I had to look up, it seems I have a lot to learn if I want to use it to its fullest potential.
You must not fool yourself, and you are the easiest person to fool - Richard Feynman

apochronaut
Posts: 6327
Joined: Fri May 15, 2015 12:15 am

Re: No power to light sources on Leica DMRE

#9 Post by apochronaut » Fri Jan 05, 2024 3:29 pm

lxlotl wrote:
Fri Jan 05, 2024 2:18 pm
Hobbyst46 wrote:
Thu Jan 04, 2024 10:25 pm
How about both bulbs are burnt out, normal end of life ?
I wasn't able to directly probe the plug on the microscope for the light sources with a multimeter, but I was able to probe the connection within one of them, and there wasn't any voltage across them. Not super experienced with a multimeter, so I could've done something wrong, so I suppose it would be worth trying new bulbs.
Since you have a meter, just check continuity across the pins. Sometimes bulb filaments can fry close to the bracket and it is hard to see visually if they are broken.
One other place where those illuminators can give problems over time is in the socket itself. Once a pin loses contact just a little, it heats and oxides build up, causing weaker contact, more heat. Eventually the socket gives out.
Sometimes, a thin point file or emery paper can clean the contacts up and restore the socket but they can be horribly pitted and burned.
It is suspicious that both illuminators are out though. It does sound like a power supply issue.
I found a reference that says 140 watts. It should be on a tag somewhere on the instrument.

Alexander
Posts: 413
Joined: Fri Jun 11, 2021 7:10 pm

Re: No power to light sources on Leica DMRE

#10 Post by Alexander » Fri Jan 05, 2024 10:27 pm

The upper lamp housing looks very much like a HBO which needs a special power supply anyway.
On the picture the lower lamp's power line is not connected to the frame.

Is a bulb installed?
What resistance reads the multimeter on the pins of the lamp housing?

Post Reply