Testing high NA darkfield condenser
Testing high NA darkfield condenser
Here follow the thread on the unsold "ultracondenser", darkfield for oil immersion.
viewtopic.php?f=16&t=14790&sid=99ac660a ... 641bf528d1
New/refurbished tools to check out how the illumination is coming out:
- slide with pinhole 0.3mm (aluminium foil with an hole made with a pin, mounted under a coverslip);
- test slide, UNI-Posca pencil ink under coverslip; it is composed of nice fine pigment grains;
- Ground screen, a BK7 glass block with one side covered with milky grease; to visualize the illumination;
- Prism apertometer, a BK7 45 deg prism, with a pinhole in the center of hypotenuse (downward in the photo) and graduated NA scale on the sides;
- Fake immersion oil bottle (refilled with frying oil, new);
. .
Lenses used (dream team):
Lomo 60x 1.0-0.7 (actually goes to 0.2), good conditions;
CZJ 60x 1.40, clean but miscollimated;
Olympus Plan 100x 1.25 short, collimated but dirty;
Nikon PA 20x 0.75, supposedly good.
Eyepiece Olympus WHK 10x/20 goes well (only) with the Oly 100x;
The condenser to be tested, darkfield manufactured few days ago.
.
viewtopic.php?f=16&t=14790&sid=99ac660a ... 641bf528d1
New/refurbished tools to check out how the illumination is coming out:
- slide with pinhole 0.3mm (aluminium foil with an hole made with a pin, mounted under a coverslip);
- test slide, UNI-Posca pencil ink under coverslip; it is composed of nice fine pigment grains;
- Ground screen, a BK7 glass block with one side covered with milky grease; to visualize the illumination;
- Prism apertometer, a BK7 45 deg prism, with a pinhole in the center of hypotenuse (downward in the photo) and graduated NA scale on the sides;
- Fake immersion oil bottle (refilled with frying oil, new);
. .
Lenses used (dream team):
Lomo 60x 1.0-0.7 (actually goes to 0.2), good conditions;
CZJ 60x 1.40, clean but miscollimated;
Olympus Plan 100x 1.25 short, collimated but dirty;
Nikon PA 20x 0.75, supposedly good.
Eyepiece Olympus WHK 10x/20 goes well (only) with the Oly 100x;
The condenser to be tested, darkfield manufactured few days ago.
.
Last edited by patta on Sat Feb 12, 2022 8:49 am, edited 2 times in total.
Re: Testing high NA darkfield condenser
According to the prism apertometer, the condenser emits a narrow cone with NA around 1.4 (I'd say 1.38-1.45), over a 1.2 mm slide and like 0.2mm of condenser-slide spacing; oil at all interfaces. The illumination does not pass though water (total internal reflection). This only for the center.
No photos for this, not much interesting.
Below the photo of the illumination on the ground screen; it is the hollow cone that hits the diffuser plane, after passing through the pinhole placed where the specimen should be.
It is a relatively neat stripe, with no color fringes up or down. The bright bands are from the condenser ring mirror, that is a thin aluminum foil glued and pressed over 3D printed (relief-banded) surface.
. .
As comparison, below a refractive Abbe condenser (not achromatic), with a "darkfield patch" in the center; about 0.7-0.8; color fringes are visible
. .
Edit Saturday morning
A drawing that *should* illustrate the above photos; the "smile" in the photos is a vertical plane section of the illumination cone:
. .
And managed to take a similar photo with the 45 prism apertometer.
.
No photos for this, not much interesting.
Below the photo of the illumination on the ground screen; it is the hollow cone that hits the diffuser plane, after passing through the pinhole placed where the specimen should be.
It is a relatively neat stripe, with no color fringes up or down. The bright bands are from the condenser ring mirror, that is a thin aluminum foil glued and pressed over 3D printed (relief-banded) surface.
. .
As comparison, below a refractive Abbe condenser (not achromatic), with a "darkfield patch" in the center; about 0.7-0.8; color fringes are visible
. .
Edit Saturday morning
A drawing that *should* illustrate the above photos; the "smile" in the photos is a vertical plane section of the illumination cone:
. .
And managed to take a similar photo with the 45 prism apertometer.
.
Re: Testing high NA darkfield condenser
Now some photos.
The subject is acrylic paint ink from blue "posca" pencils, that is made of pigment grains about 0.5 micron in size. At the border of a line, the grains are few and separate. The black part is black because there are no scattering particles, just transparent mountant. If it's black, it's good, meaning the condenser+objective are working in darkfield mode.
No I don't have nice diatom test slides, just made this "grains slide".
Lamp is a 6W white led; imaging is afocal:
Visual eyepiece Olympus WHK 10x 20;
Objective Canon 50mm f/1.8
Sensor Canon 600D aps-c
It captures a field of about 12mm diagonal (that's my best)
The condenser is always the same.
Objective Lomo 60x at max aperture (NA 1.0), full field:
. .
Lomo 60x with the iris a bit closed (let's say NA 0.50):
. .
Center crop of said Lomo at 1.0
.
The subject is acrylic paint ink from blue "posca" pencils, that is made of pigment grains about 0.5 micron in size. At the border of a line, the grains are few and separate. The black part is black because there are no scattering particles, just transparent mountant. If it's black, it's good, meaning the condenser+objective are working in darkfield mode.
No I don't have nice diatom test slides, just made this "grains slide".
Lamp is a 6W white led; imaging is afocal:
Visual eyepiece Olympus WHK 10x 20;
Objective Canon 50mm f/1.8
Sensor Canon 600D aps-c
It captures a field of about 12mm diagonal (that's my best)
The condenser is always the same.
Objective Lomo 60x at max aperture (NA 1.0), full field:
. .
Lomo 60x with the iris a bit closed (let's say NA 0.50):
. .
Center crop of said Lomo at 1.0
.
Last edited by patta on Sat Feb 12, 2022 1:14 pm, edited 5 times in total.
Re: Testing high NA darkfield condenser
Same afocal setup as above, with the almighty objective Zeiss Jena Apochromat 60x 1.40
This objective is a bit miscollimated because I've opened it for cleaning once. Not really visible in the photos, but visually it shows some issues on-axis.
Full-field, focused in the center:
. .
Crop of above photo:
. .
now this objective isn't Plan, definitely. Let's try to refocus it midway in the field:
. .
And that's the enlarged mash-up:
. .
Pretty horrible eh? Anyway, the point here is so check that the CONDENSER is illuminating and is still dark away from the center.
Let's assume that the rainbow CA above arises from a deep misunderstanding between the Deutsch objective and the 日本 eyepiece, and not from my beloved condenser.
Ultra-resolution, I'm not seeing much of that. Let's blame the sunflower oil immersion.
This objective is a bit miscollimated because I've opened it for cleaning once. Not really visible in the photos, but visually it shows some issues on-axis.
Full-field, focused in the center:
. .
Crop of above photo:
. .
now this objective isn't Plan, definitely. Let's try to refocus it midway in the field:
. .
And that's the enlarged mash-up:
. .
Pretty horrible eh? Anyway, the point here is so check that the CONDENSER is illuminating and is still dark away from the center.
Let's assume that the rainbow CA above arises from a deep misunderstanding between the Deutsch objective and the 日本 eyepiece, and not from my beloved condenser.
Ultra-resolution, I'm not seeing much of that. Let's blame the sunflower oil immersion.
Last edited by patta on Sat Feb 12, 2022 8:33 pm, edited 4 times in total.
Re: Testing high NA darkfield condenser
Here finally a proper Plan with more or less proper eyepiece, Olympus 100x 1.25 HI Plan
. .
. .
Last edited by patta on Sat Feb 12, 2022 11:33 am, edited 1 time in total.
Re: Testing high NA darkfield condenser
The above images were Jpegs, resized straight from the camera - no adjustments. The background appears pitch-black, but visually ain't so, it is slightly bluish.
Here a photo with phone camera over the eyepiece, HDR. That's actually a bit exaggerate in term of exposure, the scattering grains are fully burnt in this image. Anyway, there is some stray light and haze, but not so much. I think was the Oly 100?
. .
The Olympus 100x in particular had, visually, a lot of haze. Looking at it with "phase telescope", revealed that indeed the top lens was "a bit dirty". Believe or not, the photo at the previous post has been taken through those geological deposits.
.
Here a photo with phone camera over the eyepiece, HDR. That's actually a bit exaggerate in term of exposure, the scattering grains are fully burnt in this image. Anyway, there is some stray light and haze, but not so much. I think was the Oly 100?
. .
The Olympus 100x in particular had, visually, a lot of haze. Looking at it with "phase telescope", revealed that indeed the top lens was "a bit dirty". Believe or not, the photo at the previous post has been taken through those geological deposits.
.
Last edited by patta on Sat Feb 12, 2022 12:47 pm, edited 5 times in total.
Re: Testing high NA darkfield condenser
Impressive project, patta!!
I would try some monochromatic light with it, perhaps a bandpass filter with the same LED.
I would try some monochromatic light with it, perhaps a bandpass filter with the same LED.
Re: Testing high NA darkfield condenser
Great, thanks. Now I don't have any strong monochromatic source or proper filter; oh, wait, a weak blue one; one of next days.
Last for today, let's check how much large field the condenser is covering, use the Nikon Planapo 20x 0.75; mounted on the newest, exclusive CFI stand. Black enamel is back in fashion! The field is limited by a 23mm eyepiece bore. No eyepiece, direct imaging with a camera objective as tube lens (either 90 or 125mm)
. .
The tube lens isn't working perfectly, but more or less a 1mm field is observable. The subject are the lil' points of grit on black background.
Here the slide is oiled to the condenser; the objective is dry.
.
Last for today, let's check how much large field the condenser is covering, use the Nikon Planapo 20x 0.75; mounted on the newest, exclusive CFI stand. Black enamel is back in fashion! The field is limited by a 23mm eyepiece bore. No eyepiece, direct imaging with a camera objective as tube lens (either 90 or 125mm)
. .
The tube lens isn't working perfectly, but more or less a 1mm field is observable. The subject are the lil' points of grit on black background.
Here the slide is oiled to the condenser; the objective is dry.
.
Last edited by patta on Sat Feb 12, 2022 1:32 pm, edited 12 times in total.
Re: Testing high NA darkfield condenser
I've put the 3D design of the condenser here:
https://www.thingiverse.com/thing:5241935
Uploaded also the FreeCAD model that is a bit messy, can be adapted by modifying the Master Sketch.
Materials:
3D printed parts. about 15 grams total
Aluminum foil
One half-ball lens 9mm diameter (a disk may also usable?)
One round coverslip 18mm
Canada balsam or cement to glue the coverslip
https://www.thingiverse.com/thing:5241935
Uploaded also the FreeCAD model that is a bit messy, can be adapted by modifying the Master Sketch.
Materials:
3D printed parts. about 15 grams total
Aluminum foil
One half-ball lens 9mm diameter (a disk may also usable?)
One round coverslip 18mm
Canada balsam or cement to glue the coverslip
Re: Testing high NA darkfield condenser
wait a minute, monochromatic LED.. the old trick of the phone screen; it is composed of OLEDS, if one visualize a green image, well' that's emitting green.
Unfortunately turned out too weak for this inefficient darkfield; I've managed only with the Green color, since the screen is more powerful in green and the camera is also more sensitive there. Anyway not strong enough for convenient live-view focusing. It looks sharper indeed, but the extreme low brightness spoils the result.
Hope the Germans sort that out https://www.mikroskopie-forum.de/index. ... ic=43184.0
Tested only with the Olympus 100x Plan:
. .
Unfortunately turned out too weak for this inefficient darkfield; I've managed only with the Green color, since the screen is more powerful in green and the camera is also more sensitive there. Anyway not strong enough for convenient live-view focusing. It looks sharper indeed, but the extreme low brightness spoils the result.
Hope the Germans sort that out https://www.mikroskopie-forum.de/index. ... ic=43184.0
Tested only with the Olympus 100x Plan:
. .
- Attachments
-
- Shoots_Centercrop2x_Afocal_Oly_Green.jpg (148.26 KiB) Viewed 10481 times
Re: Testing high NA darkfield condenser
It is very nice to achieve true DF with an 1.25NA objective , without an iris that lowers the NA.
I have two old 100x1.25 Zeiss achromat objectives. One of them is 160/-, the other bears no mark of tube length and coverslip (yes/no/both). Unexpectedly, the first one yields a not-too-bad DF with the oil 1.2-1.4 DF cardiodid condenser. But not nearly as nice as your DF above !
My second 100X does not yield DF at all - as expected.
I have two old 100x1.25 Zeiss achromat objectives. One of them is 160/-, the other bears no mark of tube length and coverslip (yes/no/both). Unexpectedly, the first one yields a not-too-bad DF with the oil 1.2-1.4 DF cardiodid condenser. But not nearly as nice as your DF above !
My second 100X does not yield DF at all - as expected.
Re: Testing high NA darkfield condenser
This is quite stunning (literaly) ... it seems to work well. Before you posted the photos using your mobile phone screen as a light source, I guessed that it had to be a pretty inefficient system.
Soooo the question is; why didnt anyone make a df condenser like this, before ?????
(When one reads the descriptions of old oil immersion df condensers most of them say that a funnel stop must be used with oil objectives to limit the numerical aperture).
Apochronaut mentions chromatic effects, do you think that might be a reason?
BTW was the half-spherical lens exactly half or was it a lens that fell out of the top of a condenser, that might be slightly more than half?
Soooo the question is; why didnt anyone make a df condenser like this, before ?????
(When one reads the descriptions of old oil immersion df condensers most of them say that a funnel stop must be used with oil objectives to limit the numerical aperture).
Apochronaut mentions chromatic effects, do you think that might be a reason?
BTW was the half-spherical lens exactly half or was it a lens that fell out of the top of a condenser, that might be slightly more than half?
Re: Testing high NA darkfield condenser
Yes the phone screen it is very dim and inefficient in terms of energy - but quite effective in terms of setup time (search "screen green image" on the phone, and there you have it )
This high NA could be obtained also (possibly better) with a Cardioid design. I'm sure has been done in the past; the fancy "Colonel Woodward" illuminator in the previous thread, looks like it's capable of doing such high NA oblique darkfield. Today, I think it is not common because of practical issues:
- not many suitable subjects. The subject needs to be very thin, mounted in oil or balsam. No live ciliates in water mount! No live blood analysis!
- Darkfield haze: from the images above, the field is really dark; but near the particles, it gets hazy. my guess is that haze comes from the nearby particles out of focus; since the illumination is a very shallow cone (high NA) a lot of haze is produced. With smaller NA, less haze. See sketch attached below; and in post #3 above, the objective with iris closed, cuts off haze. This issue can get very bad with complicated, packed biological subjects.
- Chromatic aberration, after above testing I'd say it is not so troublesome, and if it is, one could go monochromatic, like in phase contrast. The violent rainbows of the 60x 1.40, I'd say is improper eyepiece compensation rather than condenser itself- but not sure yet.
- Modest increase in resolution: sad truth; going from NA 1.25 to 1.4 is a big hassle; but resolution improvement is just some 10% (while price increase of objective is some 1000%)
.
This high NA could be obtained also (possibly better) with a Cardioid design. I'm sure has been done in the past; the fancy "Colonel Woodward" illuminator in the previous thread, looks like it's capable of doing such high NA oblique darkfield. Today, I think it is not common because of practical issues:
- not many suitable subjects. The subject needs to be very thin, mounted in oil or balsam. No live ciliates in water mount! No live blood analysis!
- Darkfield haze: from the images above, the field is really dark; but near the particles, it gets hazy. my guess is that haze comes from the nearby particles out of focus; since the illumination is a very shallow cone (high NA) a lot of haze is produced. With smaller NA, less haze. See sketch attached below; and in post #3 above, the objective with iris closed, cuts off haze. This issue can get very bad with complicated, packed biological subjects.
- Chromatic aberration, after above testing I'd say it is not so troublesome, and if it is, one could go monochromatic, like in phase contrast. The violent rainbows of the 60x 1.40, I'd say is improper eyepiece compensation rather than condenser itself- but not sure yet.
- Modest increase in resolution: sad truth; going from NA 1.25 to 1.4 is a big hassle; but resolution improvement is just some 10% (while price increase of objective is some 1000%)
.
- Attachments
-
- DF_highNA_haze.jpg (47.54 KiB) Viewed 10419 times
Last edited by patta on Mon Feb 14, 2022 1:58 pm, edited 4 times in total.
Re: Testing high NA darkfield condenser
Ah forgot, the half-ball lens is really half-sphere, I bought a bunch for another project from Aliexpress Mowen/Bohr (they were 1.5$ each, but they doubled the price last week well, still 1/10th than other suppliers.) I chose it for this condenser because it's a standard piece available in stock, only spec is its radius. There are also made of Silica if one fancies UV.
Ideally it should be a bit less than half sphere, such that spherical surface is concentric with the subject.
That was why initially the condenser was made for use with thin slide (the subject isn't much off-center); Now, over a 1.2mm slide, the subject is pushed up off-center, the light isn't impinging orthogonally on the sphere surface, so there is some refraction and hence there must be some CA.
Last photos for this testing campaign: Radial polarization
Normally polarization is linear; with this condenser it is easy to get "radial" polarization, meaning the polarization is different along the hollow cone.
It seems to be used in lasers, or to improve resolution.
This polarization can be achieved for a straight beam, either by combining slices of linear polarizers, like a recombined cake; or with some swirling liquid crystals.
With this modern half-ball condenser, the radial polarizer can be done over the converging cone by rolling a film of linear polarizer:
. . .
Photo results: Nothing! Not even with a linear pol analyzer after the objective. Well It's darkfield, we see only the scattered light; Hopefully will be more interesting with other subjects or with COL.
. . .
Ideally it should be a bit less than half sphere, such that spherical surface is concentric with the subject.
That was why initially the condenser was made for use with thin slide (the subject isn't much off-center); Now, over a 1.2mm slide, the subject is pushed up off-center, the light isn't impinging orthogonally on the sphere surface, so there is some refraction and hence there must be some CA.
Last photos for this testing campaign: Radial polarization
Normally polarization is linear; with this condenser it is easy to get "radial" polarization, meaning the polarization is different along the hollow cone.
It seems to be used in lasers, or to improve resolution.
This polarization can be achieved for a straight beam, either by combining slices of linear polarizers, like a recombined cake; or with some swirling liquid crystals.
With this modern half-ball condenser, the radial polarizer can be done over the converging cone by rolling a film of linear polarizer:
. . .
Photo results: Nothing! Not even with a linear pol analyzer after the objective. Well It's darkfield, we see only the scattered light; Hopefully will be more interesting with other subjects or with COL.
. . .
Last edited by patta on Mon Feb 14, 2022 1:32 pm, edited 1 time in total.
Re: Testing high NA darkfield condenser
About rainbows, here's how the calibration slide looks like in DF (plated chrome over glass)
The same effect can be seen with naked eye, the 0.01mm ruler of the calibration slide looks colored from some angle, like diffraction grating.
.
The same effect can be seen with naked eye, the 0.01mm ruler of the calibration slide looks colored from some angle, like diffraction grating.
.
Re: Testing high NA darkfield condenser
I was looking at the description of Watson's immersion paraboloid:
a] That if one shuts down the iris with this condenser head, then the lowest NA light is blocked first
b] This condenser prefers a brilliant source of light (in contrast to the Cassegrain where a pearl-frosted bulb is recommended)
I am sort of wondering whether your condensers behaviour might, in someways, be more similar to the paraboloid's than the cassegrain's.
?
The text made two interesting comments:a] That if one shuts down the iris with this condenser head, then the lowest NA light is blocked first
b] This condenser prefers a brilliant source of light (in contrast to the Cassegrain where a pearl-frosted bulb is recommended)
I am sort of wondering whether your condensers behaviour might, in someways, be more similar to the paraboloid's than the cassegrain's.
?
Re: Testing high NA darkfield condenser
Would a high index ball lens be helpful here at all? You can get some 2.0 index balls and half balls from Edmund Optics. Pricey, but not insane pricey.
The highest quality live-streamed microscopy in the world.
Sundays around 8PST: https://www.youtube.com/@diettoms/streams
Occasionally (for now): https://www.twitch.tv/diettoms
Join the Discord: https://discord.gg/FgpUUnJaSE
Sundays around 8PST: https://www.youtube.com/@diettoms/streams
Occasionally (for now): https://www.twitch.tv/diettoms
Join the Discord: https://discord.gg/FgpUUnJaSE
-
- Posts: 6327
- Joined: Fri May 15, 2015 12:15 am
Re: Testing high NA darkfield condenser
Paraboloids are like this one , single reflecting condensers, so the illumination beam gets reversed with the iris diaphragm first closing on the part of the light cone that ends up being the interior or low N.A. Double reflecting condensers such as Cassegrain or Cardioid or Spherical reverse that again.Chas wrote: ↑Mon Feb 14, 2022 8:57 pmI was looking at the description of Watson's immersion paraboloid:
Watson immersion paraboloid resized.jpg
The text made two interesting comments:
a] That if one shuts down the iris with this condenser head, then the lowest NA light is blocked first
b] This condenser prefers a brilliant source of light (in contrast to the Cassegrain where a pearl-frosted bulb is recommended)
I am sort of wondering whether your condensers behaviour might, in someways, be more similar to the paraboloid's than the cassegrain's.
?
Re: Testing high NA darkfield condenser
Yes the condenser here is in practice like a Paraboloid. Below the examples, the first one is the half-ball of this thread.
It would be nice also to remake an old style dry paraboloid, like the dry Wenham.
Cassegrain, I think there there are 2 different:
- Cass - "Schwarzchild" like a telescope, with two mirrors; also like the "reflective" microscope objectives. It is still in use, with the reflective objectives; don't know if it is used also for darkfield.
- Cass - "Nelson", I've seen it only on Watson's catalogue, with two pieces of glasses and a nasty double internal reflection; in the tiny characters below write NA 1.37 - 1.47, oh yeah.
I cant understand the why of the high NA Nelson-Cassegrain; a normal Paraboloid or Cardioid should be capable as well of high NA. Maybe at the time the glass available had too many air bubbles in it, so DF was impaired; while the thin top of the Cassegrain could be blown or stamped to remove bubbles? or the second reflection further reduces stray light?
. . .
Images from microscope-antique, Quekett, Archive.org etc.
It would be nice also to remake an old style dry paraboloid, like the dry Wenham.
Cassegrain, I think there there are 2 different:
- Cass - "Schwarzchild" like a telescope, with two mirrors; also like the "reflective" microscope objectives. It is still in use, with the reflective objectives; don't know if it is used also for darkfield.
- Cass - "Nelson", I've seen it only on Watson's catalogue, with two pieces of glasses and a nasty double internal reflection; in the tiny characters below write NA 1.37 - 1.47, oh yeah.
I cant understand the why of the high NA Nelson-Cassegrain; a normal Paraboloid or Cardioid should be capable as well of high NA. Maybe at the time the glass available had too many air bubbles in it, so DF was impaired; while the thin top of the Cassegrain could be blown or stamped to remove bubbles? or the second reflection further reduces stray light?
. . .
Images from microscope-antique, Quekett, Archive.org etc.
Re: Testing high NA darkfield condenser
That seems like a good explanation. The way it’s drawn, it looks like that second reflection is designed to be right near the critical angle. So any light which would tend to go into a lower NA cone would not TIR?
The highest quality live-streamed microscopy in the world.
Sundays around 8PST: https://www.youtube.com/@diettoms/streams
Occasionally (for now): https://www.twitch.tv/diettoms
Join the Discord: https://discord.gg/FgpUUnJaSE
Sundays around 8PST: https://www.youtube.com/@diettoms/streams
Occasionally (for now): https://www.twitch.tv/diettoms
Join the Discord: https://discord.gg/FgpUUnJaSE
Re: Testing high NA darkfield condenser
"a normal Paraboloid or Cardioid should be capable as well of high NA"
Here is a table of the performances of presumably modern implementations of the designs from the Olympus site:
https://www.olympus-lifescience.com/en/ ... darkfield/
Or another one, attributed to Needham, ('The microscope a practical guide'??):
Taken from:
https://www.chem.uci.edu/~dmitryf/manua ... oscopy.pdf
Which also contains some drawings of the Spot Ring Bicentric et al
Here is a table of the performances of presumably modern implementations of the designs from the Olympus site:
https://www.olympus-lifescience.com/en/ ... darkfield/
Or another one, attributed to Needham, ('The microscope a practical guide'??):
Taken from:
https://www.chem.uci.edu/~dmitryf/manua ... oscopy.pdf
Which also contains some drawings of the Spot Ring Bicentric et al
Re: Testing high NA darkfield condenser
And..as a little distraction; attached is E.M.Nelson's obituary:
http://www.microscopist.co.uk/about/fav ... opy-paper/
And just in case you are wondering about 'Norbert's 19th band' here is a further distracting link:http://www.microscopist.co.uk/about/fav ... opy-paper/
-
- Posts: 6327
- Joined: Fri May 15, 2015 12:15 am
Re: Testing high NA darkfield condenser
Those sites, such as the Olympus Lifescience .com appear to be prepared by several authors under contract. The same information appears on a Nikon page as well as others, slightly specific to the host corporation's product line but highly generalized to explain various principles of microscopy. The chart of DF types is also highly generalized. Several of the types, Paraboloid and Cardioid for instance, have many permutations from many manufacturers and the specifications are different, depending on which condenser is being considered. A Cassegrain type design on the other hand does not seem to have been embraced by too many companies, possibly only a couple of designs exist, with each originating with the same inventor.Chas wrote: ↑Tue Feb 15, 2022 8:47 pm"a normal Paraboloid or Cardioid should be capable as well of high NA"
Here is a table of the performances of presumably modern implementations of the designs from the Olympus site:
DF condensers from olympus site.jpg
https://www.olympus-lifescience.com/en/ ... darkfield/
Or another one, attributed to Needham, ('The microscope a practical guide'??):
DF Table orginaly from Needham.jpg
Taken from:
https://www.chem.uci.edu/~dmitryf/manua ... oscopy.pdf
Which also contains some drawings of the Spot Ring Bicentric et al
The array on that site has not much to do with Olympus. It is a historical compendium.
-
- Posts: 6327
- Joined: Fri May 15, 2015 12:15 am
Re: Testing high NA darkfield condenser
Optical glass was worked and worked to remove air bubbles and then shattered and inspected for imperfections. I can't see that glass with air bubbles would have been used to make anything in microscope optics, after a certain point in time : mid.19th century or before.patta wrote: ↑Tue Feb 15, 2022 2:22 pmYes the condenser here is in practice like a Paraboloid. Below the examples, the first one is the half-ball of this thread.
It would be nice also to remake an old style dry paraboloid, like the dry Wenham.
Cassegrain, I think there there are 2 different:
- Cass - "Schwarzchild" like a telescope, with two mirrors; also like the "reflective" microscope objectives. It is still in use, with the reflective objectives; don't know if it is used also for darkfield.
- Cass - "Nelson", I've seen it only on Watson's catalogue, with two pieces of glasses and a nasty double internal reflection; in the tiny characters below write NA 1.37 - 1.47, oh yeah.
I cant understand the why of the high NA Nelson-Cassegrain; a normal Paraboloid or Cardioid should be capable as well of high NA. Maybe at the time the glass available had too many air bubbles in it, so DF was impaired; while the thin top of the Cassegrain could be blown or stamped to remove bubbles? or the second reflection further reduces stray light?
.
Vintage_DF_condensers.jpg
.
Vintage_DF_condensers_cassegrains.jpg
.
Images from microscope-antique, Quekett, Archive.org etc.
One thing worth considering Patta, is the effect that sunflower oil is having on the N.A. of your entire optical system. With a maximum refractive index of 1.473 but likely less , your objective for starters is no longer a 1.4 N.A., plus the glass to glass homogeneity between the upper condenser element and slide is being broken, reducing the angle and N.A. You are working with lower angles than if homogeneity was intact. Maybe you should calculate this out and see what the angles actually are? Maybe closer to 1.25 for the objective?
I have a Reichert glycerin immersion 63X objective, which I occasionally used to use for DF along with an oil 100X. Sometimes, I have absent mindedly dipped the oil objective into the glycerin before I cleaned off the slide with a resultant dramatic, reduction in resolution. Glycerin has pretty close to the same N.A. as sunflower oil.
Re: Testing high NA darkfield condenser
I have only just noticed this... but Watson's immersion Paraboloid was sold with a "supply of suitable slips"
And the thickness of the slides "...must be within 20% of thickness engraved on each of each of the paraboloid mountings"
And they werent very generous with the slides ! :
https://collection.sciencemuseumgroup.o ... condensers
(that one says 1.10 mm on the side)
This one says 1.6 mm on the side:
https://www.microscope-antiques.com/HIP.html
And the thickness of the slides "...must be within 20% of thickness engraved on each of each of the paraboloid mountings"
And they werent very generous with the slides ! :
https://collection.sciencemuseumgroup.o ... condensers
(that one says 1.10 mm on the side)
This one says 1.6 mm on the side:
https://www.microscope-antiques.com/HIP.html
-
- Posts: 6327
- Joined: Fri May 15, 2015 12:15 am
Re: Testing high NA darkfield condenser
The principle of DF requires that the rays entering the slide do not pass through the upper surface, they reflect off of it and undergo TIR in the slide. Essentially the slide becomes like a fiber optic. If the slide thickness is too thick or thin, then one gets a partial DF with an uneven background and if the thickness is too far off the field brightens. Most DF condensers are quite specitic regarding slide thickness , with some stamping the specifications right on the condenser.
Re: Testing high NA darkfield condenser
On both tables above there is the word "aplanatic"; and also in the Nelson's obituary! So should be important for high NA, even if the illumination is just a thin ring. At NA 1.4 and 1.5 the angles are so shallow, aberrations likely get gigantic. So the condenser needs to be well made and relatively sharp, like a high NA objective - with its shallow focus depth, hence tight slide thickness. Two mirror surfaces are needed to get aplanatic, one isn't enough.
Below an illustration of how I think the current condenser get around this issue, with cheats:
-The mirror ain't conic nor parabolic, but is rough and wobbly, so it actually diffuses the light in many directions, instead of focusing it;
- The baffle stops that determine the aperture are after the mirror, so final NA isn't affected by aberrations (assuming that the glass half-ball has no significant effects)
So in practice this condenser is working as a "diffuse ring light + thin annular slits"; very inefficient in terms of light use, but no worries about aberrations. For illumination it needed 6 Watts LED at full power, which in brightfield would be "scorching sun".
Yes I'll get asap a gallon of proper immersion oil and test again (more rational choice than getting a refractometer for vegetable oil); extra perk, the runny sunflower oil was giving much less bubbles than the thick "B" imm oil.
Working NA, the "apertometer" used (photos in post #1 &2 of this thread; it is a prism from binocular) it's measuring the angle at which light cone comes out, after slide and oil layers; at the time I've had measured its RI, and checked it against objectives, it was working ok; so the aperture shouldn't be too much off; however that's also true that the NA of the condenser came out always higher than designed...?
I've stopped working on this condenser, will just use it and test a bit more. Well, ideas about modifications always welcome.
Thanks everybody for the historical contributions, now I know much more about condensers than two weeks ago. Hopefully some manufacturer will start again making the "Nelson Cassegrain"!
. .
Below an illustration of how I think the current condenser get around this issue, with cheats:
-The mirror ain't conic nor parabolic, but is rough and wobbly, so it actually diffuses the light in many directions, instead of focusing it;
- The baffle stops that determine the aperture are after the mirror, so final NA isn't affected by aberrations (assuming that the glass half-ball has no significant effects)
So in practice this condenser is working as a "diffuse ring light + thin annular slits"; very inefficient in terms of light use, but no worries about aberrations. For illumination it needed 6 Watts LED at full power, which in brightfield would be "scorching sun".
Yes I'll get asap a gallon of proper immersion oil and test again (more rational choice than getting a refractometer for vegetable oil); extra perk, the runny sunflower oil was giving much less bubbles than the thick "B" imm oil.
Working NA, the "apertometer" used (photos in post #1 &2 of this thread; it is a prism from binocular) it's measuring the angle at which light cone comes out, after slide and oil layers; at the time I've had measured its RI, and checked it against objectives, it was working ok; so the aperture shouldn't be too much off; however that's also true that the NA of the condenser came out always higher than designed...?
I've stopped working on this condenser, will just use it and test a bit more. Well, ideas about modifications always welcome.
Thanks everybody for the historical contributions, now I know much more about condensers than two weeks ago. Hopefully some manufacturer will start again making the "Nelson Cassegrain"!
. .
Re: Testing high NA darkfield condenser
This was the Mk.I soft drinks can Cobb slide holder:
The carboard ends do help some mechical stages to 'get a grip'.
Mistakes (apart from trying to flatten the holder whilst the coverslips were inside) :
1] The slightly jagged edges can break the corners off the slips, so a more continuous scissor cut after getting the size about right, would be an improvement.
2] The can material was just a bit to springy, I think I should have attempted to anneal it first.
-It has a 3/4 inch punch hole in the middle and and 'half a hole' puched each side of it.The carboard ends do help some mechical stages to 'get a grip'.
Mistakes (apart from trying to flatten the holder whilst the coverslips were inside) :
1] The slightly jagged edges can break the corners off the slips, so a more continuous scissor cut after getting the size about right, would be an improvement.
2] The can material was just a bit to springy, I think I should have attempted to anneal it first.
Re: Testing high NA darkfield condenser
Apart coverslip holders annealing (and maybe a compressor..)
I'm working a bit more on the condenser, will make the base to fit other stands.
Could somebody help me with the dimensions? I have only the ancient Swift and the Leitz dovetail, and can't find reliable data on the sizes of other makers.
Need Fit Diameter and the approximate height of condensers, as drawing below;
The more the better; at least the "standard 37mm"; maybe also other more sophisticate fittings like dovetails
.
I'm working a bit more on the condenser, will make the base to fit other stands.
Could somebody help me with the dimensions? I have only the ancient Swift and the Leitz dovetail, and can't find reliable data on the sizes of other makers.
Need Fit Diameter and the approximate height of condensers, as drawing below;
The more the better; at least the "standard 37mm"; maybe also other more sophisticate fittings like dovetails
.
Re: Testing high NA darkfield condenser
Does "Standard" refer to Zeiss? if so, I can provide some dimensions later this evening.