#1 Post by MichaelG. » Fri Mar 29, 2024 8:34 am

I have recently acquired a fairly basic IMT-2 which is causing me some confusion

The Halogen lamp-house looks very much like the one for the BH-2 and does not have the very distinctive umbilical cord that seems to feature in all the available IMT-2 literature.

It is also very clearly labelled as using a 100W lamp [which is verboten according to that literature].

Can anyone please advise ?

My enquiry to Evident in the U.K. was met with a very friendly, but ultimately unhelpful, response that the microscope is very old and they have no documentation on file.

Thanks
MichaelG.
.

.
IMG_9342.jpeg
IMG_9342.jpeg (131.73 KiB) Viewed 2979 times
IMG_9342.jpeg
IMG_9342.jpeg (146.23 KiB) Viewed 2979 times
IMG_9342.jpeg
IMG_9342.jpeg (147.01 KiB) Viewed 2979 times
Too many 'projects'

FredH
Posts: 20
Joined: Sun Sep 04, 2022 7:45 pm
Location: Wisconsin

Re: Olympus IMT-2 mystery

#2 Post by FredH » Fri Mar 29, 2024 2:45 pm

The lamp housing that came with my IMT-2 is labeled "LH50A" and says "Replacement Lamp 12 V 50 W HAL". However, it is lacking the umbilical cord that plugs into the microscope body, and instead has a cord that goes to a crude homemade power supply.

I suspect that the lamp housing on my IMT-2 is original to the microscope, because it looks like the lamp housing seen in a few of the photos pointed to in the thread "Restoring an Olympus IMT-2" by essence25:

viewtopic.php?f=5&t=16114

essence25 implied his lamp housing used a 12 V 50W halogen lamp.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#3 Post by MichaelG. » Fri Mar 29, 2024 3:36 pm

Thanks for the comments, and the link

Just to be clear … my IMT-2 has no umbilical, and no provision for plugging one in.

The lamp is powered internally … which presumably means that microscope PSU design was upgraded at some stage in the product development.

It may be worth mentioning that the Serial Number 405001 is very clearly stamped on the rating plate.

MichaelG.

.
P.S. _ This, from the available instruction booklet, shows the standard lamp-house and gives the warning about not using 100W bulbs:
.
IMG_9342.jpeg
IMG_9342.jpeg (134.12 KiB) Viewed 2943 times
Too many 'projects'

PeteM
Posts: 3028
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Olympus IMT-2 mystery

#4 Post by PeteM » Fri Mar 29, 2024 6:49 pm

Michael, I suspect your scope should have the black metal 50-watt halogen illuminator - and that someone switched lamp housings along the way.

If you check the electrical plate for your scope, it may shed light (metaphorical kind) on what lamp wattage the power supply can reliably support. If the input rating isn't some safe margin (V x A) above 100 watts, I'd suspect you might still be able to use the housing with a 50-watt bulb. If it looks capable of supporting (say with an input of at least 130 to 150 watts), then maybe it's a later model.

Your IMT-2 is a really nice (if massive) scope. The mechanical build is excellent and the optics available still near state-of-the-art for brightfield and phase contrast.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#5 Post by MichaelG. » Fri Mar 29, 2024 9:03 pm

PeteM wrote:
Fri Mar 29, 2024 6:49 pm
Michael, I suspect your scope should have the black metal 50-watt halogen illuminator - and that someone switched lamp housings along the way.

If you check the electrical plate for your scope, it may shed light (metaphorical kind) on what lamp wattage the power supply can reliably support. If the input rating isn't some safe margin (V x A) above 100 watts, I'd suspect you might still be able to use the housing with a 50-watt bulb. If it looks capable of supporting (say with an input of at least 130 to 150 watts), then maybe it's a later model.

Your IMT-2 is a really nice (if massive) scope. The mechanical build is excellent and the optics available still near state-of-the-art for brightfield and phase contrast.
Thanks for the thought, Pete … but I don’t see how that [emboldened in the quote] could be
There is no electrical connector, and not even a blanking plate, at the top of the ‘scope … the bulb holder appears to be integrated into the frame.

I will take some detail photos tomorrow … but meanwhile; the 50W lamp-house is detailed on FIG 7 of this:
https://www.alanwood.net/downloads/olym ... agrams.pdf

I am loving it so far … just a little confused !

MichaelG.
.
Rating Plate  … QED
Rating Plate … QED
IMG_9342.jpeg (62.18 KiB) Viewed 2900 times
Too many 'projects'

Hobbyst46
Posts: 4291
Joined: Mon Aug 21, 2017 9:02 pm

Re: Olympus IMT-2 mystery

#6 Post by Hobbyst46 » Fri Mar 29, 2024 9:34 pm

MichaelG. wrote:
Fri Mar 29, 2024 9:03 pm
...
I will take some detail photos tomorrow … but meanwhile; the 50W lamp-house is detailed on FIG 7 of this:
https://www.alanwood.net/downloads/olym ... agrams.pdf
Michael, Fig 20 on that list hints (maybe) that there was more than one illumination housing. So, while a 50W lamp seems the default, perhaps there was a 100 W - for more demanding modalities ? fluorescence with a 100W halogen lamp comes to mind, although the during the IMT-2 era, epifluorescence was well established, and 100W transmission excitation seems very unlikely...

PeteM
Posts: 3028
Joined: Tue Dec 06, 2016 6:22 am
Location: N. California

Re: Olympus IMT-2 mystery

#7 Post by PeteM » Fri Mar 29, 2024 9:36 pm

Michael, the rating plate on your microscope suggests the power supply can support 100 watt lamps (100+ volts x 2A = 200 or so watts input). I've only seen versions of the IMT-2 with that 50-watt lamp housing. But, odds are, you're OK.
Last edited by PeteM on Fri Mar 29, 2024 9:38 pm, edited 1 time in total.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#8 Post by MichaelG. » Fri Mar 29, 2024 9:38 pm

Indeed it does, Pete

In fact it explicitly states LAMP 12V 100W

MichaelG.
Too many 'projects'

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#9 Post by MichaelG. » Fri Mar 29, 2024 10:03 pm

Hobbyst46 wrote:
Fri Mar 29, 2024 9:34 pm
Michael, Fig 20 on that list hints (maybe) that there was more than one illumination housing. So, while a 50W lamp seems the default, perhaps there was a 100 W - for more demanding modalities ? fluorescence with a 100W halogen lamp comes to mind, although the during the IMT-2 era, epifluorescence was well established, and 100W transmission excitation seems very unlikely...
[/quote]

The epi-fluorescence lighting [which I do not have, and am unlikely to ever acquire] is a very different “kettle of fish” which inserts into a dedicated port.

The full system [of which mine is but a small part] is illustrated in all its glory, here:
https://www.alanwood.net/downloads/olym ... ochure.pdf

… With the notable exception of my mystery lamp-house !

MichaelG.
Too many 'projects'

Tom Jones
Posts: 340
Joined: Wed Apr 13, 2016 3:47 pm

Re: Olympus IMT-2 mystery

#10 Post by Tom Jones » Sat Mar 30, 2024 2:56 am

I had an IMT-2 with a 100w power supply and lamp house as well. I ended up selling it off, but they do exist.
I don't think I've ever seen it in a manual. My manual only lists the 50w P/S and standard lamp house. It even says not to use a 100w bulb or the power supply may be damaged, but that's with the 50w P/S.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#11 Post by MichaelG. » Sat Mar 30, 2024 7:11 am

.
Thanks for the sanity-check, Tom

… I wonder how many they made
… or how few :?

Do you happen to recall the serial number of yours ?

MichaelG.
Too many 'projects'

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#12 Post by MichaelG. » Sat Mar 30, 2024 8:03 am

I hope these snapshots will clarify

The housing is simply a shielded and ventilated box that clamps on with a single knurled screw
… beautifully built, but just a cover.

The bulb-holder is a fixed installation in the frame of the microscope
… the white wiring can just be seen in one image.

In short, this bears very little resemblance to the well-documented 50W version.

MichaelG.
.
IMG_9499.jpeg
IMG_9499.jpeg (146.96 KiB) Viewed 2835 times
.
IMG_9500.jpeg
IMG_9500.jpeg (187.24 KiB) Viewed 2835 times
.
IMG_9501.jpeg
IMG_9501.jpeg (163.97 KiB) Viewed 2835 times
.
IMG_9502.jpeg
IMG_9502.jpeg (178.74 KiB) Viewed 2835 times
Too many 'projects'

Rorschach
Posts: 377
Joined: Sun Mar 10, 2019 11:44 am

Re: Olympus IMT-2 mystery

#13 Post by Rorschach » Sat Mar 30, 2024 8:55 am

You have a nice scope in your hands there! I like the IMT-2 a lot.

We have one in our unit and we use it mainly for teaching, plankton for the most part. It was originally equipped with phase and epi-fluorescence, although we never used the fluorescence system. Now I have equipped if for DIC and will also get a quality LED-conversion plus a Zeiss Axiocam 208 Color camera, which is a 4K camera with a particularly good live image (very important for teaching). The camera will serve both the IMT-2 and the soon-to-arrive Zeiss Jena Sedival inverted scope.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#14 Post by MichaelG. » Sat Mar 30, 2024 5:13 pm

Rorschach wrote:
Sat Mar 30, 2024 8:55 am
You have a nice scope in your hands there! I like the IMT-2 a lot.
[…]
Thanks for the encouraging endorsement … I must say that I am delighted with it so far

… and the ‘correction collar’ adjustment for 0 to 2 mm glass thickness is astonishing, and effective.

MichaelG.
Too many 'projects'

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#15 Post by MichaelG. » Mon Apr 01, 2024 5:00 pm

A different mystery today …

How do I free-up the CT focus ring ?

This is stacked above the Magnification Changer, and illustrated in the exploded parts diagram:
.
IMG_9342.jpeg
IMG_9342.jpeg (190.18 KiB) Viewed 2679 times
.
But it is not obvious how it actually operates.

… The two rings currently appear to be locked firmly together, but I suspect that’s just dried grease and lack of use.

How far should it move ?
and does a vertical gap appear between the two rings when focussing the CT ?

Grateful for any advice from the wise, as I don’t really want to start dismantling.

MichaelG.
Too many 'projects'

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#16 Post by MichaelG. » Wed Apr 03, 2024 9:33 am

MichaelG. wrote:
Mon Apr 01, 2024 5:00 pm
A different mystery today …

How do I free-up the CT focus ring ?

[ … ]
MichaelG.
.
Panic over

I first removed the two filter cubes and stored them safely in a clean box.
… Then wedged the changer wheel in place and gently warmed the the whole wheel assembly with a little hot-air gun
[ at a guess it might have reached 35°C but no higher ]

Some pressure on the CT focus ring then freed its rotation, and after exercising it a little all seems well.

MichaelG.
Too many 'projects'

Rorschach
Posts: 377
Joined: Sun Mar 10, 2019 11:44 am

Re: Olympus IMT-2 mystery

#17 Post by Rorschach » Tue Apr 09, 2024 4:15 pm

I am glad you found a solution! These old scopes sometimes require us to come up with some tricks. We are getting two additional IMT-2 units in the next few weeks. They are refurbished and come with phase equipment. It's finally time to replace those Soviet Union inverted scopes that have been in a dismal shape for years. Various problems with those, I am told by the teachers who had to use them.

MichaelG.
Posts: 4042
Joined: Mon Apr 10, 2017 8:24 am
Location: North Wales

Re: Olympus IMT-2 mystery

#18 Post by MichaelG. » Tue Apr 09, 2024 5:55 pm

Thanks, Rorschach

Please do let me know if either of your new acquisitions has any similar oddities to mine.

MichaelG.